    Introduction

    This chapter deals with HeNe lasers that have their frequency actively regulated either directly, or indirectly via control of intensity. Their output is generally either a single longitudinal mode (single optical frequency), a pair of adjacent longitudinal modes (two optical frequencies separated by 600 MHz to 1 GHz typical), or a pair of optical frequencies close together produced by Zeeman splitting of a single longitudinal mode (250 kHz to 8 Mhz typical) or an Acousto-Optic Modulator (AOM) generating a sideband of a single longitudinal mode (20 MHz typical).

    Unstabilized HeNe lasers - the more conventional type that most people are familiar with from Physics 101 - are covered in the chapter: Commercial HeNe Lasers.

    For current production lasers, the manufacturers' Web sites often provide basic specifications. A Google search is usually the easiest way to find them, but most are also linked from the chapter: Laser and Parts Sources. For older lasers, it's often difficult to obtain detailed specs so estimates based on physical size, and then testing may be the only option.

    An introduction to stabilized HeNe lasers and typical locking schemes can be found in the chapter: Helium-Neon Lasers.

    Comparison Chart of Commercial Stabilized HeNe Lasers

    Here is a summary of the frequency stability specifications for most of the stabilized HeNe lasers that are presently commercially available, or have been in the past. The best of the mode-stabilized lasers may approach the performance of a much more complex and expensive iodine-stabilied HeNe laser (also included in this chart).

    Nearly every model of stabilized HeNe laser ever sold commercially is listed in the chart below. Many are described in separate sections of this chapter, arranged approximately in alphabetical order by manufacturer. For these, the level of detail here is probably several orders of magnitude greater than from any other source, except perhaps the operation and service manual for the laser (which with few exceptions, is generally not available in the public domain). Where there is no entry for a particular laser, a Google search using the manufacturer (with or without model number) will usually locate what little information is available. Sometimes, a research paper referencing the specific laser will even have the most information!

    Unless otherwise noted (below), these data were obtained from manufacturers' Web sites, brochures, spec sheets, or user manuals for each laser. Contributions (including stabilized HeNe lasers I've missed) and corrections are welcome. Please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    All stability values are in parts per billion (ppb).

                                         <--- Frequency Stability Time Scale --->
        Model                    Type/AP  Sec    Min   Hour    Day   Year    Life
    -------------------------------------------------------------------------------
      Aerotech LZR2000             SM M
      Aerotech LZR3000             SM M
     *Aerotech S100 (2)            SM S         +/-1   +/-2   +/-3  (8 hrs)
     *Axsys 150 (8)                DM M  +/-2   +/-6          +/-20 (24 hrs)
     *Coherent 200                 DM S         +/-2 (5 min)         +/-10 (long)
     *Excel 1001A/B/F              AZ M        20 (unspecified time)
      Frazier 100                  I2 S
     *HP-5500A/B/C, 5501A/B        AZ M                       +/-2          +/-20
     *HP/Agilent 5517 (all)        AZ M                       +/-2          +/-20
     *HP/Agilent 5518A, 5519A/B    AZ M                       +/-2          +/-20
      JENAer ZL 600 (10)           DM M                2.0                   20
     *Lab for Science 200          DM S  0.03   0.05   0.2     0.5
     *Lab for Science 210          DM S  0.03   0.05   0.2     0.5
     *Lab for Science 220          TZ S  0.01   0.02   0.05    0.2
     *Lab for Science 260          TM S  0.02   0.02   0.1     0.4
     *Laseangle RB-1 (3)           DM S  0.01   0.1            
      Laser Metric Systems SFL-1   DM M         2 (unspecified time)
      Limtek LS 10.3               GP M        20 (unspecified time)
      LINOS FS Series              DM S         +/-2   +/-10  +/-20
      LINOS FAS Series             DM S         +/-1   +/-2   +/-10
     *Mark-Tech 7900               DM M                       +/-2 (const. temp.)
      Mark-Tech 7910 (6)           DM M                       +/-2 (const. temp.)
     *Melles Griot STP-901 (4)     DM S         +/-1   +/-4   +/-6 (8 hrs)
     *Melles Griot STP-909/911 (2) SM S         +/-1   +/-2   +/-3 (8 hrs)
     *Melles Griot STP-910/912 (2) SM S         +/-1   +/-2   +/-3 (8 hrs)
      Micro-G  ML-1                DM S  0.2 (10 ms)          1.6 (const. temp.)
      NEOARK NEO-262               TZ M          1 (unspecified time)
      NEOARK MODEL-430 (11)        DM S         30 (unspecified time)
      NEOARK 430-R4 (11)           LD S         10 (unspecified time)
      NEOARK NEO-9111   (11)       AZ S                 1     10 (1 wk)
      NEOARK NEO-92SI-NF (11)      I2 S                                     0.025
      NEOARK NEO-OL101K (11)       OL S  0.0001 (10 seconds)
      NEOARK NEO-2MSS (11)         PS S
      NEOARK NEO-5MSS (11)         PS S
     *Newport NL-1 (3)             DM S  0.01   0.1            
     *Nikon NKL-85 (14)            LD S
      Nikon NN-1                   I2 S
      NPL Hexagon (13)             ?? S               0.01                  +/-2
      NPL I2 543 nm                I2 S                                     +/-0.25
      NPL I2 633 nm                I2 S                                     +/-0.2
     *Optodyne L-109               DM M
     *Optra Optralite              AZ M
     *PLASMA LGN-212               AZ M         10 (unspecified time)
     *PLASMA LGN-302               DM S         10 (unspecified time)
     *PLASMA LGN-303               DM S         10 (unspecified time)
     *PLASMA LGN-304               DM S         10 (unspecified time)
     *Renishaw HS-10               DM M     +/-100 (unspecified time)
     *Renishaw ML-10               DM M                +/-20         +/-50        
     *Renishaw XL-80               DM M      +/-50 (unspecified time)
      REO SHL                      DM S         +/-2   +/-4   +/-6
      SIOS SL 02                   DM S         +/-2   +/-10  +/-20
      SIOS SL 03                   DM S         +/-1   +/-2   +/-10
     *Spectra-Physics 117 (5)      DM S
     *Spectra-Physics 117A (4)     DM S         +/-1   +/-4   +/-6 (8 hrs)
     *Spectra-Physics 117C (5)     DM S
     *Spectra-Physics 119 (7)      LD S                       +/-2
     *Teletrac 150-IV (8)          DM M
     *Teletrac 150 (Long) (8)      DM M  +/-2   +/-6          +/-20 (24 hrs)
     *Teletrac 150 (Short) (8)     DM M  +/-2   +/-6          +/-20 (24 hrs)
      Tropel 100                   DM S
     *Tropel 200 (15)              DM S         +/-2 (5 min)         +/-10 (long)
      Winters 100                  I2 S                                     0.025
      Winters 200 (12)             I2 S                                     0.025
     *Zygo 7701/7702 (9)           DM M                +/-2   +/-10  +/-100
     *Zygo 7705 (9)                AZ M                +/-10  +/-20  +/-200
     *Zygo 7712/7714 (9)           DM M                +/-0.5 +/-1
     *Zygo 7722/7724 (9)           DM M                +/-0.5               +/-10
    

    The * denotes lasers that are covered in detail elsewhere in this chapter.

    Type Legend:

    AP (Application) Legend:

    A metrology laser can generally also be used for scientific/research applications since all have very tightly controlled optical frequency. And while the converse is often (but not always) true in principle, it's not usually practical or worthwhile except for experimental purposes since metrology systems may require laser characteristics (like two-frequency) that aren't present in laboratory stabilized lasers. In addition, the optics and cabling/electronics requirements would likely make their adaptation potentially complex, if possible at all.

    Notes:

    1. Whether "+/-" is used for the ppb value depends on how they were listed in the spec sheet, research paper, or other source.

    2. The Melles Griot 25-STP-909/10/11/12 and Aerotech S100, on which they are based, are assumed to have similar performance. The frequency stability specifications listed are from the Melles Griot Web site.

    3. Specifications for the Laseangle RB-1 were obtained from the paper: "Frequency stability measurements on polarization-stabilized He-Ne lasers", T. M. Niebauer, James E. Faller, H. M. Godwin, John L. Hall, and R. L. Barger, Applied Optics, vol. 27, no. 7, 1 April 1988, pp. 1285-1289. The Newport NL-1, which evolved from the Laseangle RB-1 is assumed to have similar specifications.

    4. The Melles Griot 05-STP-901 and Spectra-Physics 117A, which is nearly identical, are assumed to have the same specifications.

    5. While the original Spectra-Physics 117 laser uses essentially the same laser head as the 117A, its electronics differ so the specifications may not be the same. However, the OEM Spectra-Physics 117C probably has specifications similar to the 117A if installed in a suitable enclosure.

    6. The Mark-Tech 7910 seems to be the same laser as the 7900 but with the addition of an internal optical receiver. So, the laser probably has the same specifications.

    7. The SP-119 with the Servo Option had a stability specification of +/-10 ppb per day in a 1964 brochure (which probably predated the actual laser!). But this was improved to +/-2 ppb in the 1966 Operator Manual. The SP-119 without the Servo Option has a stability specification of +/-75 ppb per day using a constant temperature controller without optical feedback.

    8. The Teletrac/Axsys 150 lasers may output either a single circularly polarized mode or a single linearly polarized mode oriented at 45 degrees depending on model/version.

    9. The Zygo 7701/01/12/14/22/24 use dual mode stabilization with an AOM to split the output into two frequency components 20 MHz apart. The Zygo 7705 is an axial Zeeman two-frequency HeNe laser similar in performance to an HP/Agilent 5517D.

    10. The JENAer ZL 600 is a two-frequency HeNe laser which uses the normal longitudinal modes of a ~240 mm long random polarized HeNe laser tube resulting in a difference frequency of about 640 MHz. The laser itself is therefore similar to one like the Spectra-Physics 117/A/C.

    11. The NEOARK MODEL-430 uses laser current for stabilization and a heater for overall cavity length control. The NEOARK NEO-430-R4 is an IR (1.523 um) stabilized HeNe laser using Lamb Dip stabilization. The NEOARK NEO-OL101K requires another reference laser to which it may be locked with an offset of up to +200/-500 MHz. The NEO-2MSS and NEO-5MSS are intensity-stabilized HeNe lasers.

    12. The Winters model 200 is an offset-locked iodine stabilized HeNe laser all in one unit with a fiber-coupled output.

    13. The locking scheme for the NPL Hexagon laser is not known but may be dual mode with enhancements to improve stability. Versions operating at 632.8, 611.9, 594.1, and 543.5 nm are available. Supposedly, these lasers are now manufactured by a separate company, Hexagon Metrology, but there are no references to stabilized lasers on their Web site.

    14. The locking scheme used by the Nikon NKL-85 laser is not known with certainty, but is assumed to be Lamb Dip stabilization based on the characteristics of the laser tube and construction of the laser head.

    15. The Tropel 200 and Coherent 200 are similar if not identical lasers. The specifications for the Coherent 200 are shown.



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Aerotech Syncrolase 100

    Description of the Aerotech Syncrolase 100

    The Syncrolase 100 is an amplitude stabilized HeNe laser that combines the laser tube and locking circuitry into a single compact unit with only DC and/or HV power applied externally. Everything related to stabilization is contained in a small module called the "SFA Adapter" that screws onto the end of a substantially standard 1 or 2 mW HeNe laser head. Included are the polarizing optics and beam sampler, a single photodiode, and all the feedback control circuitry.

    With only a single photodiode sampling the beam, only the amplitude of one of the two polarized modes can be stabilized. Even so, the frequency stability will still be quite good once the laser tube has reached thermal equilibrium and its power has leveled off. As the tube ages and its power declines, the output power from the laser will remain constant until it exceeds what's available from the HeNe laser tube,. At that time, it will lose lock and may even be damaged, more below. With dual mode frequency stabilization, locking will still be possible even when the power output from the tube is very low because it is the difference or ratio of the polarized mode amplitudes that produces the error signal, not a specific value. In addition, with intensity stabilization, the frequency will drift as the tube gets weaker and the lock point moves with respect to the neon gain curve. Why frequency stabilization was not implemented instead, or in addition to intensity stabilization as in the 05-STP-901, is a mystery as it would have been a very straightforward enhancement - primarily a second photodiode! Nearly everything else is already there.

    Lasers based on the Aerotech technology are now sold as the Melles Griot STP1 with specific model numbers of 25-STP-910 and 25-STP-912. [They may also be found as 05-STP-910 and 05-STP-912. Whether "05" or "25" is used simply depends on if it is considered a component (05) or system (25), and sometimes at random!] The Melles Griot lasers are physically and functionally very similar to the Syncrolase 100 but both of these use a separate HeNe laser power supply instead having one inside the laser head. (The Melles Griot 25-STP-909 and 25-STP-911 had the built-in power supply but have been discontinued.) While, it is not known if the electronics are the identical to the S100 version, all indications are that very little has been changed since obtaining the technology as part of the acquisition of the HeNe laser division of Aerotech. Melles Griot calls them frequency stabilized lasers though the description indicates that the same amplitude stabilization technique as the Syncrolase is used. Searching for "Melles Griot 25-STP-910" or "Melles Griot STP1" should return a description and spec sheet. Here is a summary of the specifications for the Melles Griot versions:

    And, if you'd like to order a few, the Melles Griot price (in 2010) is $3,635 each for the low power version (0.5 to 0.95 mW) and $3,861 each for the high power version (0.6 to 1.4 mW). I wonder how they came up with those exact prices.

    The output power is a user adjustment (a trimpot) that sets the intensity stabilization point, and indirectly the operating frequency. In addition to versions based on output power, the Syncrolase came in two versions based on whether a pair of DC wall adapters were used to power an internal HeNe laser power supply and the locking controller, or whether the laser head had a standard Alden connector to attach to an external lab-style HeNe laser power supply, which is included in the price, along with the wall adapter for the locking controller! :) Now, why weren't the two combined, given the warning in the operation manual: "Application of power to the SFA adapter (locking controller) in excess of 5 minutes with the head de-energized may damage the SFA adapter".? There's more on this below.

    One of the unique features of this system is that rather than using a resistance heater over a substantial part of the HeNe laser tube as is done in most commercial stabilized HeNe lasers, it uses a coil surrounding the OC mirror mount stem to directly heat the metal mount via RF induction. A very simple MOSFET driver can provide over 10 W directly to the mount resulting in a very rapid response. Based on tests I've done, I estimate that at maximum RF power, it will increase the temperature of the mirror mount stem itself by greater than 1 °C per second. This is more than an order of magnitude faster than traditional resistance heaters surrounding the glass portion of the tube. A thermocouple in close proximity to the mirror mount stem senses its temperature and is used both to switch the feedback loop on when hot enough, as well as to shut the heater off if the temperature goes too high. Warmup to fully stable operation still takes 30 minutes or so because the rest of the laser has to come into thermal equilibrium as well as the mirror mount stem. But, initial locking is very quick - a couple of minutes - and once locked, it should use less power and be more immune to ambient temperature variations, and the faster response also improves frequency stability. In addition, the use of this technique allows for the possibility at least in principle of converting almost any HeNe laser tube with a suitable mode structure into a stabilized laser by simply attaching the very compact controller to its output end. However, in practice, minor details like the mirror mount stem dimensions and the exhaust tip-off usually being in they way make this rather difficult.

    The DC adapter (either version, 1 or 2 required depending on the model of the laser) is rated 13 VDC, 1.3 A. Measurements show it to have an open circuit output of 16.5 V. The plug is 2.5 mm center positive. I do not know the official specifications for the the external HeNe laser power supply (where required), but based on the length of the tube and other typical Aerotech tubes, it probably around 1,500 V ad 5 mA. Since there is a 7812 +12 V regulator in the controller (see the schematic below), the output of that DC adapter must be greater than about 14.5 V to assure proper regulation. So, at least once the feedback loop is closed, the input voltage should never dip below 14.5 V.

    The HeNe laser tube in the Syncrolase 100 is about 7 inches long. A common 6 to 9 inch tube with cathode-end output (high voltage far away from the electronics!) would probably work except that the mirror mount stem needs to be a about an inch long with the exhaust tip-off cut off close to the end-cap so as not to interfere with the coupling coil assembly. Very few tubes have these characteristics, though some are close enough to be usable in a pinch. In addition, using too long a tube might result in a second longitudinal mode being present if the Output Adjustment is set so the main lasing line is too close to the neon gain center.

    The gate of a power MOSFET is driven by a simple oscillator, running at between 500 kHz and 1 MHz (I measured about 700 kHz on one unit). The feedback signal is summed into the gate junction from the error amp and serves to modulate the output of the induction heater to maintain lock once the operating temperature has been reached. The coil is just short of 9 turns of #24 AWG wire close wound on a 1.35 cm form. Due to the way the leads enter through the back of the form, the final turn is short changed! :) This is probably not terribly critical though.

    For details on theory and implementation see U.S. Patent #4,819,246: Single Frequency Adapter.

    A schematic diagram of the electronics for the Syncrolase 100 can be found at Schematic of Aerotech Syncrolase 100 Controller. This may not yet be quite complete and numerous errors are possible since the PCB is tight, it is a 4 layer board, and the soldermask is almost totally opaque. It was not much fun to trace the circuit. Part numbers are not available for a half dozen components because (1) they might have been obscured and (2) there were several added parts that appear to be in the "oops" category. :-) But I bet this schematic provides infinitely more information than what's available anywhere else! :)

    The RF driver consists of a HEX Schmitt trigger (MC14584BCP similar to a CMOS 40106) with one section used as the oscillator and the remaining sections paralleled to buffer its output. An RC network converts the squarewave of the oscillator to a roughly triangle waveshape at the MOSFET gate. The output of the Error Integrator feeds into the gate as well with the effect of modifying its DC offset. Since the MOSFET gate threshold is fixed, this produces a modulation effect which is a combination of amplitude and pulse width, with the net result of controlling the amount of RF power transferred to the HeNe laser tube mirror mount stem. A significant part of the capacitance in the waveshaping network is the internal input capacitance of the MOSFET gate itself, and this may exceed 1 nF. Thus, it's possible that if the MOSFET needs to be replaced, the value of the capacitor between the gate and ground (C13) may need to be adjusted as well to maintain approximately the same net capacitance and waveshape. The MOSFET gate capacitance can vary by a factor of over 2:1 between MOSFETs with the same part number, or by even more if a MOSFET with otherwise acceptable specifications is substituted. On the unit I have, it was about 1.3 nF.

    Newer versions include a ULN2003 Darlington array for something. But I haven't dug deep enough to be sure if it's part of the MOSFET driving circuitry or replaces the MOSFET entirely. They may also use a semiconductor sensor in place of the thermocouple - it looks like a 1N4148 with no markings. That's probably much cheaper!

    The control functions are implemented in the four sections of a TLC27L4CN quad op-amp as follows:

    When powered up, the temperature sensor is initially cool so the RF driver comes on at full power. When the mirror mount stem reaches the operating temperature (something like 80 °C in 30 seconds or so), the feedback loop becomes active and the Sync LED comes on. However, since the remainder of the laser tube is still increasing in temperature due to the normal heating of the discharge and hasn't reached thermal equilibrium, lock may be lost several times as the overall tube expands and the controller then needs to keep *reducing* the temperature of the mirror mount stem to maintain the distance between the mirrors constant. When the mirror mount temperature gets to be too low, the system will go back to continuous heating based on the hysteresis of the Sync Enable Comparator. After a half hour or so, the laser tube will reach thermal equilibrium and the system will then remain locked forever. (Unfortunately, many people take this literally and leave the laser on until it dies, which is considerably sooner than forever!)

    Here are some photos:

    The coupling coil assembly on the first Syncrolase 100 of mine had disintegrated due to excessive temperature. (Actually the magnet wire and its insulation is in fine shape but the plastic form on which the coil was wound or embedded is no more and it's not even possible to determine much about it.) I've tested the induction heating winding a test coil on a tube made from insulating plastic sheet. The effect is impressive considering the simplicity of the circuitry (see the schematic below) raising the temperature of a dummy mirror mount stem by more than 1 °C per second even with a coil that is probably far from optimal.

    I do not know for sure if the cause of the destroyed coil form was due to a part failure rather than simply a result of the laser being been left on for 7 years continuously! :) The HeNe laser power supply was indeed dead, probably due to the tube being very hard to start and impossible to run for more than a few seconds regardless of power supply or ballast resistance. So it's possible that when the tube decided it was tired of doing its thing and the power supply shorted out, the controller ended up cycling on the over-temperature condition. The Melles Griot manual does warn against running without the laser on. And, electrical tests seem to indicate that the controller is working properly.

    So, the Over-Temperature (OT) adjustment might have been incorrect and too high all along. Since it's not something that affects normal behavior, it would be all too easy to neglect setting it properly! I've also been told by the former owner that this laser always ran very hot. If the tube fails - even if someone forgets to plug in its wall adapter! - the heater tends to be on and bad things can then happen if the OT setting is too high. Ask me how I found out. :( :) OK, I'll tell you. I acquired another Syncrolase with a good tube but that would not stabilize. I traced the problem to what I believe may have been a short in the temperature sensor and then adjusted it to operate at a reasonable temperature set-point. But I accidentally left the controller powered after turning off the laser and went away. When I returned (after lunch!), the entire assembly was too hot to touch and the platic coil form at is cover had melted!!! Apparently, either the OT setting was way too high (it's possible someone before me messed with it) or it isn't effective or was broken.

    Interestingly, on one of those rare occasions where I was able to get the tube to remain on long enough with a lab power supply to watch a few mode sweep cycles, it is a classic FLIPPER! I suppose that the flipperitis could have happened in its old age (it is also weak - about 0.7 mW - and with brown crud in the bore), but normally the flipper or non-flipper status of a tube doesn't change over the course of its life. I do have another Aerotech laser head that would screw right on to the controller but it too is a flipper! :( :) In fact, its behavior shown in Plot of "Flipper" Aerotech OEM1R HeNe Laser Head During First Part of Warmup and the merged version in Plot of "Flipper" Aerotech OEM1R HeNe Laser Head During First Part of Warmup (Combined) looks virtually identical to that of the Syncrolase tube (over the few mode sweep cycles I could see before the tube went out). But, even more interstingly, the flipping of the tube in the plots ceases entirely and it becomes perfectly well behaved once nearly warmed up as shown in Plot of "Flipper" Aerotech OEM1R HeNe Laser Head at Transition to Normal Behavior (Combined). Perhaps that tube was intended for a Syncrolase as it in unusual in having the required long mirror mount stem and short cutoff exhaust pipe. Perhaps it was a reject due to the flipping. Or perhaps for unknown reasons, all these tubes flip when cold. Since the Syncrolase 100 would be operating well beyond this point, there's a chance that the flipping is irrelevant and it would work just fine. In fact, that one working genuine Syncrolase tube is also a flipper until it warmed up! More on this below.

    I built a replacement coil using the wire from the first dead Syncrolase on a roll of plastic. It works, though the temperature response is faster probably because the thermocouple is not in the same location as the original. So, it locks more quickly, but also loses lock more frequently during warmup but is otherwise functional. Perhaps changing the temperature set-point would correct that. It's amazing how much variability can be tolerated with this design.

    Adjusting the temperature set-point is an interesting exercise. Ideally, it should be slightly above the equilibrium temperature of the laser head with only the laser tube powered. Set too high and the laser will run excessively hot, but there will be a fewer number of lost lock events during warmup. Set too low and it may lose lock eventually when the tube equilibrium temperature exceeds the set-point temperature.

    One way to do the adjustment might be to initially set the Temp. Gain pot (R1) fully CW (for a very low temperature) and power *only* the laser head (not the controller) for at least an hour so it reaches thermal equilibrium. Then, power up the controller and slowly turn R1 CCW to slightly beyond the point where the SYNC LED goes out. Monitoring the Temperature Amp output (A1 pin 1) will indicate how effective this is. The voltage on A1 pin 1 should remain between approximately 1.5 V and 2.75 V when the laser is locked. If it goes below about 1.5 V, the feedback loop is disabled and the heater turns on full (SYNC LED OFF). This state continues until the temperature increases to the point where A1 pin 1 exceeds about 2.75 V and the feedback loop is enabled (SYNC LED ON). Better to start out with the temperature set-point adjusted too low should the over-temperature protection fail. :( :)

    I built another temporary coil for the first laser to check it out. This coil is wound on a plastic cylinder found in a junk pile that was glued to the remains of the original coil form. The Epoxy seems to stick rather well, which is a bit surprising. I didn't have any #24 AWG magnet wire, so I used #20, which just fits 9 turns in the available space. The laser works quite well now except that the speed of heating is not quite as fast, possibly due to the coil being slightly longer and larger in diameter. However, this is probably of little consequence in the grand scheme of the Universe. :) Lowering the RF frequency improved the response, though there was no resonance.

    Finally, I built a new coil form for the third laser. It has approximately the same dimensions as the original so it behaves very well. But the plastic is too think and there is very little clearance between the form and mirror mount stem. So the genuine Syncrolase laser head won't fit because its tube is too off-center. (This must have been a result of the way it was manufactured since its beam is well centered.) But my "flipper" head fits just fine and works just fine. :)

    If anyone has an Aerotech Syncrolase or Melles Griot 05-STP-909/910/911/912 laser, dead or alive, that they no longer need, or one that they'd like evaluated), please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    Improving Over-Temperature Protection of the Syncrolase

    Of three similar Syncrolase 100s I've seen, 1 had already melted, 1 I melted, and the other is still fine. :) There are several fundamental deficiencies in the design that can lead to fireworks.

    Another modification (or complete redesign depending on your point of view!) that would enable the Syncrolase (or any thermally-stabilized laser) to run at the minimum temperature to assure reliable operation would to have a temperature set-point that is based on the ambient temperature of the environment, not a fixed setting. In principle, this can easily be accomplished by counting mode cycles from a cold start. Since each mode cycle represents a precise change in temperature, this would enable the laser to operate at a temperature of ambient plus a constant known to be greater than the heating from the laser tube current. A microcontroller could be used for the implementation, left as an exercise for the student. :)

    This of course assumes that the ambient temperature remains relatively constant, but this is often the case with real lab environments. The Zygo metrology lasers with digital controllers compute the number of mode cycles (they call them "mode slews) needed to reach operating temperature based on the actual tube temperature when the laser is switched on, though they may still operate at the temperature required for worst case conditions.

    While the Melles Griot 05-STP-9XX is not supposed to melt down due to a fault condition like not powering the laser, it is not known if the design has actually been improved or rather that they are simply depending careful adjustment of the over temperature pot.

    Using the Syncrolase Controller with Other HeNe Lasers

    I acquired a second Syncrolase 100 on eBay which also had a dead tube, but with a fully intact controller module. So, I tried experiments with two random polarized HeNe laser tubes. These were selected mainly to be cathode output and to have enough OC mirror stem accessible for coupling to the induction heater coil.

    So, while two data points may not be conclusive, it would seem that that almost any tube that can be stabilized using the conventional heating blanket technique can also be stabilized using the Syncrolase controller if its mirror mount stem will fit inside and extend far enough into the induction heater coil. Where the tip-off is not too long but interferes with the coil assembly, simply removing the plastic cover may gain enough clearance. Of course, if you happen to be friendly with the tip-off person at a HeNe laser tube manufacturer, simply ask them to pinch-off and trim the tip-off closer to the tube! :) For longer higher power tubes, the internal preamp gain would need to be reduced to allow the Output Adjust pot to lock at higher power. Of course, for such tubes, the position on the gain curve over which the output is pure single mode would be reduced.

    And flippers will work just fine, thank you. :-) And as noted above, 3 of 3 Aerotech tubes from Syncrolase lasers were flippers, at least when cold!

    Adding True Frequency Stabilized Mode to the Syncrolase

    While Melles Griot describes the 05-STP-909/910/911/912 as a frequency stabilized HeNe laser, it really is not since only a single photodiode is used for optical feedback. However, in principle, it would be very straightforward to convert this laser or the Syncrolase to use dual mode stabilization. The electronics would be trivial: If an amplitude stabilized mode is no longer desired, then it's just a matter of adding a second photodiode in parallel with the first one, with the opposite polarity. Otherwise, an additional op-amp for the second photodiode would be needed to feed the Error Integrator with a switch to enable it.

    The real challenge would be to find space for the modified photodiode assembly. One option would be to replace the polarizing beam-splitter cube with a 45 degree 5 percent plate beam-splitter. Then, mount a 5 mm diameter PBS above this with the photodiodes attached to it. As far as the additional electronics. the newer versions of the Syncrolase, and presumably the 05-STP-909/910/911/912 as well have less space available, but there's probably still enough space to float an IC above the other components. :) Or, if everything were converted to surface mount, there would be plenty of space!



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Coherent Stabilized HeNe Lasers

    Coherent Model 200 Single Frequency Stabilized HeNe Laser

    (Some of these are badged Tropel, which is presumably the company that originally developed the laser before merging with or being taken over by Coherent.)

    The CO-200 is a HeNe laser that operates in a single longitudinal mode. It consists of a cylindrical laser head and separate controller/power supply as shown in Coherent Model 200 Stabilized HeNe Laser. The only front panel control is the power switch on the left. The only indicator is a red "Wait"/green "Ready" lamp on the right. Although the CO-200 design dates from the early to mid-1980s, I've seen systems with post-2000 manufacturing dates, likely for some OEM customer, possibly used in a high accuracy wavemeter as the reference laser.

    The HeNe laser head is powered from a standard Laser Drive 6.5 mA, 2,100 V power supply brick via a HV BNC connector. There is no special control or regulation of this supply - it's turned on by the main power switch. But some thoughtful engineer included a high resistance bleeder to discharge the HV caps in the power supply brick after power is removed. :)

    The HeNe laser tube itself is a Melles Griot (not made by Coherent!) model, labeled 05-LHR-219-158. It has similar dimemsions to an 05-LHR-120, a common 2 mW (rated) random polarized laser. But, the -158 may mean it has been specially selected to have a well behaved mode sweep cycle (not a flipper!) for this application. It may also be filled with isotopically pure gases and an AR-coated HR (to minimize back-reflections from the HR's outer surface). The tube itself puts out more than 2 mW when new - possibly up to 4 mW or even more - but the polarizing and beam sampling optics sucks up some of it. In addition, depending on the particular version, there is either a dielectric filter or polarizing filter in the end-cap. The dielectric filter cuts the output by about half but the this can be varied by 10 percent or so (though I'm not sure if this is intentional or just a byproduct of it being angled). The polarizing filter allows continuous adjustment of output power. (In both cases, the adjustment is done by loosening a set-screw and rotating the end-cap). According to the CDRH sticker, the output beam is supposed to be less than 1 mW. Given the wide swings in output power during warmup (see below), even with 50 percent attenuation, the peak output power may approach 1 mW. But regardless of the type of end-cap, only a single polarization ever exits the laser since the internal beam sampler blocks the other one.

    There is a thin film heater attached to a thick rubber jacket between the tube and laser head cylinder. A beam sampler assembly consists of a pair of Beam-Splitter Cubes (BSCs) in series and two photodiodes, each associated with one of the BSCs. The first BSC is a polarizing beam-splitter and reflects the full power of one polarized mode to its photodiode. Thus, the beam that passes through it is linearly polarized with the orthogonal orientation. The second BSC reflects 10 or 20 percent of this mode to its photodiode. So, the output beam from the laser is pure linearly polarized and has slightly less output power than one of the polarized modes of the tube. The controller monitors the lasing modes and maintain cavity length using the heater so that a pair of orthogonally polarized longitudinal modes straddle the gain curve. The beam sensor assembly can be rotated to align the photosensors with the 2 orthogonal lasing modes as this is arbitrary from tube to tube, and orientation within the cylinder, but should remain fixed for the life of the tube.

    The controller can be set up to run on various input voltages from 100 VAC to 240 VAC by changing the position of a small PCB that plugs into the AC entrance assembly, and plugging in the appropriate fuse. However, it seems that the HeNe laser power supply always runs on 115 VAC from a tap on the main power transformer so it doesn't need to be capable of 230 VAC operation, even though the one that's in there has that option - the wire for 230 VAC is not used! The output of the HeNe laser power supply is rated 2,100 V at 6.5 mA with no start delay.

    The user controls consist of one (1) power switch. There are indicators for AC power and Status. After a warmup period of 20 minutes or so for the laser head to reach operating temperature, the Status indicator will change from Wait (red) to Ready (green). Doing anything that causes lock to be lost will result in a shorter delay of a couple minutes to re-establish it.

    The internal circuitry of the controller box is relatively simple and includes a pair of LM3403 quad op-amps, a 741 op-amp, and LM311 voltage comparator, along with a TO5 power transistor on a heatsink to drive the heater.

    Here is the pinout of the circular control connector as determined by my measurements. There may be errors.

      Pins  Wire Color  Function      Comments
     --------------------------------------------------------------------------
      1,2    Blk/Wht    Heater Power  ~22 ohms
      3,4    Blk/Red    Temp Senseor  ~880 ohms at 25 °C, ~1.2K when locked
      5,6    Blk/Blu    Photodiode 1  Anode is pin 5; Approximately 250 uA max
      7,8    Blk/Grn    Photodiode 2  Anode is pin 8; Approximately 50 uA max
    

    It would appear that the difference in sensitivities is the way it's supposed to be since this was similar on 3 heads. (However, the readings on an analog VOM for the photodiodes did differ on 2 heads I tested - I'm not sure what, if any significance, that has.) This makes sense given that the sampling is done from the main beam. One polarization orientation is blocked entirely and thus the associated photodiode gets its full intensity. The other mode would then seem to be sampled at about 20 percent intensity. The controller and laser head are normally a matched pair and there is an adjustment inside the controller to equalize the responses.

    The heater consists of a serpentine thin file metal pattern on a rubbery backing material that wraps completely once around the tube.

    The temperature sensor extends the length of the tube and is buried within the heater backing, technology unknown.

    I picked up a controller and 3 laser heads in two separate eBay auctions for a grand total of $22.50 + shipping. The serial number on one of the heads matched that of the controller and while this head was initially hard to start, after running it for awhile on my HeNe laser test supply, it now starts normally.

    The controller originally had a dead HeNe laser power supply brick (Laser Drive 314S-2100-6.5-2, 2,100 V at 6.5 mA) which is likely the reason it was taken out of service. I replaced that with an Aerotech LSS-5(6.5) which seems to be happy enough. Using a laser power meter, one of the two modes of the laser (the one present in the output beam) could be seen cycling up and down between about 0.60 and 1.40 mW with the orientation of the beam sensor assembly adjusted for maximum peak power. Each cycle took longer and longer as the tube warmed up to operating temperature, helped along by the heater. After about 15 minutes, it would appear to try to "catch" at certain power levels but couldn't quite remain there. (This behavior may have had nothing to do with the feedback control though.) Then suddenly, after about 20 minutes, the Ready light came on and a few seconds later, it locked rock stable at 0.95 mW. :) A second laser head behaved in a similar manner but with a slightly higher final output power of 1.02 mW. No adjustments were needed inside the controller despite the fact that the second head's serial number didn't match the controller's serial number. Possibly, even better stability or slightly higher stabilized output power could be achieved with some fine tuning. (The 1.02 mW head actually had higher peak power than the 0.95 mW head. The difference is probably in part due to the photodiode sensitivities.) With the fixed filter end-caps installed, the output power dropped to around 0.50 mW. I rather suspect that these are normal power levels for this system. (This was later confirmed when a manual with detailed specifications turned up.) The third head had its cables cut but I finally scrounged a replacement control connector from a box of junk in the garage and jerry-rigged the HV BNC for testing. That laser head now works as well. It also came with an adjustable polarizer in its end-cap. With that installed on either of the other heads, the output power could be varied continuously from near 0 mW to about 1 mW.

    Note that the Ready light comes on and then the laser locks in at the proper phase of the next mode cycle. So, basically the pea brain in the controller (no actual CPU of any kind!) decides that conditions are suitable and enables the feedback loop. The final "decision" is based the cycle duration being longer than some magic number (around 1 minute). :) I've also seen the ready light come on even if the laser doesn't start and when one of the previously locked heads was plugged back in after a few minutes of cooling. In the latter case, the laser was indeed locked though it might not have been able to maintain it continuously since the tube was probably no longer really warm enough.

    There are actually two feedback loops in the controller. During warmup, the heater is driven to a fixed temperature based on the resistance between pins 3 and 4 of the Control connector. Once the period of the mode cycle exceeds a fixed time (guessing somewhere around 60 seconds), the control loop based on the difference of the photodiode outputs is enabled. The same signal that switches from the temperature feedback to mode feedback turns the Wait indicator goes off and the Ready indicator on. More on this in the next section.

    Plot of Coherent Model 200 Stabilized HeNe Laser Head During Warmup and Plot of Coherent Model 200 Stabilized HeNe Laser Head Near End of Warmup show the output power variation due to mode cycling. Note how it seems to "snap" into regulation once the time is right. :) There are roughly 90 mode cycles during warmup prior to lock. The internal optics account for the large variation in output power. The HeNe laser tube itself has a normal mode sweep of only a few percent.

    Another Coherent 200 system I have has a fully functional controller but a fully dead laser head. It is very hard start, impossible to run, and way beyond end-of-life. So, that gave me an excuse to go inside.

    The Coherent 200 laser head can be disassembled in a reversible manner with fewer individual parts than the Spectra-Physics 117/A or the essentially identical Melles Griot 05-STP-901. However, it doesn't come apart as easily, using a press-fit for the tube/heater sandwich.

    As noted above, the tube was found to be way beyond end-of-life. If it could be convinced to start (on a lab power supply), it would not run at any reasonable current and produced no output at all. There was sputtered aluminum coating on the holes near the cathode end-cap and even through holes in the cathode can near the center of the tube. This system had obviously been left on continuously for a large number of years. It was probably not even in use for a good portion of that time, forgotten and lonely in a corner of a lab, wasting its life producing coherent stabilized photons no one was using until there were no more! :) That seems to be the destiny of so many stabilized HeNe lasers. I'll be searching for a suitable replacement tube. The original tube, a 05-LHR-219 (with or without a -158), doesn't show up in any list I've seen) but an 05-LHR-120 has nearly the same dimensions and will run on the same power supply. So, as long as one can be found that is well behaved (non-flipper, wedged HR), it will almost certainly work fine. Other random polarized laser tubes of similar length can also be adapted but may require replacing the HeNe laser power supply and coming up with a creative mounting scheme if diameter is smaller.

    An operation manual and application notes for the Coherent 200 can be found at Ajax Electronics Laser/Optics Manuals under "Coherent".

    Schematics of Coherent Model 200 Controller

    I have reverse engineered the very nearly complete circuit diagram for the Coherent Model 200 controller. The only parts not included in the diagram are the AC line front-end and voltage selection circuitry, the power transformer for the low voltage power supply, the Laser Drive HeNe laser power supply, and the detailed wiring of some of the internal connectors. (The only internal connector on the drawing is J2, for the cable from the PCB to the laser head Control connector on the rear panel. It is labeled J2 on the PCB copper.)

    Everything is in Schematic of Coherent Model 200 Stabilized HeNe Laser. Note that most of the part numbering is totally arbitrary as there were *no* part numbers on the PCB except for the PCB connectors (and I only have J2 in the drawing). This is a late revision with PCB artwork dated 1997, though that probably only means that there was a PCB fab run in 1997, since the artwork itself was obviously hand taped. :) I guess some important customer just had to have more of these lasers made well after they would have been considered very obsolete by Coherent. :)

    The controller has two feedback loops. The Preheat Loop, which is active while the tube is warming up, drives the heater in the laser head to a fixed temperature (set by a pot). The temperature sensor in the laser head is not a common NTC thermistor, but something that increases in value with increasing temperature. It has a resistance of around 800 to 900 ohms at room temperature, but well over 1K ohms at operating temperature. The preheat loop prevents the mode feedback loop from going active until the temperature is sufficiently high. Only after this occurs, does a timer begin to look at mode changes, and switches from the preheat loop to the mode feedback loop once their period exceeds around 60 seconds. The mode feedback loop uses the difference between the orthogonally polarized A and B modes in a simple PI control loop to drive the heater. Should the laser not stabilize as evidenced by mode changes still occurring, the preheat loop will be switched back on to try again. At least, that seems to be how it's supposed to work. However, a system with a laser tube that doesn't start (or a bad HeNe laser power supply) will likely turn on READY shortly after being powered up even though it is obviously not working correctly. Well, I guess it IS quite stable - dead with a frequency of exactly 0.0000000000 Hz and an output power of exactly 0.0000000000 mW! :)

    Adjustment of Coherent 200

    The Coherent Model 200 stabilized HeNe laser controller and laser head are a matched set and do not automagically mate with each other. There are 3 pots inside the controller. Two of them are likely to need some fine tuning when attaching a laser head that was not originally sold with the controller. Although another laser head may appear to lock, performance with respect to stability and ambient temperature range may compromised. There may also have been some subtle changes in laser head construction over the years, particularly with respect to the temperature sensor. Two late model controllers I acquired would not lock with an older laser head due to the temperature set-point being too low. But aside from component drift in the controller, which should be minimal, there is generally no need to adjust the controller over the life of a laser head.

    Here is the adjustment procedure. A multimeter (preferably an analog VOM, with a needle!) or oscilloscope is required. A 14 pin "DIP Clip" will come in handy, and a laser power meter and temperature probe are desirable but not essential. A hex wrench to set the output polarizer orientation and small flat blade screwdriver to adjust the pots will also be needed.

    This should be done from a cold start at an ambient temperature close to where the laser will typically be used. If the laser had been on, it should be turned off and allowed to cool down for a half hour minimum before proceeding.

    A printout of the Schematic of Coherent Model 200 Stabilized HeNe Laser will come in handy.

    Preparation

    1. Near the output end of the laser head is a slot with a hex screw positioned somewhere along its length. This allows the orientation of the output polarizer and beam sampler to be optimally set to line up the polarization axes of the laser tube. If this is still covered or sealed, there is probably no need to touch it as this will not likely change over the life of the tube. Otherwise, or just to be sure, have the proper hex wrench available to turn this screw.

    2. Remove the cover of the controller. This requires taking out two semi-recessed Philips screws in the rear and sliding the chassis out of the box. On some units, the large main filter capacitor will get stuck on the front bezel and require pressing down the PCB to free it.

    3. Connect the laser HV and control cables, and the power cord.

    4. Locate U1 and U2 (my numbering). U1 is the LM3403 furthest in the rear and U2 is the one in front of it. Also locate the labeled test points (TP-A through TP-G) near the front of the PCB.

    Mode A and B adjustment

    The balance between the two polarized modes will affect the location of the lasing line on the neon gain curve. The following sets the two mode amplitudes to be equal, which places the modes equidistant on either side of the gain curve. However, it should be possible to offset the modes if desired, if a different location or slightly more output power in Mode A (the output beam) is desired. However, it's not possible to place either mode precisely at the top of the gain curve.

    1. Attach the positive probe of the multimeter to TP-D (VC+, the midpoint reference). Attach the negative probe of the multimeter to U1-1. (The polarity doesn't matter for a DMM or scope, but for an analog multimeter, this way around will result in a positive reading.

    2. Turn the laser on. If a laser power meter is available, it can be used to monitor the output. If the Ready indicator comes on initially. wait until the Wait indicator comes on, so the laser is in Preheat mode,

    3. As the laser tube heats up and expands, the modes will sweep through the neon gain curve and the voltage on U1-1 will vary periodically from a minimum to a maximum. The laser power meter will show a similar behavior since this is Mode A which is the same mode that exits as the main beam.

    4. If the hex screw in that slot near the end of the laser head is still covered or sealed in place, this step can be skipped as the orientation of the beam sampler with respect to the tube is unlikely to have changed. Otherwise - or just to be sure - loosen slightly - but don't remove - the hex screw. Slide it back and forth in the slot to maximize the voltage swing of Mode A and tighten the hex screw. The maximum will typically be around 1 V and may occur with the screw positioned anywhere in the slot or at one end. On the one I checked, the sensitivity was actually about 1/2 V per mW of laser output power in Mode A.

    5. Note the maximum reading of the Mode A voltage.

    6. Move the probe from U1-1 to U1-14.

    7. Adjust the Mode B pot (R14, next to U1), so that the maximum reading is the same as in step 9, above. (Mode B is blocked by the polarizer at the output of the laser head.) Counterclockwise rotation increases gain.

    8. Go back to U1-1 and check that it is still about the same as before. As the tube warms up, the output power will typically increase somewhat. If this is the case, readjust R14 to make Modes A and B voltage equal.

    9. Attach the multimeter between the case or TP-F (signal ground, negative) and U1-7 or TP-A (the Mode Difference Amplifier, positive) to confirm that its output is swinging from near 0 V to over 20 V along with the mode sweep. Unless the laser tube is very weak, this voltage will probably be at each extreme most of the time and will move relatively rapidly between them.

    Temperature adjustment

    The HeNe laser tube and ballast resistors dissipate almost 12 W (1.8 kV at 6.5 mA). The temperature set-point must be selected such that it is slightly above what would result from the tube and ballast power alone. At an ambient temperature of 18 °C, the required temperature set-point ends up being around 40 °C, a difference of 22 °C. I do not know exactly how this is affected by a change in ambient temperature. If the difference remains constant, the head must run at 62 °C for the maximum allowable operating temperature of 40 °C (from the specifications in the Coherent manual). Such a high operating temperature seems unrealistic.

    One way to estimate the value for the temperature set-point is power only the laser HeNe laser tube (not the heater) by disconnecting the Control cable and allow it to reach thermal equilibrium (at least 1/2 hour). Measure its temperature and then reconnect the Control cable and adjust the Temperature set-point to be about 5 °C higher, or so that the mode sweep goes through an additional 15 full cycles.

    The following assumes an ambient temperature of 18 °C:

    1. Monitor the temperature of the laser head cylinder. If you have a temperature probe, tape it to the center of the cylinder with some thermal insulation covering it. If not, your calibrated thumb will have to do.

    2. Watch the mode changes either with a laser power meter or with the multimeter on TP-A or U1-7.

    3. Once they slow down to a full cycle in 30 seconds or so, check the temperature of the laser head. It should be about 40 °C. If it is lower, slowly turn the Temperature pot (R1, next to U2), clockwise to increase the temperature set-point. This will increase the frequency of the mode changes. If it is higher than 40 °C, turn R1 counterclockwise. If this causes the mode changes to cease, and the laser to lock, it may be necessary to shut it off for 5 minutes to allow the tube to cool a bit. The goal is for the laser head temperature to stabilize just above the maximum ambient temperature of the Coherent Model 200 laser specifications. (If no temperature probe is available, 40 °C, is warm but not hot, slightly above normal human body temperature of 37 °C.)

    4. Give the laser time to stabilize. This will occur once the full mode sweep cycle is more than about 1 minute.

    5. After the green Ready indicator comes on, the laser output power on the laser power meter should quickly settle down and then remain very nearly constant approximately mid-way between the minimum and maximum. The short term fluctuations should be no more than a fraction of 1 percent. Check the voltage on TP-G, the drive voltage to the heater. It should be relatively constant in the range 5 to 10 V.

      Note that the range of 5 to 10 V is my estimate. The Coherent manual shows a graph with the voltage at 12 V at the time of lock (which would then likely drop down to under 10 V after thermal equilibrium). But there is no description or indication of what ambient temperature was used. Perhaps some key piece of information is missing. While there's no problem adjusting the temperature so the laser locks and is stable at any given ambient temperature or a reasonable range around it like +/-5 °C, I don't see any practical way the laser could be set up to operate over the entire 0 to 40 °C range spec'd in the manual without running excessively hot, especially under typical conditions (below 25 °C). It would make more sense if R2 was a sensor for ambient temperature so that the temperature set-point was an offset from ambient rather than actual temperature, but R2 looks like an ordinary resistor.

      • If the voltage is too close to 0 V, the laser will probably lose lock after awhile since the need for heater power tends to decline slightly as all parts of the laser heat reach thermal equilibrium and there's then no way to go low enough.

      • If the voltage is between 10 and 20 V, the temperature set-point is too high and the laser will be running hotter than necessary, but will probably still remain locked.

      • If the voltage continues to fluctuate over a wide range, or remains stuck at 0 V or above 20 V even after Ready has been on for a minute or more, then the laser is unable to lock because the temperature set-point is too low or too high. In this case, the Wait and Ready indicators will probably switch back and forth at a very slow rate (minutes) - forever.

    6. Assuming the voltage on TP-G is in a reasonable range and the ambient temperature is similar to what it will be during use, monitor TP-G for a couple hours. The voltage on TP-G will likely decline slightly as all parts of the laser head come to thermal equilibrium, but should never be less than 3 or 4 V. If it does go close to 0 V, power off for 5 minutes and power on to reset to Preheat Mode, and turn the Temperature pot clockwise a couple turns to increase the temperature set-point. If the voltage is too high, turn the pot a couple turns counterclockwise to lower the temperature set-point. Then check TP-G after it lock. (If the laser isn't off for long enough, the Ready indicator will come on immediately and it may lock immediately. In that case, leave it off for longer and try again!)

      If the laser will be used in an environment where the ambient temperature is much different than where it was tested, readjustment may be needed. The official Coherent Adjustment Procedure (CAP) probably sets the temperature so high that this would not be required over the full spec'd temperature range of 0 to 40 °C, but that shouldn't be necessary unless the laser is to be used near in a sauna. :)

    Mode feedback gain adjustment

    1. The third pot, Loop Gain (R20, near the front of the PCB), can probably be left alone. The controllers I've checked were all set to 0 ohms. It might make a difference with a very weak tube - or it may just be filling a set of holes! But in the case where your tube is way under 1 mW for the peak output, increasing the Loop Gain might help short term stability. Otherwise, there's probably no reason to touch it unless someone before you twiddled everything.

    Finally, power off for 1/2 hour and confirm that the laser will then stabilize properly (after the warmup period) when powered back on.

    One other thing that's recommended while the case is opened is to check R39 and R40, the third and forth resistors from the right in the first row at the front of the PCB. These are the current limiting resistors for the Wait and Ready indicators, respectively, and were originally 510 and 1K ohms, both apparently 1/4 W (by size and appearance). There are other current limiting resistors inside the indicator packages, but the voltage across R39 and R40 may still be high enough to greatly exceed the 1/4 W ratings of the original resistors. If so, the PCB will probably be darkened beneath them as well. Measure the voltage across R39 and R40 when their respective indicator is lit. If either is more than 20 V and the resistor is only 1/4 W, replacement is highly desirable, especially for R40 which will be stressed possibly for years on end. :) Suitable values are 1K, at least 1/2 W for both. Yes, Ready won't be quite as bright but it will be much happier! Proper replacement will require removing the PCB but this is just five screws and several connectors. Space the new resistors off the PCB a bit to further aid in cooling. The PCB is easily damaged, so use a proper desoldering tool to remove the old resistors and clean up the holes. Or just cut the leads off at the bodies of the old resistors and solder to those.

    Replacement of HeNe Laser Tube in Coherent 200

    Some samples of the CO-200 found on eBay have hard-start but usable tubes (if you're very lucky), but most have tubes that are terminally ill and sputter with very low output power (end-of-life), or are simply dead as a glass bottle. The HeNe laser tube in the CO-200 is a Melles Griot 05-LHR-219, which runs on 2,100 V at 6.5 mA. This is physically and electrically identical (or close enough for government work!) to the common 05-LHR-120, a tube with a rated output power of 2 mW, though new ones may approach 4 mW.

    The first step in tube replacement is to find a suitable tube. Melles Griot probably won't even sell you a tube, and if they did, it would cost $300 to $400! Although a common type, this seems to be harder to find surplus than it would appear. Most of those that turn up on eBay seem to be the 05-LHP-120 - the polarized version - which is useless for this purpose. Once a suitable candidate tube has been found, it needs to be tested for non-flipper behavior. A tube that is a flipper may still be useful if the flipping is consistent, or if it disappears when the tube warms up, but a totally well behaved non-flipper is most desirable.

    CO-200 laser head disassembly:

    1. Remove the front bezel. This either unscrews (there may be a set-screw locking it in place) or is held on with a set-screw. The front bezel usually contains an angled neutral density filter.

    2. Remove any black RTV silicone securing the angular adjustment of the beam sampler and then free the beam sampler from the laser head cylinder. It should be possible to pull it far enough out that the 4 conductor ribbon cable is accessible. Cut this about 1 inch from the beam sampler and set the beam sampler aside. Make a note of the position of the rubber heater jacket is inside the laser head cylinder.

    3. Remove the rear end-cap with the cables by unscrewing the single set-screw (it may be concealed by the yellow safety label) and setting it aside in a safe place. Gently pull the end-cap out of the cylinder. If it doesn't come freely, the anode wire is preventing it from moving so it may be necessary to push the entire tube out of the laser head from the front. Stacked pill bottles are useful for this. Since the old tube is dead, extreme care isn't required. But if reasonable force doesn't budge the tube, it may be necessary to free up the outer plastic liner from its 12 blobs of RTV silicone securing it to the inside of the laser head cylinder. There are 12 small access holes through which a dental pick or narrow piece of steel can be wedged to cut it free.

    4. Once the tube/heater assembly has been freed, it will become obvious how clunky this setup really is. The tube is simply held in place by the heater jacket - it's not glued to it. Make a note of the location of the tube relative to the heater jacket. Make a note of the orientation relative to the heater jacker of the photodiode ribbon cable and green cathode wire that runs the length of the laser head

    5. Regardless of whether the outer plastic liner came free initially, it will probably be necessary to remove it so that the replacement tube with the heater jacket can be installed without excessive force. Assuming this is the case, remove as much of the RTV residue from the inside of the laser head cylinder as possible.

    6. Use a 6-32 tap to thread six of the holes from which the RTV was removed (3 at each end) to accept nylon screws. These will be very useful for centering the tube and aligning its beam, especially with the beam sampler, which is rather critical.

    CO-200 laser head reassembly:

    1. Peel off the tape securing the heater jacket to the laser tube and install the replacement tube at the same longitudinal location in the heater jacket and secure with fresh tape. Add the photodiode ribbon cable and cathode wire and secure with tape before installing the outer plastic wrap. I recommend also adding an extra layer of high voltage insulating plastic around the tube running about 2 inches beyond the glass at the anode end. This in addition to the smaller insulating sleeve that was there to begin with. Don't forget to install the anode HV clip on the mirror mount stem and the cathode wire on the exhaust tip-off!

    2. With luck, it should now be possible to push the entire assembly consisting of the tube, heater jacket, plastic sleeve, photodiode ribbon cable and cathode wire back into the laser head. However, this is rather tricky since the anode wire is just long enough to reach to the mirror mount stem, and needs to be relatively straight to minimize the possibility of high voltage arcing, though the dual layer of insulating sleeves should take care of this. Once the assembly is part way in, it may be necessary to pull it the remaining distance from the front. The orientation really doesn't matter that much except that there is a hole in which the set-screw for the rear end-cap seats. Otherwise, the polarization axes will be determined after everything's buttoned up. The beam sampler orientation will take care of optimizing that. Once the rear end-cap is all the way in

    3. At this point, it is probably a good idea to test the HeNe laser tube This can be easily done from the CO-200 controller. There is no interlock so the laser should come on when AC power is applied after a delay of a few seconds even without the control connector attached. In fact, it's best to do it that way as any major oops like a HV short will not destroy the controller!

    4. Assuming a successful test, reconnect the 4 photodiode wires to the ribbon cable. Use heat-shrink tubing to insulate them. Take care not to flex the wires to the photodiodes as they are somewhat fragile.

    5. Replace the beam sampler. Install six nylon screws to adjust the centering of the beam in the output of the beam sampler. CAUTION: Don't force the nylon screws, just gentle finger tightness!

    6. Go to Adjustment of Coherent 200, with the added step that the beam centering may need to be find tuned to maximize the photodiode signals.

    I found an old 05-LHR-121 laser head with a good tube, extracted the tube, and spent way too much time installing it in a CO-200 laser head that had a nearly dead tube. This included a liberal application of duct tape and bailing wire. :) But it works. I knew that this particular tube was a flipper and expected to simply pick the proper mode polarity such that it would lock on the opposite side of the gain curve from the one that flipped. However, it turned out to only flip until it warms up for about 4 minutes or 113 half-mode cycles, then abruptly it stops flipping and becomes well behaved. I have an Aerotech tube with similar behavior, cause unknown.

    I don't think this is in what might be called original condition, but it does start right up without problems (no hard-start tube!) and has decent power (3.2 mW or more total from the tube). It locks normally with 1.2+ mW in a single mode.

    All in all though, much more effort is required to do a tube replacement on the CO-200 than the SP-117/A.

    The controller that went along with this laser head also had minor problems. I had to replace the usual toated dropping resistor for the READY LED but also had to totally rebuild the READY LED assembly itself - both LEDs and their current limiting resistors were fried to a crisp. :( :)



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Excel Precision Stabilized HeNe Lasers

    Excel Precision was another manufacturer of metrology systems based on laser interferometry. It's rather doubtful that are still in business! The last update on their Web site is dated 1998! Excel was never as well known as HP/Agilent or Zygo depite the claim on their moldy homepage "to be the leading manufacturer of laser interferometers". Their product line was much more limited but did include the same types of system components including their own two-frequency HeNe lasers, interferometer optics, environmental compensators, and processing electronics. Laser Interferometry Tools for Precision Measurement is a summary of their product offerings. Many of the components have direct HP/Agilent counterparts. Unfortunately, beyond this, there isn't much in the way of useful information on their Web site except one page on the 1100B 6DOF Calibration System.

    Excel had only a single type of laser, the 1001, a Zeeman-split HeNe laser with a split/REF frequency between 1.5 and 3.0 MHz. But there are at least 2 different case styles. The 1001A and 1001F are about the same size as the smaller HP/Agilent lasers with a similar mounting arrangement, and have connectors and signals compatible with the 5501B and 5517, respectively. However, an Excel laser would be a drop-in replacement for an HP/Agilent laser only if selected for split frequency: 1.5 to 2.0 MHz for the 5501B and 5517A (1001A), 1.9 to 2.4 MHz for the 5517B or 2.4 to 3.0 MHz for the 5517C (1001F). The 1001B is in a larger rectangular case but the specifications appear to be the same as the 1001A/F. The 1001B might be a replacement for the HP 5517A so it's likely that the connector and its pinout, and mounting arrangement are the same, even if the case looks more normal. :) But as with the 1001A/F, to be truly compatible, the split frequency may need to be selected to be in be between 1.5 to 2.0 MHz. Although the Excel model 1100B 6DOF Calibration System is based on the 1001 laser technology, it may use only the internal components of the 1001 laser packaged along with additional optics and electronics in a single enclosure.

    Here are the specifications for the Excel 1001A/F lasers (mostly from the file linked above). The difference(s), if any, between the 1001A and 1001F are probably only in the connectors (5501B for the 1001A and 5517A for the 1001F). While I haven't seen full specifications for the 1001B, the connector is almost certainly 5517-compatible and everything below is probably the same:

    Excel 1001F Two-Frequency HeNe Laser Head

    The Excel 1001F is physically similar to the small 5517 lasers and the cable connector and electrical requirements are also compatible. However, the spec'd REF frequency for the 1001F is from 1.5 to 3 MHz, so it's not really equivalent to any of them. Possibly, one can request a particular REF frequency range via an option, rather than a different model laser. Or, a typical laser starts out close to 1.5 MHz and increases with use and normal power decline until 3.0 MHz is considered end-of-life. The nominal wavelength spec is also the same as that of the 5501A/B, 5517A/B, and 5518A - 632.99137 nm. If it is possible to specify a REF frequency range, then the 1001F could be compatible with the 5517B (1.9 to 2.4 MHz) or 5517C (2.4 to 3.0 MHz) except that the nominal wavelength is slightly different for the 5517C (632.991372 nm versus 632.991354 nm). But this discrepency of about 28 ppb may have simply been a change in the specifications by HP, and is of no real consequence anyhow.

    For a summary of the specifications for the 1001F laser, see the previous section.

    Several photos of the Excel 1001F laser can be found in the Laser Equipment Gallery (Version 3.00 or higher) under "Excel Precision HeNe Lasers".

    The two most interesting ones are:

    Here are some observations/comments:

    So, I need to find a dead Excel laser to go inside. :)

    Excel Power and Reference Cable Pinout

    The Excel 1001A Power and Reference Cable is similar to the HP/Agilent cable for the 5501A and 5501B lasers.

    The Excel 1001F Power and Reference Cable is similar to the HP/Agilent 10791 and the two may be used interchangeably. "Wire Color" is that of the power connections with ring lugs:

             Wire
       Pin   Color    Function
     ----------------------------------------------------------------------
        A*            No Connection on 1001F (MEAS signal level on 5508A)
        B*            ~MEAS (Not used on 1001B)
        C*             MEAS    "          "
        D*            Signal Return (MEAS)
        E             ~REF (Zeeman beat signal from internal optical
        F              REF  receiver's differential line driver.)
        G    Black    Ground
        H    Green    Ground
        J    Orange   +15 VDC
        K    Red      +15 VDC
        L    White    NC (-15 VDC on HP/Agilent cable)
        M             +15 VDC
       N,P            NC (Cable Shield on HP/Agilent cable)
        R             Signal Return (REF)
        S             Ground (Optical Receiver)
        T             +15 VDC (OPtical Receiver)
        U             NC (Cable Shield on HP/Agilent cable)
    

    * Connections to pins A-D are not present on the 1001F cable. For the Excel 1001B, they probably correspond to those shown for the HP 5518A.



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Hewlett-Packard/Agilent Stabilized HeNe Lasers

    General Overview of HP Interferometry Lasers

    Hewlett-Packard (HP) manufactured a variety of stabilized two frequency HeNe lasers for use in metrology applications including very precise measurement of displacement, velocity, straightness, and angle using the wavelength of light as the extremely precise reference. Although these lasers are now sold under the Agilent name, since they were developed by HP and are virtually identical to them, that's the designation we'll continue to use most often. While several other companies have competing product lines, all indications are that the vast majority of HeNe metrology lasers, as well as the associated optics and electronics the explored Universe have been made by HP or Agilent. This statistic is confirmed by some very reliable scientific evidence confirmed by multiple well funded research studies: Most of the metrology lasers appearing on eBay are from HP or Agilent! :-)

    The general approach to precision measurement used by all systems based on two-frequency HeNe lasers such as those from HP/Agilent is shown in Interferometer Using Two Frequency HeNe Laser. The capabilities are quite impressive. A typical example is the HP-5501B laser head from the HP-5501A Laser Interferometry Measurement System, which enables a position/distance resolution down to better than 10 nm (that's nanometer as in 0.000000001 meter!). And that's one of the earliest implementations. More information on inteferometers based on two frequency lasers including descriptions of the optical components can be found in the section: Interferometers Using Two Frequency Lasers. What follows relates mainly to the laser technology.

    Here is a comparison of most of the HP two frequency metrology laser models:

                                         (6)
                    (4,5)  Reference   Maximum   Beam
     Model  Case   Tuning  Frequency   Velocity  Diam.   Comments
    ------------------------------------------------------------------------------
     5500A  Huge :)  PZT  1.5-2.0 MHz  0.4 m/s   6 mm    (1)
     5500B  Huge :)   "      "     "    "   "    6 mm    (1)
     5500C  Huge :)   "      "     "    "   "   6,9 mm   (2)
     5501A  Small     "      "     "    "   "    "  "    (3)
     5501B  Small  Thermal   "     "    "   "    "  "    (3)
     5517A  Large     "      "     "    "   "    6 mm
     5517B  Small     "   1.9-2.4 MHz  0.5 m/s    "
     5517BL Small     "      "     "    "   "     "
     5517C  Small     "   2.4-3.0 Mhz  0.7 m/s 6,3,9 mm
     5517D  Small     "   3.4-4.0 MHz  1.0 m/s   6 mm
     5517DL Small     "      >4.4 MHz >1.1 m/s  6,9 mm
     5517E  Small     "   5.5-6.5 MHz  1.6 m/s   6 mm    (8)
     5517EL Small     "   ???-??? MHz  ??? m/s   6 mm    (8)
     5517F  Small     "      >7.0 MHz >1.7 m/s 6,3,9 mm  (8)
     5517FL Small     "      "    MHz   "  m/s   "   mm  (8)
     5517G  Small?    "   ???-??? MHz  ??? m/s   ???     (8)
     5517GL Small?    "   ???-??? MHz  ??? m/s   ???     (8)
     5518A  Large     "   1.5-2.1 MHz  0.4 m/s   6 mm    S/N below 2532A02139 (2)
      " "    " "      "   1.7-2.4 MHz 0.453 m/s   "      S/N 2532A02139, above (2)
     5519A  Large     "   2.4-3.0 MHz  0.7 m/s    "      (2)
     5519B  Large     "   3.4-4.0 MHz  1.0 m/s    "      (2)
    

    Of the lasers listed, as of Winter 2010, only the 5517A/B/C/D (and their variants) and the 5519A/B are listed on the Agilent Web site as still being in general production and "orderable". The 5517E/F/G (and their variants) may be limited editions for special applications or secret Government projects or to satisfy the fantasies of laser jocks :) since there is virtually no information about them anywhere. There are also variations such as higher power or higher REF/Split frequency for the above lasers depending on options.

    Notes:

    1. The 5500A was the original HP interferometery laser, with built-in linear interferometer optics so that the only required external optical component was a retro-reflector (cube-corner) on the moving tool. The 5500B seems to be physically and functionally similar to the 5500A with the same tube but perhaps upgraded electronics.

    2. The 5500C is the only laser that has a set of 2 optical receivers for an external return beam built-in. Each one uses a pair of photodiodes on either side of the beam, which may be selected to be horizontal or vertical by the orientation of the front shutter plate depending on the interferometer setup. This is part of the reason for the 5500C's huge case. (They all have an internal optical receiver for monitoring the beam to confirm that the beat frequency is present and that there is adequate output power.)

      The 5518A and 5519A/B have a single optical receiver built-in. And of all the HP metrology lasers, the 5519s are unique is having a built-in DC power supply so they simply plug into the wall and feed their REF and MEAS signals to the associated measurement processor/display.

      Like the 5500C, the 5518A or 5519A/B can be used in the normal way (e.g., in a 5528A Laser Measurement System), but are generally intended to be set up stand-alone without any additional optical receivers in a 5529A Dynamic Calibrator. For example, the 5519A laser head can be mounted on a cart and aimed through interferometer optics at a cube-corner (retro-reflector) or plane mirror on a tool whose motion needs to be measured precisely.

    3. The 5501B is a drop-in functional replacement for the 5501A. However, since the 5501B uses thermal rather than PZT tuning, the time from power on to lock is typically 5-9 minutes instead of a few seconds. And the 5501A has an additional connector, not present on the 5501B, so if associated equipment depended on signals from the laser, substitutes would have to be provided - somehow. Going the other way - replacing a 5501B with a 5501A will also work subject to one requirement: After several hours, especially from a cold start, the 5501A may need a "Retune" cycle, normally initiated by a pushbutton on the rear panel or an external signal.

    4. The stabilization in all of these lasers is done by balancing the DC amplitude of the two linearly polarized Zeeman modes, F1 and F2.

      • The 5500A/B/C and 5501A use a PZT to control cavity length so locking is achieved within about 10 seconds. They use a tube assembly which is totally different than those in all the other HP/Agilent lasers including the 5501B. The original patent for the 5500A/B laser tube is: U.S. Patent #3,771,066: Gas Laser and the substantially identical "Divisional Application", now U.S. Patent #3,889,207: Frequency Stabilized Gas Laser. The 5500C and 5501A tubes are slightly different, mainly in that there is a single terminal post for the PZT connection with no waste beam out the back, and the optics (waveplates and beam expander) are part of the tube assembly.

        The polarizing beam-splitter that detects the modes is deliberately oriented so that the separation isn't perfect and a small amount both F1 and F2 are present in each. This results in a beat frequency being generated which is used to produce the reference signal (REF) and to confirm that there is enough beam power to be usable.

      • All the other lasers use a heater inside the laser tube (thermal tuning) to control cavity length so locking typically requires 4 minutes. Laser with the "L" suffix have a vented cover. (Some may spec this as an option like "CO5" rather than an "L".) Though my 5517E also has a vented cover so maybe it's really a 5517EL even if it's labeled 5517E! The beam height for the "L" versions also differs slightly - 80.8 mm instead of 79.5 mm. Can you spell "kludge"? :)

        Although locking typically occurs in 4 minutes (READY comes on solid), some lasers (perhaps the 5517F and 5517G) may require 20 minutes. My 5517E takes about 9 or 10 minutes. But a fully stable frequency output requires 90 minutes for lasers with a non-vented cover or a vented cover but no fan. Those with a vented cover and a fan require only 45 minutes. (It's not known if the temperature set-point for the tube is lower for these compared to the non-vented variety. But if so, this would explain the "Low heat" option since power dissipation would be reduced by running at a lower temperature.)

        From my observations, the frequency oscillates slightly immediately after locking with a period of order of minutes. The amplitude of these oscillations gradually decreases with time and eventually becomes very small. However, the laser likely still meets accuracy specifications during this time.

        An internal optical receiver samples part of the output beam and is used to generate the references signal and to confirm that there is enough beam power to be usable.

    5. All of the lasers that use thermal tuning except for the 5501B have virtually identical electronics, though the Control PCBs are not physically interchangeable between lasers in the small and large cases mostly due to mounting configurations. Some of the newest versions (by date) of the 5517s, have had at least two major redesigns of the Control PCBs (reason unknown). These appear to be much more complex and use FPGA/FPLDs and/or DSPs in place of simple logic. This may have been done by Agilent for some specific option or OEM requirement. The four LEDs on the outside are the same but there are a few more on the Control PCB. On the more common of these fancy Control PCBs, I believe the functions of the jumpers and even the temperature set-point pot are similar. But a much more complex later version (only found on a sample of a 5517E) has changed dramatically.

      The 5501B is the only laser to use Pulse Width Modulation (PWM) rather than pure analog to drive the heater inside the laser tube. This was probably done to reduce power dissipation in the electronics, but does result in modulation of the optical frequency by the PWM.

    6. The usable maximum velocity will depend on the reference frequency (REF) of the specific laser. The values given are the worst case for the minimum of the specification for REF in the direction that reduces the MEAS Dopplar frequency. The approximate maximum velocity for a given laser in meters/second can be calculated using the equation: Vmax=(REF-100kHz)*3.16E-7. The maximum velocity is cut in half with the plane mirror interferometer since it doubles the Dopplar frequency shift (and doubles resolution). (Note that the value of REF really doesn't limit the velocity in the opposite direction, though the electronics will impose restrictions depending on optical receiver bandwidth and data processing implementation.)

    7. The tube assemblies in the 5501B and 5517B/C/D including the tube and magnet, beam expander/collimator, and waveplates, are physically similar, though of course the reference frequencies differ. And the specific collimators will depend on the beam diameter option. The structure on these are all nicely machined aluminum except for the cast base metal base. A diagram is shown in Internal Structure of Hewlett Packard 5501B and 5517B/C/D Laser Tube Assemblies and a photo of an intact one in Tube Assembly Used in HP-5501B and HP-5517B/C/D Two-Frequency HeNe Lasers, and after being taken apart in Major Components of HP/Agilent 5517B/C/D or 5501B Tube Assembly. The tube assemblies in the 5517A, 5518A, and 5519A/B lasers are quite different, being of cast base metal and also much larger, though the actual glass tube, magnet, waveplates, and optics themselves are the same as in the other lasers. The cast assemlies do have more heatsink surface area but and this improves stability, though it seems excessive given the relatively low power dissipation of all these lasers. They do have precise keying pegs so no alignment is needed when swapping tube assemblies. Except for the physical mounting, the diagram, above, applies to these as well. However, it is possible that the length of the Backing Cylinder in front of the OC mirror may depend on the particular model to shorten the cavity for those with a higher REF frequency, but I haven't seen the inside of enough of them yet to know for sure! The 5517E and 5517FL tubes are a bit shorter. I have not seen any 5517Gs but assume them to be similar. See Internal Structure of Agilent 5517E/F/G Laser Tube Assemblies. The 5501B also differs in that the waveplates at the output are oriented to put the F1 and F2 frequency components at 90 degress compared to the other thermally tuned lasers (but is the same as the 5501A). I have yet to hear any explanation, (let alone a credible one!) of why HP changed the orientation after the 5501A/B lasers!

      There's no way to tell the version (e.g., 5517C) or reference frequency (e.g., 2.3 MHz) of the tube itself by inspection of the assembly or from its label. They don't have that information explicitly, only a part number:

      • 5500A: ???.
      • 5500B: ???.
      • 5500C: 05500-60033.
      • 5501A: 05501-60006.
      • 5501B: 05501-60102, 05501-69202 (???).
      • 5517A: 05517-60301, 05517-60501.
      • 5517B: 05517-60201 (???), 05517-68201 (???), 05517-69201 (???).
      • 5517C: 05517-68218 (3 mm?), 05517-68217 (6 mm?). 05517-68249 (3 mm?).
      • 5517D: 05517-68224, 05517-68234 (C29?).
      • 5517FL: 05517-68253.
      • 5518A: 05518-60301.
      • 5519A: 05519-60301.

      The difference in tube part numbers for same model lasers isn't entirely clear. It may be a combination of the size of the beam optics and other special features like a particularly high REF frequency or high output power option.

      As noted, the older 5501A and 5500C use physically identical tubes with PZT tuning. The tubes in the 5500A and 5500B are the same and functionally similar to those in the 5501A and 5500C, but the construction differs enough to make it impractical to substitute for those. None of these tubes are compatible with any of the other lasers. The chart below shows the Physical (P) and Reference frequency (R) compatibility of the various thermally-tuned HP/Agilent lasers:

                                      (a) (b)
                   5   5   5   5   5   5   5   5   5
                   5   5   5   5   5   5   5   5   5
                   0   1   1   1   1   1   1   1   1
                   1   7   7   7   7   8   8   9   9
            (c,d)  B   A   B   C   D   A   A   A   B
       -----------------------------------------------
        (e) 5501B  PR   R  P   P   P    R
            5517A   R  PR              PR  P   P   P
            5517B  P       PR  P   P        R
            5517C  P       P   PR  P            R
            5517D  P       P   P   PR               R
        (a) 5518A   R  PR              PR  P   P   P
        (b) 5518A      P    R          P   PR  P   P
            5519A      P        R      P   P   PR  P
            5519B      P            R  P   P   P   PR
      

      Notes:

      1. 5518A S/N below 2532A02139.

      2. 5518A S/N 2532A02139 and above.

      3. Where both "P" and "R" are present in the matrix, it should be possible to install the tube and achieve original laser performance specifications. Where only "P" is present, the tube may be installed but the performance specifications will be those associated with the tube. changes.

      4. The metal casting of older 5517A tube assemblies may require minor trimming if being installed in 5519A/B lasers to provide clearance for the internal switchmode power supply.

      5. If installing a 5501B tube in a 5517B/C/D laser or vice-versa, the waveplates must be adjusted to rotate the output polarization 90 degrees and swap the axes of the F1 and F2 modes - which are interchanged compared for the 5501 compared to the 5517 (reason unknown). The laser will not lock at the proper location - or at all - otherwise since the slopes of the split gain curves at the lock point would have the wrong sign. And even if it did lock somehow, the sign of position changes would be reversed. I had originally assumed this would just involve rotating the inner barrel of the half waveplate (the waveplate assembly at the very end of the tube assembly) 45 degrees. But on the one sample I tried, this seemed to result in impure modes. So, perhaps that waveplate isn't 1/2 wave in the 5501B, or it might have needed adjustment of the outer barrel to make it pure 1/2 wave. Or, to avoid any need for waveplate tinkering, simply swap the drive signals to the LCD on the beam sampler PCB.

      6. It is not known whether 5517E/F/G tubes may be swapped with other 5517 tubes, or among themselves. Although the 5517E case is the same size as the others, its tube is slightly shorter. But it does come with a series resistor to make the heater resistance similar to the others.

    8. Specifications for the 5517E are mostly unknown except for the beam diameter and REF frequency (estimated based on measurements of one sample) and the corresponding maximum velocity (1.6 m/s). The REF frequency and maximum velocity is even higher for the 5517FL - greater than 7.0 MHz and 1.7 m/s, respectively (7.2 MHz with the -300 option). Nothing at all is known for the 5517G/GL. See the section: Agilent 5517E/F/G.

    9. Minimum output power:

      • 5500A: 120 µW.
      • 5500B: 120 µW.
      • 5500C: 120 µW.
      • 5501A: 120 µW.
      • 5501B: 180 µW.
      • 5517A: 180 µW.
      • 5517B: 180 µW. Option H10 is 400 µW and 2.2 to 2.4 MHz REF. There is/was also an option H07 which may also be 400 µW (as with the 5517C).
      • 5517BL: 120 µW.
      • 5517C: 180 µW. Option H03 is 300 µW, option H05 is 240 µW, and option H07 is 400 µW.
      • 5517D: 180 µW. Option H01 is 300 µW, option H03 is ??? µW.
      • 5517DL: 120 µW. Option H01 is 300 µW.
      • 5517E: ??? µW?
      • 5517EL: ??? µW?
      • 5517F: ??? µW?
      • 5517FL: 65 µW.
      • 5517G: ??? µW?
      • 5517GL: ??? µW?
      • 5518A: 180 µW.
      • 5519A: 180 µW.
      • 5519B: 180 µW.

    10. Additional options. Many of these are not documented anywhere. In fact, even with a laser in-hand having some of these options, it's difficult or impossible to tell what they mean! Sometimes they are for a special high power or slightly higher REF frequency. For the most part, if a given model laser (e.g., 5517C) has the desired beam diameter and acceptable output power, it will work in most tools calling for that model laser. Or, they may be a combination of other options. Only in rare instances will there be issues with something like a higher than spec'd REF frequency:

      • 5500A: ???
      • 5500B: ???
      • 5500C: ???
      • 5501A: ???
      • 5501B: ???
      • 5517A: 003, 006.
      • 5517B: 003, 006, 009, 030.
      • 5517BL: ???
      • 5517C: 003, 006, 009, 030, C01, C05, C07, N05, X01, X02.
      • 5517D:006, 009, C01, C05, C06, C07, C11, C13, C15, C16, C19, C39, N06, N07, N13.
      • 5517DL: 006, 009, 039, 300.
      • 5517E: C01.
      • 5517EL: 003, 006, 009, 030, 039.
      • 5517F: ???
      • 5517FL: 003, 006, 009, 030, 039, 300.
      • 5517G: ???
      • 5517GL: 006, 009, 030, 039.
      • 5518A: ???
      • 5519A: ???
      • 5519B: ???

      Legend:

      • 003: 3 mm beam.
      • 006: 6 mm beam.
      • 009: 9 mm beam.
      • 030: Electrically insulated feet.
      • 039: Standard mounting.
      • A6J: ANSI/NCSL Z540-1-1994 compliant calibration.
      • C01: Long range (5519).
      • C39: 9 mm beam and 300 µW output (5517D).
      • STD: Standard performance.
      • X01: Combination of 006 and STD.
      • X02: Combination of 003 and STD.

      If anyone has additional info defining what these other options mean, please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    There are photos of various HP/Agilent metrology lasers in the Laser Equipment Gallery (Version 2.32 or higher) under "Hewlett Packard/Agilent HeNe Lasers". These include the 5501A laser head and tube, the 5501B laser head, (which is physically similar to the 5517B/C/D except for the connectors), and the 5517A, 5519A, and 5519A/B laser heads. I think the older 5501A tube looks much cooler than the newer ones. :)

    The most significant difference between the various lasers is in the Zeeman split reference frequency. A higher frequency enables a faster slew rate for position and velocity measurements. As of Winter, 2010, all the 5517s and 5519s are current Agilent products. General information, descriptions, and specifications may be found by going to Agilent and searching for "laser positioning laser heads" or a specific model number like "5517C". Some of the specifications from the datasheet:

    These sound quite incredible but 1 ppm is a frequency of about 474 MHz (1/1,000,000 of 474 THz, the optical frequency corresponding to a wavelength of 633 nm). Thus 0.1 ppm is 47.4 MHz, 0.02 ppm is 9.5 MHz and 0.002 ppm is 0.95 MHz. So, still impressive, but quite reasonable for a well designed stabilized HeNe laser. However, what is somewhat unique about the 5517 and some of the other HP/Agilent lasers is that this absolute accuracy is achieved without the need for any periodic testing or adjustments, by virtue of the design of the mode sampling and locking electronics.

    With respect to selecting among the various laser models, if your application has no need for the higher REF frequency (often called the split frequency), there is no advantage to getting a laser like a 5517D as opposed to a 5517B. In fact, the lasers with a lower REF frequency tend to have higher output power and thus may be easier to set up and align especially in multiple-axis configurations. They also tend to be less expensive on the surplus market, though the Agilent price isn't that much different. The only disadvantage of a laser with higher output power is that there can be enough of a detected MEAS signal due to slight angular misalignment of interferometer optics like the 10706A to result in a reading even if the beam to the tool or whatever whose position or velocity is to be measured is blocked or misaligned. The interferometer cube contains a polarizing beamsplitter and if the F1 and F2 orientation are not precisely aligned with the polarizer, there will be a small amount of F1 mixed with F2 and vice-versa even without the reflection from the mirror on tool. With a 400 µW laser and single axis, the required angular accuracy to avoid a false MEAS signal is well under 1 degree with the optical receiver threshold at its default most sensitive setting. And even if the alignment is perfect, polarizing beam splitter and AR coatings are not perfect so there can still be residual mixing. None of this matters once the return beam is aligned since the MEAS signal will be much stronger than the bogus one, but it can be confusing. Increasing the threshold may be desirable to avoid the issue.

    And a note about that impressive spec'd lifetime of 50,000 hours - about 6-1/4 years of continuous use. HP lasers used to last a long long time and it wasn't unusual to find an HP laser running fine after 8 years. But I rather suspect this is no longer the case. I've seen many late model (2004 to 2006) Agilent 5517s that were going down hill well before 6 years including at least one that was essentially dead after less than 3 years. These were standard 5517Bs or 5517Cs pulled from semiconductor fabs, either because they failed in normal use, or because they were rejected during preventive maintenance due to low power or the REF frequency going out of spec (which is usually related to the power decline). Thus, even a late manufacturing date is no longer assurance of a healthy laser. Nor would even a close inspection of the HeNe laser tube, as the they appear identical except for the Agilent label - perhaps that's enough! So if you are buying these things new, it probably pays to go for the extended warranty. :)

    HP/Agilent Laser Principles of Operation

    All the HP/Agilent laser heads use what is called an axial Zeeman-split HeNe laser, which produces two orthogonally polarized modes differing in frequency by between 1.5 and 4 MHz for most models. The beam is single spatial mode (TEM00) though the outer tail may be clipped by the beam expander. The two polarized modes are actually a pair of lasing lines operating on the same longitudinal mode of the laser cavity but slightly shifted due to the Zeeman effect resulting from the laser tube operating inside an axial magnetic field. However, it's not strictly accurate to think of the longitudinal mode oscillating on its own and then somehow being split into two sub-modes by the magnetic field. Rather, it is the Doppler-broadened neon gain curve itself that splits and the two halves are actually separated by 100s of Mhz. Lasing on one of these gain curves will result in right-hand circular polarization and lasing on the other will result in leff-hand circular polarization. Then, any given longitudinal mode will see some amount of gain from each of the split gain curves, and if above threshold, will lase there. This means that the location where the two Zeeman split modes are equal in amplitude is not at the peak of the neon gain curve but part way down as shown in Axial Zeeman Split HeNe Laser Mode Behavior, a simplified diagram like one probably found in a textbook. (More below.) Note that while the cavity may be long enough for more than one longitudinal mode to be present (as in the diagram), only when the a longitudinal mode is approximately centered on the Zeeman-split neon gain curve will it be able to create a Zeeman-split mode pair and beat signal in the frequency range that matters (a few MHz rather than many 100s of MHz). In addition, in geniune HP/Agilent lasers, conditions are such that when locked (READY on solid), only the one pair of Zeeman-split modes can oscillate (no "rogue" modes). Rogue modes can potentially result in problems in the interferometer.

    Waveplates at the output of the HeNe laser tube convert the left and right-hand circularly polarized Zeeman split modes to linearly polarized modes that are orthogonal and aligned with the horixontal and vertical axes of the laser. These two modes usually differ in optical frequency by between 1.5 and 4 Mhz (depending on the specific laser). (Some recent versions of the 5517 may actually go to 6 MHz or more.) The X and Y polarizations are sent down different paths in the metrology application. One is generally a reference length and the other is the distance to be measured or tracked. (It's the change in path length difference that matters so they could both move if desired.) The two beams rather than creating an interference pattern are beat together in a detector that outputs a difference (or heterodyne) signal. If the relative distance between the two beam paths changes by one half wavelength of the laser (about 632.8 nm but accurate to many significant digits!), the phase of the difference signal will change by 360 degrees. The laser also generates an electrical signal from beating the signals together internally. This constant reference is compared to the detector signal and an electronics package counts off the phase shifts and uses it to determine the distance traveled.

    A moderately powerful cylindrical permanent magnet does the Zeeman splitting resulting in a pair of circularly polarized outputs at two very slightly different frequencies. F1 is designated the lower frequency and F2 is designated the higher frequency. For the 5501A/B, F1 is vertical (perpenticular to laser base) while F2 is horizontal (parallel to laser base). For the 5517A/B/C/D, 5518A, and 5519A/B, F1 is horizontal (parallel to laser base) while F2 is vertical (perpendicular to laser base). (Exactly why HP switched orientations between the two model series is not clear as there is no benefit to one over the other and it just causes confusion.) The difference between F1 and F2 ranges from 1.5 to 4 MHz for most of the HP/Agilent lasers depending on the model (as listed above) and also the specific sample of the laser. The cavity length of the early HP lasers (5500A/B/C and 5501A) is PZT controlled using feedback based on the amplitude of the two modes. For the 5501B and all later lasers like the various 5517s, cavity length is adjusted by a heating coil wrapped around the bore inside the tube. In all cases, the feedback is used to maintain the position of the lasing lines symmetric on the Zeeman split neon gain curves as shown in Axial Zeeman Split HeNe Laser Mode Behavior. A 1/4 waveplate converts the circular polarized output to orthogonal horizontal and vertical polarized components which are used externally. F1 is reflected from the thing being measured or tested (e.g., disk drive servo writer) and F2 is reflected from a fixed reference. The difference frequencies (F1-F2) and (F1-F2)+dF1 are then analyzed to determine precise position, velocity, or whatever. This approach has lower noise, greater stability, and is therefore more accurate than the common single frequency interferometer. By using cavity length control to force the mode position to be centered, the actual optical wavelength/frequency can be set very accurately. Using the MHz range beat signals makes signal processing straightforward and is more immune to noise than the baseband optical signals.

    Interestingly, the actual beat or reference frequency (REF) does NOT need to be super stable over the long term. Rather, it is the difference between REF and the return signals that matters and that only depends on the motion of the target reflector, the optical frequency of the meausrement beam, and the speed of light. Thus, although the optical frequency needs to be known to high precision (+/-0.1 ppm for the standard lasers; +/-0.02 ppm for those calibrated to MIL STD-45662), the exact beat frequency of each laser is not precisely controlled or even precisely measured and recorded or used anywhere in the calculations. This is one reason why the listings above include only a range of values. Any given sample will operate somewhere within that range but the exact value is somewhat random depending on the specific characteristics of the tube/magnet assembly and the specific place on the neon gain curve that the lasing line is parked. In fact, REF tends to increase over the life of the laser as the gain and thus output power decline with use. As long as REF remains within the spec'd range for the specific model laser, then the system in which it is installed will work properly. What exactly a machine will do if REF goes out of range is implementation dependent, but one reason for a laser such as this to be replaced is for REF to approach or exceed the high end of the spec'd frequency range.

    While one might think that locking the difference frequency to a crystal reference would be even better - and the technique is patented - it's not clear that this would be better and might actually be worse. The difference frequency relative to the mode position can change for any number of reasons. But forcing the amplitude of the two modes to be equal as is done in the HP lasers centers the modes on the split neon gain curves, which should be very stable in terms of wavelength.

    All of the HP lasers use conventional dual polarization mode stabilization to lock the lasing lines to the split neon gain curve. However, the two signals are not from adjacent longitudinal modes as with most common laboratory stabilized HeNe lasers, but are the two Zeeman split sub-modes differing in frequency by a few MHz instead of many 100s of MHz. In fact, both are the same cavity mode but shifted slightly higher and lower than would be predicted by c/2*L. One twist on the implementation is that the 5501B and those below it on the chart use a Liquid Crystal Device (LCD) polarization rotator to alternately sample the horizontal and vertical polarized modes, and subtracting sample-and-hold to compare them in the error amp driving the heater, rather than the polarizing beamsplitter and dual photodiodes used in many other dual polarization mode stabilized lasers including the 5500C and 5501A. The LCD approach does have a sort of elegance as well as practical benefits. Since the same optical path and photodiode is used for both polarization modes, the sensitivity is identical so the mode balance should be perfect assuming the LCD polarization rotation is 90 degrees. Since the intent is to park the modes symmetrically on the split neon gain curve, this is perfect and thus requires no offset adjustment over the life of the laser as the output power of the tube declines. And, the LCD and associated electronics may in fact be cheaper than a high quality polarizing beam splitter. However, it also creates some artifacts as a result of the digital switching, resulting in small cyclical variations in optical frequency over a period of 2 or 3 seconds. These are of no consequence for most metrology applications, but do detract from the elegance of these lasers.

    In fact, the thermally tuned lasers have only one adjustment associated with stabilization, and that is for the temperature setpoint at which the controller switches from pre-heating to optical locking. The resistance change of the actual heater coil is used to sense temperature and there may apparently be variation from one tube to the next. But this is an extremely non-critical setting and won't affect accuracy, only possibly the temperature range over which the laser will remain locked. (5517s with the Newest Digital Control PCB may have no adjustments, or at least none that are obvious!)

    One oddity with respect to the thermally tuned laser tubes is the patent reference that appears on the label of all newer ones at least: "Licensed by Patlex Corporation Under Patent No, 4,704,583". The title of this patent is: "Light Amplifiers Employing Collisions to Produce a Population Inversion", filed in 1977 but not granted until November of 1987. The most curious thing is that there appears to be very little of relevance in the patent other than its association with laser action! Nothing in the patent diagrams or text has any obvious connection to the tube assembly design. In fact, the exact same text exists on other more mundane things like a Carl Zeiss-badged Siemens LGK 7634, a bog standard 2 mW random polarized HeNe laser head. I've heard that Patlex is actually a bunch of lawyers and I bet they made out or are making out quite well. :)

    Speculation on Axial Zeeman HeNe Laser Behavior

    I've yet to see an explanation of the axial Zeeman effect that isn't either total hand waving or pages of hairy math! I can hand-wave with the best of them, but would welcome something more satisfying. :) All fundamental physical phenomena have fundamentally simple explanations that at least capture the spirit, if not all the details). (OK, well, except perhaps for high energy physics where hypothetical particles and extra dimensions are created at the pleasure of high energy physicists who need to attract equally high energy grant money.) I'll even accept that an axial magnetic field splits the neon gain curve and results in the separate shifted gain curves for left and right circularly polarized modes. But why do the two lasing lines move apart?

    There are two possible "simple" causes of the lasing frequency shifts resulting from Zeeman splitting: Mode pulling (which tends to attract each lasing mode towards its respective gain center) and magnetically-induced birefringence of the plasma (which results in the effective cavity length differing for each lasing mode's polarization). Normal mode pulling seems unlikely as its effect in a HeNe laser is typically in the 10s of kHz range at most, 2 orders of magnitude lower than the MHz difference frequency present with these lasers. So, that would leave birefringence, but birefringence effects are even worse at predicting some of the observed behavior. For example, why does cavity loss affect split frequency so dramatically?

    The following explanation may be totally bogus but it has the attractive property that using some not-so-hairy math, it is able to predict the approximate behavior of real Zeeman lasers. So here goes:

    The mechanism for the shift of the Zeeman modes away from the cavity modes is a type of mode pulling. Normal mode pulling is a form of mode competition where two or more lasing modes are competing for the same population of excited particles. The lasing modes will then tend to be attracted slightly toward the center of the gain bandwidth curve. This will be true in a solid state laser where the gain is homogeneously broadened - all atoms are in the same environment since they are fixed in a solid, usually a crystal lattice. The only time it will be true with a HeNe laser with a Fabry-Perot cavity is where modes are equally positioned on opposite sides of the Doppler-broadened neon gain curve. Since the Doppler broadening is due to motion, photons traveling one way will stimulate excited atoms on one side of the gain curve and photons of the same frequency traveling in the opposite direction will stimulate excited atoms on the opposite side of the gain curve. Even though the Zeeman-split modes are shifted in frequency with respect to each-other, they satisfy this condition when they are near the center of the split gain curves - which is the normal locked position.

    The basic mode pulling equation (with gobs of assumptions!) is:

                         GB
      FS = FSR * (1 - ---------)
                       GB + CB
    

    Where:

    CB will be equal to FSR/Finesse for high-R mirrors with equal reflectivity, Finesse = π*sqrt(R)/(1-R). Where one of the mirrors is HR as is the case with HeNe lasers, this will actually be near 2*π*sqrt(R)/(1-R).

    There should probably be a factor of 2 thrown in since what we're interested in is not the shift of a single mode, but the change in distance between the two modes due to both of them shifting by the same amount in opposite directions. However, this can arguably be offset by the fact that the modes are positioned about half way down the gain curves, not far down where the mode pulling effect would be greatest. So, accept that this hand waving cancels out. :)

    After plugging in some numbers, here are 2 lasers with OC mirrors with reflectivity that has been measured (dissected Melles Griot 05-LHR-006 barcode scanner tube and a 5517B tube) and several others where the reflectivities can be estimated based on difference frequency (REF) specifications (5517A/C/D/E/F/G). There is also a Melles Griot 05-LHR-007 (same as the Spectra-Physics 007), which is about the shortest commercially available barcode scanner tube:

    Barcode scanner tubes:

                  Cavity              OC Mirror   Cavity   |<--- F2-F1 (REF) --->|
      Tube Type   Length      FSR    Reflectance  Finesse   Predicted   Measured
     ------------------------------------------------------------------------------
      05-LHR-006  135 mm   1.071 GHz    0.990       625     1.23 MHz   1.2-1.6 MHz
      05-LHR-007  110 mm   1.360 GHz    0.990       625     1.85 MHz   1.5-1.7 MHz
    

    HP/Agilent lasers:

                  Cavity              OC Mirror   Cavity   |<--- F2-F1 (REF) --->|
      Tube Type   Length      FSR    Reflectance  Finesse   Predicted   Specified
     ------------------------------------------------------------------------------
        5517A     127 mm   1.180 GHz    0.988       520     1.67 MHz   1.5-2.0 MHz
        5517B      "  "      "   "      0.985       416     2.09 MHz   1.9-2.4 MHz
        5517C      "  "      "   "      0.980       311     3.00 MHz   3.0-3.4 MHz
        5517D      "  "      "   "      0.975       248     3.50 MHz   3.4-4.0 MHz
        5517E    101.6 mm  1.475 GHz    0.973       230     5.90 MHz   5.5-6.5 MHz
        5517F      "  "      "   "      0.968       193     7.01 MHz   7.0-??? MHz
        5517G      "  "      "   "      0.965       176     7.67 MHz   ???-??? MHz
    

    Note that the predicted values assume a constant (though not known) magnetic field, but its actual strength also strongly influences F1-F2. However, based on measurements of many HP/Agilent lasers, the magnetic field may differ by up to nearly a factor of 2 depending on model and specific sample. Higher REF frequency lasers on average have stronger magnetic fields, but there are some notable exceptions! (See the section: HP/Agilent 5517 Laser Construction.)

    The actual mirror reflectivity for the SP-007 tube is not known. Being so short, it may be slightly higher than 0.99, in which case the predicted value of F1-F2 would be lower.

    OK, so I kind of picked the reflectivities for the 5517A/C/D/E/F/G mirrors to make the results reasonable. :) With the increasing cavity loss, the output power of lasers with higher REF frequencies will tend to be lower. However 0.965, or even 0.975 may simply be too low to lase at all or with useful power on a tube of this length. So, perhaps I should compromise on that factor of 2 gave up above. :) However, the actual change in discharge length going from the 127 mm to the 101.6 mm cavity is small, perhaps around 10 mm, so not that much gain is lost. Eventually, I will measure the reflectivity of OCs for other HP/Agilent lasers. But I don't have any tubes that I'm willing to take to bits at the present time, partially due to (1) the physical and emotional trauma that would result and (2) the fact that I haven't located the special chants and incantations required for metrology laser tube sacrifice. :) If anyone has done this, has certifiable 5517A/C/D tube bits, or has a 5517 tube that's already cracked or broken they'd be willing to contribute to the cause, please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    This approach also predicts that the shift and thus beat frequency will be a maximum with the modes centered as they are when locked. Of course, all these nice results based on numerous assumptions may be wishful coincidence, but they are close to what is observed and don't require delving into esoteric plasma physics. Whew! :)

    In addition, measurements suggest that the magnets used in the various 5517A/B/C/D/E lasers and 5501B tend to have a higher strength for higher beat frequencies, but not always. So, mirror reflectivity alone is probably not used for beat frequency selection and there may in fact be some "mix and match" going on mating tubes with magnets to achieve the desired result.

    Now back to your regularly scheduled programming. :)

    HP/Aglilent Laser Output Power and REF Frequency versus Use

    With common HeNe lasers, output power may decline slightly with thousands of hours of use and then drop precipitously near end-of-life. However, with some, power will actually increase initially due to the tube being overfilled. And in many cases, the decline in output power is mostly due to misalignment of the mirrors that occurs after many power cycles. This is fully recoverable with realignment. There is of course, no REF frequency for such vanilla (or cherry!) flavored HeNe lasers. :)

    I do not have any full life data for HP/Agilent lasers. But in general it would appear that the trend is for output power to decline, usually accompanied by an increase in REF frequency, especially near end-of-life. But not always - and that's somewhat of a mystery. Since HP/Agilent lasers have mirrors whose alignment is permanently fixed by glass-glass contact, changes in alignment cannot be the cause for either power decline or REF frequency increase. However, the gas-fill He:Ne ratio and pressure does change with use which affecting gain. And, optics can degrade increasing cavity loss. The strength of the permanent magnetic field also affects output power and REF frequency, but if it were to change at all, it would most likely decrease, resulting in an opposite effect compared to what's in evidence here. And while there may be a very small effect from external losses in the optics due to changes in beam expander alignment and contamination over the years, it's almost always insignificant.

        Parameter       Output        REF
       (Decreases)       Power     Frequency
     ----------------------------------------
          Gain         Decreases   Increases
       Cavity Loss     Increases   Decreases
      Magnetic Field   Increases   Decreases
      External Loss    Increases   Unchanged
    

    The following table shows how the output power and REF/split frequency changes with use for recent vintage (2004-2006) Agilent 5517B lasers. Though the actual number of on-time hours is not known, these lasers were all labeled by Agilent with the original output power and REF frequency. ID #1 is believed to either be grossly mislabeled, or an unused spare, which thus has like-new performance. All the others have probably been run for thousands of hours. Several have REF frequencies that have climbed to near or beyond the upper limit for the 5517B (2.4 MHz) and were probably pulled from service for that reason. But on some others, while the output power has declined significantly, the REF frequency is unchanged or even lower than the labeled value. So, it's possible that there is indeed more than one mechanism accounting for the changes in output power and REF frequency. With gain and cavity loss being critical, each of these will degrade at different rates.

      Laser  Label PWR  Warm PWR  W-PWR/  Label REF  Warm REF  W-REF/
       ID      (µW)       (µW)    L-PWR     (MHz)     (MHz)    L-REF
     -----------------------------------------------------------------
        1      355        395      1.11      2.27      2.17     0.96
    
        2      580        525      0.91      2.29      2.39     1.04
        5      560        510      0.91      2.29      2.29     1.00
        6      607        525      0.86      2.22      2.27     1.02
        7      580        516      0.89      2.22      2.32     1.05
        9      680        530      0.78      2.20      2.47     1.12
       12      251        222      0.88      2.19      2.17     0.99
       13      560        305      0.54      2.23      2.75     1.23
       16      654        519      0.79      2.23      2.51     1.13
       17      660        574      0.87      2.24      2.35     1.05
       18      628        387      0.62      2.24      2.67     1.19
    

    (The missing laser IDs did not have label values for output power and/or REF Frequency.)

    Notes on the Waveplates in HP/Agilent HeNe Laser Tube Assemblies

    When two-frequency Zeeman lasers are discussed in textbooks and research papers, it is generally stated that the magnetic field splits the original mode into two modes that are left and right circularly polarized. Period. However, if this were true, a 1/4 waveplate alone would be sufficient to both convert them to orthogonal linear polarized modes and to align them with the X and Y axes of the laser as required for use in an interferometer. The reason is that a 1/4 waveplate takes a left-hand circular polarized input and produces a linear polarized output at +45° with respect to its fast axis, and takes a right-hand circular polarized input and produces a linear polarized output at -45° with respect to its fast axis. The input orientation doesn't matter. So, the 1/4 waveplate could be oriented to align the output linear polarizations without regard to the input.

    But HP lasers from day 1 (the original 5500A around 1969 even before it had an official model number) and Agilent lasers to the present have all had both 1/4 and 1/2 waveplates with the basic design unchanged over more than 40 years. Further, both waveplates are in mounts that allow the tilt of each one to be adjusted around one of its principle axes. Why? Some possibilities as to the reasons for this more complex setup are as follows:

    Real HeNe laser tubes exhibit some small random amount of birefringence both from the fine structure of the mirror coatings as well as from unavoidable geometric asymmetry in their construction. Without a magnetic field or explicit polarization control measures such as a Brewster window or plate, these tend to lock the polarization of the longitudinal modes to a fixed orientation about the tubes optical axis, and 90 degrees from it. Adjacent modes will almost always be orthogonally polarized. In a HeNe laser with an axial magnetic field such as one from HP/Agilent, this will result in the Zeeman modes being slightly (or not so slightly) elliptically polarized rather than pure circularly polarized. So, the orientation of the 1/4 waveplate will matter and only certain orientations (2 or 4) will convert these to orthogonal linearly polarized modes. But the resulting linealry polarized modes in general won't be aligned with the system's X and Y axes, so the 1/2 waveplate is then required to rotate them to match. (The magnetic field will also never be perfectly symmetric or uniform, though I don't know whether this is ever a significant factor in affecting the mode polarization.)

    The adjustable tilt allows the exact retardation of each waveplate to be altered slightly. I find it somewhat hard to believe that the reason is simply to be able to use cheapo waveplates that might not always be exactly 1/4 or 1/2 wave! However, this explanation can't be entirely discounted since the accuracy of the retardaion is critical to producing F1 and F2 modes that are purely linear and precisely orthogonally as required for the metrology applications. And, the waveplates are made from what looks like optical-grade mica whose discrete layers preclude the ability to select the exact retardation by controlling thickness. And whether mica waveplates were originally selected based on low cost or zero order or temperature stability or being very thin to avoid significantly shifting the beam when tilted or being what the designers had laying around is not known either. But there might be another reason for this "feature" - namely to further compensate for some deficiency in the modes coming out of the tube, again related to deviation from being purely circular. Or something. ;)

    Thus, there are in fact 4 degrees of freedom, though clearly the tilt has a much smaller effect than the rotation. And without a full understanding (possibly including hairy math!), it's difficult to really come up with an adjustment procedure that will work in general. That's the bad news.

    The good news is that from experience, swapping the entire waveplate assembly between HP/Agilent tubes is likely to result in acceptable performance without any adjustments other than making sure that the overall orientation is the same. Therefore, it would appear that these errors are usually small. But it does seem that it is sometimes necessary to touch up the tilt of the 1/4 waveplate in order to produce the best mode purity in addition to optimizing the waveplate orientations. Adjusting a waveplate assembly from scratch (random orientation and tilt) may be much more challenging though since it would first be necessary to match the orientations of each of the waveplates to a waveplate assembly to at least allow the laser to lock. Without a stable output, going further would be virtually impossible.

    So much for hand waving. :-) One might think that a good source for information relating to Hewlett Packard's metrology laser technology would be patents. But so far, patent searches have turned up almost nothing of relevance. If anyone has knowledge or references related to the waveplate issue or anything else of relevance, please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    The only actual waveplate fine tuning procedure to maximize the F1/F2 separation and achieve optimal orientation I've ever seen for one of the early HP-5500 lasers with built-in interferometer optics. The goal then is to minimize any return of the component that is supposed to be transmitted. It simply says something like:

    But there is no explanation as to why this procedure is the way it is. And they don't mention the tilt adjustment on the 1/2 waveplate for some reason.

    I did tests of WavePlate (WP) sets from three HP-5517 lasers using a linearly polarized HeNe laser. The results are as follows:

                         <--------- Orientation ---------->
        ID#    Laser     Input    1/4 WP    1/2 WP   Output
     ---------------------------------------------------------
         1   5517C 9mm    +20°     +20°     +32.5°    +45°
         2   5517B 6mm    -20°     -20°     +12.5°    +45°
         3   5517B 6mm    +20°     +20°     +32.5°    +45°
    

    (I doubt that the specific type of 5517 laser or its beam diameter makes any difference. The accuracy of my measurements on orientation is within +/-2° for input and +/-1° for output, though the latter at +45° is probably quite precise based on theory.)

    The Input is the orientation required for the polarized HeNe laser to produce a pure linearly polarized beam at the output of the WPs, and thus also of the orientation of the optical axes of the 1/4 WP. (Only if aligned with the slow or fast axis of a 1/4 WP will the polarization remain linear, a requirement for these tests.) As expected, the output orientation is the same in all cases since the desired output will be rotated by 45° to align with the X and Y axes. This is a result of the conversion from circular to linear polarization by the 1/4 WP, at 45° with respect to its optical axes. The orientation of the 1/2 WP was inferred from the transfer function from input to output. Now, it's quite possible that the orientation of the 1/4 WP was chosen at random and not actually determined for the specific tube unless it was found that the adjusting the 1/2 WP alone would not meet specifications. However, I have found that it may be necessary to iteratively tweak the 1/4 WP and 1/2 WP to achieve best purety of the F1/F2 modes - the same result could not be obtained by adjusting only the 1/2 WP. Indeed, with a genuine original HP/Agilent tube that has its WPs optimally adjusted, the purety of the F1/F2 modes is nearly perfect. So, perhaps they start at 0° (or +/-20°!) for the 1/4 WP orientation and go from there.

    With the Zeeman tube producing a pure two-frequency left and right circularly polarized beam, all of these WP assemblies would result in pure orthogonal linearly polarized outputs oriented along the X and Y axes. This was confirmed by placing another 1/4 waveplate in the linearly polarized HeNe laser's beam at 45 degrees to produce pure left and right circularly polarized inputs to the WPs. The results were linearly polarized outputs oriented along X and Y. In neither WP assembly, was there any indication that the tilt of either WP was adjusted for other than pure 1/4 or 1/2 wave as the extinction when set for optimal linearly polarized outputs was nearly perfect.

    Experiments with common barcode scanner HeNe laser tubes inside strong magnetic fields have also shown similar characteristics. With these as well, the beat frequency would come and go as the tube heated and expanded with this effect becoming more pronounced when the magnetic field encompassed the entire tube as it does with the HP lasers.

    I also did some tests with and without the presence of the waveplates from an HP laser. Without waveplates, there was absolutely no indication of a polarization preference in the output beam at any time indicating pure circular polarization. When the optics were installed and aligned to the original blue paint, the symmetry of the beat waveform, if nothing else, was polarization dependent. In addition, just after the output beat appeared as well as just before it disappeared, the polarizer would suppress the beat entirely when oriented so that its axis was parallel or at 90 degrees to the axis defined by the blue paint. These perhaps weren't quite as dramatic as the effects I was hoping for but confirmed some of the speculation at least.

    HP/Agilent Laser Wavelength/Optical Frequency

    While most types of lasers, an in particular, HeNe lasers only operate on very precise wavelengths (or equivalently, optical frequencies), the HP/Agilent specifications include the nominal values to 8 or 9 significant digits. The chart below lists the spec'd nominal wavelength in vacuum, in air (at STP assuming n = 1.00027593), and optical frequency, from HP/Agilent specifications:

      Laser  |<-------- Wavelength -------->|   Optical
      Type       Vacuum            Air          Frequency
     -------------------------------------------------------
      5501A   632.99???? nm   632.81???? nm   473.6122?? THz
      5501B   632.991372 nm   632.816759 nm   473.612234 THz
      5517A   632.991372 nm   632.816759 nm   473.612234 THz
      5517B   632.991372 nm   632.816759 nm   473.612234 THz
      5517BL  632.991372 nm   632.816759 nm   473.612234 THz
      5517C   632.991354 nm   632.816741 nm   473.612248 THz
      5517D   632.991354 nm   632.816741 nm   473.612248 THz
      5517DL  632.991354 nm   632.816741 nm   473.612248 THz
      5517E   632.99???? nm   632.81???? nm   473.6122?? THz
      5517FL  632.991354 nm   632.816741 nm   473.612248 THz
      5517G   632.99???? nm   632.81???? nm   473.6122?? THz
      5518A   632.991372 nm   632.816759 nm   473.612234 THz
      5519A   632.991354 nm   632.816741 nm   473.612248 THz
      5519B   632.991354 nm   632.816741 nm   473.612248 THz
    

    I'm not sure what accounts for the two different wavelengths (and thus optical frequencies) among these lasers. There are no obvious physical differences to account for it. The tubes, beam samplers, and relevant portions of the control electronics are all identical. So, it's possible there was a change in isotopic gas-fill or pressure or something else between 5517A/5517B/5518A/5501B lasers and those that came after them. The difference of approximately 12 MHz is still way lower than the commercial-grade error spec of +/-0.1 ppm (roughly +/-47 MHz), so it really doesn't matter. For the Military-grade lasers, the exact optical frequency is measured and included in the calibration report. But a report for one laser I saw had the optical frequency over 10 MHz away from the spec'd value anyhow. My contact at NIST doesn't even know whether it's an actual change in optical frequency or simply an upgrade to the calibration in the measurement electronics!

    I've compared the optical frequency of multiple 5517 and 5501B lasers and have found no evidence of any real difference, let alone one averaging 12 MHz as shown in the specifications above. Here are some data. These lasers are listed in more or less the order in which they were tested:

                  Locked   REF/     Balanced
     Laser Laser  Output   Split    Frequency
      ID   Type   Power    Freq.   Difference  Notes/Comments
    -------------------------------------------------------------------------------
      1    5517B  660 µW  2.3 MHz   -2.30 MHz  Faulty beam sampler was replaced
      2-0  5517B  480 µW  2.4 MHz   -1.44 MHz  Laser 2 with its (new) beam sampler
      2-1   " "    "  "    "   "    -1.35 MHz  Laser 2 with beam sampler 1
      2-2   " "    "  "    "   "    -6.75 MHz  Laser 2 with beam sampler 2
      3    5501B  450 µW  1.9 MHz    0.00 Mhz  Other lasers referenced to Laser 3
      4    5517E  120 µW  6.3 MHz   -2.10 MHz  Only laser with digital Control PCB 
      5    5517D  120 µW  3.6 MHz   -8.25 MHz
      6    5517C  260 µW  2.7 MHz  -15.60 MHz  Tube run at 4.0 mA (not 3.5 mA)
      7    5517C  240 µW  2.9 MHz   -9.00 MHz       "                   "
      8    5517C  210 µW  2.7 MHz   -8.58 MHz       "                   "
      9-0  5517D   80 µW  3.7 MHz   -9.10 MHz  Laser 9 with its beam sampler
      9-1   " "    "  "    "   "    -7.10 MHz  Laser 9 with beam sampler 1
      9-2   " "    "  "    "   "   -11.10 MHz  Laser 9 with beam sampler 2
     10-0  5517A  550 µW  1.7 MHz   -7.63 MHz  Laser 10 with its beam sampler
     10-1   " "    "  "    "   "    -2.13 MHz  Laser 10 with beam sampler 1
     10-2   " "    "  "    "   "    -9.33 MHz  Laser 10 with beam sampler 2
     10-3   " "    "  "    "   "    -3.44 MHz  Laser 10 with beam sampler 3
     11    5501B  220 µW  2.1 MHz  -10.70 Mhz
     12    5501B  150 µW  1.8 MHz   -8.65 Mhz  Tube run at 4.0 mA (not 3.5 mA)
     13    5501A  100 µW  2.0 MHz  -23.48 MHz  Really high mileage!
     14    5501A   50 µW  2.1 MHz  -25.42 MHz     "           "
     15    5501A   35 µW  2.0 MHz  -34.47 MHz     "           "
     16    5517A  410 µW  1.6 MHz   -0.34 MHz  Tube run at 4.0 mA (not 3.5 mA)
     17    5517D  405 µW  3.7 MHz  +75    MHz  Rebuilt with non-HP/Agilent tube
    

    Diagram of Test Setup for HP/Agilent Laser Optical Frequency Comparison shows the way these measurements were made. Photo of Test Setup for HP/Agilent Laser Optical Frequency Comparison shows how ugly it really is, but the scope on the right is displaying the actual beat signal of a pair of 5517Bs. For later measurements, two external HeNe laser power supplies (the white boxes and the Variac) were added to eliminate the FM introduced by the switching noise/ripple of the internal HeNe laser power supplies, but they weren't used for these tests and wouldn't have affected the results since the high frequency FM would be averaged out.

    The 5517 lasers were enclosed in standard cases (non-vented for all except the 5517E) and allowed to reach equilibrium (2 hours minimum). (Removing the cover may significantly change the optical frequency once equilibrium is reached.) The lasers in the photo don't have any cloths on, but, well, that's another matter! :) The 5501As were run naked so that the "Photodiode Offset" adjustment could be performed. (More below.) It might be best to do this via a hole in the cover as they do drift significantly with the cover in place. But I wasn't *that* enthusiastic!

    Laser 3, the first and healthiest 5501B to be tested, was chosen arbitrarily to be the reference for optical frequency. The Balanced Frequency Difference is the frequency of the mid-point between F1 and F2 for the subject (ID) laser minus the frequency of the mid-point between F1 and F2 for Laser 3. There's still a +/-1 MHz or more uncertainty due to variations in the specific lock point of the two lasers being compared during any given run.

    The three 5501As are very well used and weak, but it was easy to obtain a beat between them and the 5517A, Laser 10 (which happened to be the 5517 laser tested and thus conveniently left in place!). 5501As have a "Photodiode Offset" adjustment, which moves the lasing line on the split gain curve. It's the square pot (R4) on the Lock Reference PCB, clockwise rotation decreases optical frequency. I could have set them all to have a 0 MHz difference frequency, but this would have resulted in grossly unbalanced mode amplitudes for these high mileage lasers. So, they were adjusted according to the HP procedure - maximizing the F1-F2 REF frequency, which centers the lasing line between the split neon gain curve. Before doing this, it wasn't even possible to see the difference frequency with Laser 13 likely because it was too high for my instrumentation. This was probably because parts of Laser 13 had been swapped, including the tube, without making any adjustments. The Photodiode Offset adjustments on the other 2 5501As were quite close to optimal. However, this is a single turn pot which adjusts the mode ratio from 1:2 to 3:2, and thus the optical frequency varies significantly with very small changes in its position - possibly 50 MHz or more end-to-end. Going only by the REF frequency - which isn't perfectly stable - it's quite likely that there will be an uncertainty of 5 or 10 MHz. So, best would be to adjust this pot (make it a 10 turn pot!) through a hole in the cover after the laser has reached thermal equilibrium. And if what you want is a precise optical frequency and don't mind some possible mode imbalance, adjust it with respect to a reference laser like an iodine stabilized HeNe laser instead of for maximum REF frequency! :) And, although the optical frequency changes with the cover installed, the Photodiode Offset adjustment could still be optimal if the change is due to the tube temperature, and thus the gas pressure increasing. That would still maintain the same mode balance.

    Laser #11 may have been rebuilt by a company other than HP/Agilent. (I'm not at liberty to reveal the company name.) At the very least, its tube had been removed for some reason and then reinstalled using red RTV Silicone instead of the normal black rubbery potting compound. Exactly what was done is not known. REF at 2.1 MHz is slightly out of spec for the 5501B (1.5 to 2.0 MHz). This laser was run continuously for approximately 14 months at which point its output power had declined to about 100 µW after full warmup. While it still locked reliably, amplitude ripple was causing errors in the interferometer. (Adding anode ballast resistance and/or running the tube at a slightly higher current would probably cure this.) The REF frequency was unchanged. I do not know of what significance this is, if any. :)

    Laser #17 was rebuilt by another company other than HP/Agilent. (I'm also not at liberty to reveal the company name.) It had its original laser tube removed from the magnet/optics assembly and replaced with a laser tube that is not from Agilent. This explains the large offset in optical frequency, which could result from any number of factors. (See below.) The offset is probably of little practical consequence as long as it remains relatively constant.

    Aside from laser #17, based on these data, there is really no consistent difference in average optical frequency based on laser type and if anything, it goes the wrong way! And note the change resulting from the swap of the beam sampler. Beam Sampler 1 was originally on Laser 2 and was resulting in the optical frequency dancing around, then swapped with Beam Sampler 2 which resulted in a large frequency offset, then with a third Beam Sampler which was finally well behaved and now remains in that laser. I have no reason to suspect anything is wrong with either Beam Sampler 1 or 2 and did test them for basic functionality with a voltage source and polarizer. All beam sampler assemblies I've checked regardless of what laser they came from have exactly the same part number though it's possible that the optics inside differ in some subtle way depending on laser type. There are at least two versions of the housing - one with a small aperture for 6 mm optics and another with a large aperture for 9 mm optics, but beyond that I don't know of any differences. Lasers 3 through 7 definitely have their original beam samplers. Though I don't have minimum specs for the 5517E, Laser 4 is probably relatively high mileage. And lasers 5 and 9 are high mileage as evidenced by their low (below spec) output power. Even though the output power of Lasers 6, 7, and 8 is well within spec, they are also definitely high mileage lasers being extremely slow start and unable to run on the normal 3.5 mA discharge current. This in itself shouldn't have a large effect on optical frequency unless the tube actually runs hotter (in which case the optical frequency should increase, more below). But the lock point temperature adjustment has not been changed on these lasers, so the equilibrium bore temperature should be similar to that of the others though the equilibrium laser tube envelope and laser temperature will be slightly higher.

    So, the actual optical frequency may be dominated by the amount of use (number of hours on the tube) which also tends to correlate with a decline in output power. This may overwhelm any real or fictitious optical frequency offset found in the specifications. While I don't know what the original output power was for most of these lasers, those with 400 µW or more start very quickly or instantly and are likely relatively young (usage-wise). Laser 1 is known to have been taken out of service due to a bad LCD in the beam sampler, so it could have seen relatively little use.

    There has been research showing that the neon gain center frequency tends to decline with use due to a drop in tube pressure and other factors. Helium has an effect on lasing center frequency of about +22 MHz/Torr, so a loss of He due to gas entrapment on the tube walls or cathode, leading to a drop in its partial pressure, can easily account for these large frequency differences. (Loss of He due to diffusion through the tube walls would also result in a decline in its partial pressure, but this loss mechanism should be minimal.) Major factors include:

          Cause             Sensitivity       Comments    
     ------------------------------------------------------------------------------
      Helium Pressure      +22 MHz/Torr       Pressure of He decreases with use
      Neon Pressure        -25 MHz/Torr       Pressure of Ne decreases with use
      Neon Isotopic Ratio  +10 MHz/% of 22Ne   Ratio of 22Ne:20Ne Decreases with use
      Temperature          +280 kHz/°C        Affected by specific lock point
    

    Both He and Ne partial pressures descrease over the life of the tube but because the fill ratio is between 5:1 to 9:1 of He:Ne, the decrease in He pressure dominates and a frequency drift downward of several MHz/year is quite reasonable. If not filled with a pure Ne isotope, the Ne isotope ratio also will change slightly as the 22Ne will be trapped at a slightly a higher rate than 20Ne. Note the strong dependence on the Ne isotope ratio, a 1 GHz range! So, just over a 1 percent change in the ratio at the time of manufacture could account for the 12 MHz difference in nominal frequency specifications. And, for any given measurement, there is uncertainty in the actual lock point as the laser warms up but that's probably only a maximum of +/-1 MHz or so. The 280 kHz/°C is for a tube about 8-1/2 inches long - similar in length to most of the HP/Agilent tubes. However, note that for the HP/Agilent lasers, the temperature of the tube envelope is not controlled, only the mirror spacing rod for the 5501B and 5517s, and not at all for the 5501A. So, the temperature of the interior of the laser may have a significant impact on the optical frequency. This differs from many other stabilized HeNe lasers where a large portion of the tube is wrapped in a heater.

    The above has been distilled from the paper: "Frequency stability measurements on polarization-stabilized He-Ne lasers", T. M. Niebauer, James E. Faller, H. M. Godwin, John L. Hall, and R. L. Barger, Applied Optics, vol. 27, no. 7, 1 April 1988, pp. 1285-1289. However, a later paper states the contribution from the Ne isotope ratio as being 8.75 MHz rather than 10 MHz per percent of 22Ne.

    So, it's quite possible that any differences in the optical frequency of these lasers when they were new is totally swamped by changes due to use. For example, if a laser has been run 24/7 for 3 years (middle age for these lasers!), its optical frequency could have gone down by 10 to 15 MHz due to the decline in gas pressure and isotope ratio changes. But the ultimate conclusions may be that (1) it's not worthwhile to assume anything about the nominal optical frequency on used HP/Agilent lasers, but if the optical frequency can be measured (or compared to that of a new laser), (2) the frequency shift may be a means of estimating how many hours or years they've been on! :-)

    Despite all these potential source of variability, for an application requiring an accurate stable optical frequency reference like calibration of a wavementer, a healthy 5517 laser (any version) is probably a better choice than a laboratory stabilized HeNe laser like a Spectra-Physics 117A. The reason is that the design of the 5517 inherently locks to a balanced mode state, with no adjustments and little in the electronics to drift with age to change this. Lasers like the SP-117A have separate photodiodes and pre-amps for the two mode signals as well internal adjustments that can affect the lock point. Furthermore, the optical frequency specifications of all HP/Agilent lasers are known (even if there is an unexplained discrepancy of 12 MHz going from the 5517B to 5517C). This is not the case for many other stabilized HeNe lasers. And, if needed, the laser can easily be packed up and sent to NIST or elsewhere to have its optical frequency measured precisely without fear of it changing either from a few bumps during shipment, or over time if turned on periodically rather being run 24/7. I wouldn't recommend other HP lasers like the 5501B simply because healthy ones are becoming harder and harder to find. And the 5501A uses a different locking design which is similar to that of the other (non-HP/Agilent) lasers. However, a healthy 5518A or 5519A/B would also be suitable, using the same design as the 5517.

    Also see the section: Comparing the Optical Frequencies.

    Obtaining an Interferometry Laser for Experimentation

    While most people on this planet at least would consider an HP/Agilent laser most useful as a door-stop, there are reasons to want one other than for use in your private semiconductor foundry or custom optics diamond turning machine. :) In addition to their intended metrology applications, these lasers are also ideal as laboratory stabilized HeNe lasers and optical reference lasers. They could certainly substitute for lasers like the Coherent 200 and Spectra-Physics 117/A if one of the frequency components is blocked with a polarizer and the generally lower output power is adequate. Since the optical wavelength/frequency of the HP/Agilent 5517 laser (all versions) is known and very stable over time, they can be used for wavemeter, spectrometer, and optical spectrum analyzer calibration. If ultimate accuracy is needed, the laser can be heterodyned with an iodine-stabilized HeNe laser. Since the HP/Agilent lasers do not drift much even over years if not run 24/7, such a laser will stand in for the even more expensive iodine-stabilized laser without being tied to one. The split frequency is also handy for testing ultra-high finesse interferometers (another thing everyone needs!) And, it would also be possible to use one of these lasers as the basis for a seismometer or the ultimate laser microphone. :)

    So, if you're salivating for an HP/Agilent laser and can't live without one, they cost somewhere between $8,000 and $12,000 new depending on options! Used (or "previously owned" - which would be classier!) HP/Agilent lasers can be had much cheaper but caveat emptor. The only ones most people can afford for personal use would be found on eBay. But most of these lasers that end up being resold are taken out of service because they have an end-of-life tube. Interferometry lasers used for metrology are often run 24/7 from the day they are installed until they die. Even though the lifetime of the special HeNe laser tubes used in these lasers may be 50,000 hours, that's still only about 6-1/4 years. And guess where they then end up? :) If the seller hasn't powered the laser head (or doesn't admit to it) and lists the laser "as-is" with no returns, chances are excellent that it will serve as a nice doorstop but not much else, at least not without some effort. Unfortunately, except for the 5519A which plugs into a standard wall socket, these lasers require +/-15 VDC for power with a Military-style connector that you won't find at Radio Shack. It's easy to "hot wire" power from inside, but unless the seller is familiar with this sort of thing or has the mating power supply and cable, it may be better to just get a DOA warranty in writing and accept that you may have to pay shipping both ways if the laser is only good as a doorstop.

    These lasers also show up at surplus dealers but they tend to ask higher prices than would be considered acceptable for basic tinkering and many seem content to simply have the laser gathering dust than to let it go for a realistic price if untested. But I've also heard of at least one instance where such a laser was found at a garage sale. That price was almost certainly right!

    Also, note that when looking for lasers like this on eBay or elsewhere, the clothes these lasers wear are of little importance. Newer Agilent OEM 5517 lasers (which are mostly what show up as late model surplus in 2010) tend to have a thin cheaply made gold-ish (alodined) aluminum shroud rather than the beige or gray two piece case of most older HP lasers. It has a feeble attempt at a rubber gasket all around to seal it but this really doesn't work well and only makes reassembly a royal pain. What's inside is the same, though lasers built after the early 2000s will likely have the newer Digital Control PCB rather than the analog Control, PCB but they are otherwise identical and functionally equivalent. And in the trivial triviality department, Agilent's only concession to style seems to be in the color of the front and back plates: Beige for 5517Bs, silver for 5517Cs, and gold for 5517Ds! :) But this is only true of some samples and there doesn't appear to be any way to predict which ones.

    Regardless of who is selling the laser, if they are able to power it, the three most important things to ask of them would be:

    1. Is there a good stable output beam that doesn't sputter, flicker, or cut out and restart at any time over the next few minutes? And if so, how long does it take to appear?

      "Yes" and "a few seconds or less" means the HeNe laser tube and power supply are probably good and happy working together. A laser that takes awhile to start may still be fully functional, but it can be annoying to wait 10 minutes for a beam, and associated equipment may expect the laser to be ready within a fixed amount of time.

      Note that the 5501B is the one exception where the laser tube isn't turned on until near the end of the locking process. Thus a beam that doesn't appear immediately on the 5501B is a feature, not a bug. :) This was probably done in the design to minimize the DC current consumption during warmup to be backward compatible with the 5501A. My preference, where this isn't an issue, is to bypass the transistor switch and enable the tube immediately as with all the other lasers.

    2. Does the READY indicator come on and stay on (solid, not flashing) in about 10 seconds for the 5501A; around 4 minutes for most of the common 5517s, 5518A, and 5519A/B; and 6 to 9 minutes for the 5501B?

      A "yes" answer to this question alone is usually sufficient to confirm proper operation with usable power for many purposes. However, cold start to READY on solid may be over 10 minutes - even up to 20 minutes for a few lasers like the 5517E, 5517FL, or 5517G, and some other 5517s using the Newest Digital Control PCB, but these lasers are almost non-existent surplus. Nearly all those found on eBay even in 2010 should come ready in the typical 4 minutes. A slightly longer time like 5 minutes is of no consequence, but if a pre-2000 laser takes several minutes longer, it's probably very low power and requires the extra time for the power to increase enough as the tube warms up for the laser to be convinced there is enough power available. However, such a laser could still be useful.

      However, some types of data processing systems like the HP-5508A Measurement Display will produce a hard error if the laser takes more than 10 minutes to become ready. (Power cycling the 5508A once the laser is ready will generally get around this even if it powers the laser, as it's likely to become ready much quicker once warmed up.)

    3. What is the output power once the READY indicator comes on solid? The HP/Agilent specification for minimum power is 180 µW for most of these lasers except the 5501A (and really old 5500A/B/C) which are 120 µW. Note that the locked output power is generally at or near the minimum of the power as it varies due to mode sweep during warmup. So, simply measuring output power before then may result in an overly optimistic reading and may even be more than double the locked output power for a high mileage laser.

      For many applications, much less than spec'd minimum power is quite sufficient. Even if the seller is unable to measure the output power, as long as READY comes on solid, it is probably at least 80 µW for all the lasers except the 5501A, which will lock at much lower power - down to 40 µW or less. Even this is sufficient for a single axis system.

    Where the laser passes these tests, it will probably be more than adequate for an experimental, demo, test, educational, or research system which doesn't have many measurement axes and isn't run continuously for years. However, before considering such a laser for installation in a semiconductor wafer stepper producing next generation multi-core processors, many additional tests would need to be performed to determine its present health and life expectancy. In some installations, the laser is swapped out after a fixed number of hours, like fluorescent lamps! :) While in others, they are replaced at the point where they are at the hairy edge of meeting HP/Agilent specs. Either of these approaches makes sense where where the cost of down time is extremely high. So, even though they may start instantly, run reliably, and have decent output power much greater than the HP/Agilent minimum, if their REF frequency is found to be at or above the range for that model laser, they may be flagged for replacement during preventive maintenance. (REF frequency tends to increase with use and is related to the decline in output power.) An example would be a 5517B outputting 500 µW with a REF frequency of 2.6 MHz. The spec'd REF frequency range of a 5517B is 1.9 to 2.4 MHz. Details are beyond the scope of this presentation, but there may be a writeup in the future. Stay tuned.

    (If anyone has an HP/Agilent laser and has a record of the output power and REF frequency when new either from measurements, the label, or original paperwork, and what they are now, and if possible, an estimate of how much it has been run, please contact me via the Sci.Electronics.Repair FAQ Email Links Page. This will aid in my attempt to more accurately estimate previous use and life expectancy for these lasers.)

    Even if the laser plays dead, it could just be a blown fuse or something else that's easily repaired. Or, it could be *really* slow start.

    For details, see the section: Common Problems with HP/Agilent 5517 Lasers.

    And even if the HP/Agilent laser tube is certifiably dead, it is possible to install an inexpensive barcode scanner tube in its place that results in a usable system, at least for experimentation or demos. This isn't for the casual user, but if you're up to a modest challenge and have some basic mechanical and electronic skills, see the section: Installing a Common HeNe Laser Tube in an HP-5517 or 5501B.

    For more information on alternatives to purchasing new HP/Agilent lasers and critical issues in their selection and testing, see the companion document: Considerations in Evaluating Used or Rebuilt Hewlett Packard/Agilent Metrology Lasers.

    And, if you do come across one of these lasers at a garage sale, just splurge, pay the $2 they're asking, and take the risk. :-)

    Maximizing the Life of HP/Agilent Metrology Lasers

    Although the spec'd lifetime of the special tubes used in these lasers is around 50,000 hours or slightly over 6 years, with a sticker price of something like $10,000, it still pays to do everything reasonable to maximize their usable life. (And, there is some evidence that newer lasers typically run only 3 to 4 years before going outside Agilent specifications.) Many if not most semiconductor fabs and other similar production facilities keep these lasers powered continuously regardless of whether they are actually being used. This may be based on the belief that (1) they take a long time to reach the required wavelength accuracy, (2) stuff tends to fail when powered cycled, and (3) HeNe lasers like to be run. None of these is particularly valid for these lasers. But even so, keeping them on 24/7 still may be most cost effective where minimizing any unscheduled downtime is more critical than replacement costs. However, for educational institutions and general research and development applications, there is no reason to do this. Power up the laser an hour or so before the system is needed and power it off when not needed for several hours or more. Here some comments on each issue:

    So, in short, the laser itself won't function any better if run continuously compared to being turned on at most 90 minutes before needed as long as it doesn't affect the environment in such a way as to change the calibration. (90 minutes is HP/Agilent's spec for warmup to full accuracy on an unvented laser, only 45 minutes on one with forced air cooling). And for less critical applications, simply waiting until READY comes on solid may be adequate. It should be possible to test for the overall effect by making a measurements of a known length in each axis when the laser comes READY, after 90 (or 45) minutes, and after 24 hours. If any differences found are acceptable, there is nothing to be gained by continuous operation.

    Where the laser might be used for a few hours a week, as in a diamond turning machine at a custom optics house, this should effectively extend the life of the laser to infinity.

    Installing Newer HP/Agilent Metrology Lasers in Older Systems

    As time passes, locating functional 5501A and to a somewhat lesser extent, 5501B lasers to repair an older system is becoming increasing difficult. While it's sometimes possible to find a 5501A laser from 1981 that has a like-new tube, the vast majority of them are good for little more than doorstops or high tech sculpture.

    (The following deals with retrofitting systems using 5501A or 5501B lasers. For really old systems using 5500A/B/C lasers, a few more issues are present since the 5505A Measurement Display is more tightly coupled to the laser and somewhat more is involved to keep it happy.)

    It is often possible to install a more modern 5517 laser in place of a 5501A or 5501B. 5517 lasers are still in production and used working units are also readily available at very reasonable cost.

    Only a few relatively minor differences need to be accommodated. With only a bit of resourcefulness, the total cost (excluding the laser and labor) for this type of conversion will be under $100 in most cases:

    Repair of HP/Agilent Metrology Lasers

    The custom tubes at the heart of these lasers are what will dominate the cost of repair in most cases. Electronics problems are relatively uncommon, and generally minor when they do occur. Where the tube is found to be unusable, a replacement tube will be needed. New tubes are only available from Agilent and I wouldn't even want to think about their cost but figure 75 percent or more of the cost of a new laser. And they probably wouldn't even sell a tube without insisting on doing the installation as well (at an additional charge). One company that refurbishes tubes for precision HeNe lasers is Evergreen Laser Corporation. But I do not know if they will deal with HP/Agilent tubes, and rather doubt it due to the additional issues of gaining access to the glass tube itself buried inside the magnet assembly. And even if they will, the cost of a refurbished tube would probably exceed the cost of a complete guaranteed full spec laser on eBay! Another company I contacted who does repair these lasers said candidly that their success in refurbing tubes was spotty at best - and I think that was being overly generous. One company that does seem to claim a complete overhaul capability is Hong Ren Technology Co., Ltd. but I have no idea of their reliability. But keep in mind that for any of these, you're talking thousands of dollars, not garage sale or even eBay prices! From photos on their Web site showing a non-standard tube mount, it appears as though they may actually build their own tubes or provide tubes that are not from Agilent. Non-genuine tubes could have subtle differences in beam profile, wavelength, stability, and lifetime.

    However, even a tube deemed to be dead by Agilent due to low power or an inability to stay lit, may often be made usable for many applications (especially where only 1 or 2 measurement axes are required), for a test or educational system, or as an emergency spare, with at most some relatively minor low cost modifications to the laser, or possibly even simply an adjustment. But if the output power is so low that the beam actually disappears periodically while warming up, there won't even be a beat signal and such a tube is only good as a high tech paperweight with built-in magnetic paper clip holder. :)

    Assuming the tube is usable, except for late model Agilent 5517 lasers based on digital Control PCBs, all of these lasers are very serviceable as far as the electronics are concerned. Pre-2000 lasers - most of what's found surplus even in 2010 - will certainly have the older Analog Control PCBs. For these, even most of the HP house-numbered ICs have standard equivalents available from major electronics distributors, and none of the other electronic parts are special. Operation and service manuals are available which include detailed adjustment and troubleshooting information and complete schematics. And parts units can be obtained on eBay at low cost. Except for a blown fuse of my own doing, dried up electrolytic capacitors on really old lasers, a blown line driver chip, and bad REF photodiode, I've yet to see an Analog Control PCB with any serious problems including defective proprietary HP ICs. However, on a 5501B, the heater driver transistors and main fuses were blown as a result of dried up electronic capacitors on the Connector PCB. So, for 5501B lasers, it's probably good preventive maintenence to replace all 4 large electrolytic capacitors on the connector PCB on a laser more than 10 or 15 years old as a precaution. This is the only situation I know of where a high ESR/low uF capacitor will result in actual damge to other components. For more on the 5517 laser in general and the digital Control PCBs in particular, see the section: HP/Agilent 5517 Laser Construction.

    Now, if you're independently wealthy and would like to have Agilent repair your laser, I've heard that an evaluation is about $500. Essentially, they confirm that it's an HP or Agilent laser and then tell you how much it will cost to repair, if they are willing to repair it at all. For a single failure, the cost is a flat rate between $1,500 and $2,000, but the evaluation fee will be applied toward that, thank goodness. :) A "single failure" probably includes a blown fuse, broken resistor, dried up capacitor, or bad IC. I don't know whether something like a degraded LCD in the beam sampler or blown HeNe laser power supply would qualify as a single failure, or if two dried up capacitors would be charged (no pun....) for separately. And, it's almost certain that if you read the fine print, the flat rate would exclude a weak or dead HeNe laser tube that required replacing the tube assembly even though it is technically a "single failure". In that case Agilent would simply return the laser after collecting their evaluation fee.

    For amusement, go to Find-A-Part: Agilent's Test and Measurement Parts Catalog and enter a laser model like "5517D". If you're not independently wealthy, you better be sitting down when viewing the prices. For example, (in 2009) the cover is $344, the Control PCB is $1075, the HeNe laser power supply is $496, and a small screw is a bargain at $1.24 each. However, prices for the operation and service manuals are not totally ridiculous - $28.44 for the 5517A and $42.67 for the 5517B/C. But the exact parts available for each model laser seem to be somewhat random and forget about even being able to order a new tube assembly (or parts). They are listed as: "Not orderable, contact Agilent for repair service". Right. :-)

    And before doing something silly, getting inside HP/Agilent lasers is trivial. On the large lasers (5517A, 5518A, 5518A/B) it's just a matter of removing the 4 tiny screws on top and gently levering up the cover using a knife blade. On the small lasers (5501A/B, 5517B/C/D), rotate the front turret so the large hole is at the bottom. That will expose a slotted head screw - a 1/4 turn fastener. Push in and rotate 1/4 turn counter-clockwise and the front plate will pop off. The covers or shroud can then be removed. The only reason I've gone to this level of detail is that I had an academic type ask me if that screw was for tuning the laser frequency! :)

    Also see the section: Common Problems with HP/Agilent 5517 Lasers (which applies to other lasers like the 5501B as well). For operation and service manuals, see the section: Additional HP/Agilent Resources.

    Agilent Laser and Optics User's Manual

    This manual provides information on the lasers and optics (no surprise, huh?) aspects of Agilent (formerly Hewlett Packard) metrology systems based on two-frequency interferomters. These are links to the Laser and Optics User's Manual PDFs found on the Agilent Web site. Searching there for "Laser and Optics User's Manual" will return these PDF files but I am not aware of a linked table of contents as is present below.

    I also have backup copies of the same PDFs at Sam's Bakcup of Agilent Laser and Optics System Design Manual.

    Note that the file for Chapter 7Y does not exist on the Agilent Web site and I haven't been able to find it elsewhere. I've left the link in place should it magically appear.

    Hewlett Packard/Agilent Part Number Reference

    Here is a partial list of HP/Agilent optics, electronics, and accessories including both current and obsolete products listed by part or model number:

      Part#        Description
     ------------------------------------------------------------------------------
      05505-60048  Rack Mount Kit for 5505A
    
      05508-60021  Remote Control Unit (5528A)
    
       5500A       Laser Transducer (w/interferometer, optical receiver, 0.4 m/s)
       5500B       Laser Transducer (w/interferometer, optical receiver, 0.4 m/s)
       5500C       Laser Transducer (w/optical receiver, 0.4 m/s)
    
       5060-0049   Extender Board, 15 pin
       5060-0630   Extender Board, 22 pin
    
       5505A       Measurement Display (5526A)
        K01-5505A  Extender Board (XA-14), 52 pin
    
       5508A       Measurement Display (5528A)
    
       5510A       Automatic Compensator (5525A/5526A)
        H01-5510A  High Accuracy  Automatic Compensator (5525A/5526A)
        K15-5510A  Multiplexer for 5510A
    
       5501A       Laser Transducer (0.4 m/s)
       5501B       Laser Transducer (0.4 m/s)
    
       5507A       Electronics
    
       5517A       Laser Transducer (0.4 m/s)
       5517B/BL    Laser Transducer (0.5 m/s)
       5517C       Laser Transducer (0.7 m/s)
       5517D       Laser Transducer (1.0 m/s)
       5517DL      Laser Transducer (1.1 m/s)
       5517E       Laser Transducer (1.6 m/s)
       5517EL      Laser Transducer (??? m/s)
       5517F       Laser Transducer (1.7 m/s)
       5517FL      Laser Transducer (??? m/s)
       5517G/GL    Laser Transducer (??? m/s)
    
       5518A       Laser Transducer (w/optical receiver, 0.4 m/s), <SN2532A02139)
       5518A       Laser Transducer (w/optical receiver, 0.453 m/s, >=SN2532A02139)
    
       5519A       Laser Transducer (w/optical receiver, 0.7 m/s)
       5519B       Laser Transducer (w/optical receiver, 1.0 m/s) 
    
       5525A       Laser Measurement System
       5526A       Laser Measurement System
       5527A/B     Laser Position Transducer System
       5528A       Laser Measurement System
       5529A       Dynamic Calibrator
       5530A       Dynamic Calibrator
    
       9211-1586   Transit Case for 5500A/B/C
       9211-1587   Transit Case for 5505A
       9211-1738   Transit Case for 5510A
    
      10550A       Reflector
      10550B       Retroreflector
      10551A       Plane Mirror Convertor
      10552A       Resolution Extender
      
      10555A       Remote Interferometer
      10556A       Retroreflector
      10557A       Turning Mirror
      10558A       Beam Bender
      10559A       Reflector Mount
      10560A       Barometer
    
      10562A       Single Beam Interferometer
      10563A       Material Temperature Sensor
       H01-10563A  High Accuracy Material Temperature Sensor
      10564A       Air Temperature Sensor
      10565A       Remote Interferometer
    
      10567A       Dual Beam Splitter
    
      10579A       Straightness Adapter (Resolution Extender And Optics)
      10580A       Laser Tripod (5500C)
      10581A       Plane Mirror Converter (5526A)
    
      10585A       Metrology Program Package (5526A)
      
      10690A       Straightness Interferometer
      10691A       Straightness Interferometer
      10692A       Penta-Prism
    
      10700A       33% Beam splitter
      10700B       4% Beam splitter
      10700C       15% Beam splitter
      10701A       50% Beam splitter
      10702A       Linear Interferometer
      10703A       Linear Retroreflector
      10704A       Single Beam Retroreflector
      10705A       Single Beam Interferometer
      10705A-080   Fiber Optic Receiver Adapter
      10706A       Plane Mirror Interferometer
      10706A-080   Fiber Optic Receiver Adapter
      10706B       High Stability Plane Mirror Interferometer
      10707A       Beam Bender
      10708A       Power Supply (May Not Apply)
    
      10710A/B     Adjustable Base (Small, Beam Bender, etc.)
      10711A       Adjustable Base (Large, Linear Interferometer, etc.)
      10713B/C/D   Cube Corner
      10715A       Differential Interferometer (DI)
      10715A-001   DI (turned configuration)
      10716A       High Resolution Plane Mirror Interferometer (PMI)
      10716A-001   High Resolution PMI (turned configuration)
      10717A       Wavelength Tracker
    
      10719A       One-Axis Differential Interferometer (DI)
      10719A-C02   One-Axis DI (low thermal drift)
    
      10721A       Two Axis Differential Interferometer
      10721A-C02   Two Axis DI (low thermal drift)
      10722A       Plane Mirror Converter (5501A)
      10723A       High Stability Adapter
      10724A       Plane Mirror Reflector
      10725A       50% Beam Splitter
      10725B       4% Beam Splitter
      10725C       15% Beam Splitter
      10726A       Beam Bender
    
      10728A       Plane Mirror
    
      10735A       Three-Axis Interferometer
      10736A       Three-Axis Interferometer
      10736A-001   Three-Axis Interferometer/Beam Bender
      10737L/R     Compact Three-Axis Interferometer
    
      10740A       Coupler (5501A)
      10741A       Laser Transducer Interface (10740A card)
      10742A       Laser Transducer Counter (10740A card)
      10743A       Extender Board (10740A)
      10745A       HP-IB Interface (10740A card)
      10746A       Binary Interface (10740A card)
    
      10751A/B     Air Sensor (5528A)
      10751-60209  Laser Interferometer Cable
    
      10753A       Laser Tripod (5518A)
    
      10755A       Compensation Interface
      10756A       Manual Compensator
      10757A/B/C   Material Temperature Sensor (5528A)
      10757-60306  Laser Interferometer Cable
    
      10759A       Foot Spacing Kit
      10760A       Counter (10740A card)
      10761A       Multiplier (10740A card)
      10762A       Comparator (10740A card)
      10763A       English/Metric Output (10740A card)
      10764A/B     Fast Pulse Converter (10740A card)
      10764-60005  Laser Interferometer Cable Assembly
    
      10766A       Linear Interferometer
      10767A       Linear Retroreflector
      10767B       Lightweight Retroreflector
      10768A       Diagonal Measurement Kit
      10769A       Beam Steering Mirror
      10770A       Angular Interferometer
      10771A       Angular Reflector
      10772A       Turning Mirror
      10773A       Flatness Mirror
      10774A       Short Range Straightness Optics
      10775A       Long Range Straightness Optics
      10776A       Straightness Accessory Kit
      10777A       Optical Square
      10778A/B/C   Laser Power Cable (5501A/B)
      10779A/B/C   Reference Cable (5501A/B)
      10780A/B/C   Optical Receiver (Free Space)
      10780F/U     Optical Receiver (Fiber-Coupled)
      10781A       Pulse Converter
      10782A       Service Kit without Laser Assembly (5501A)
      10782AOP001  Laser Assembly (5501A) only
      10783A       Numeric Display
      10784A       Base
      10785A       Height Adjuster and Post
      10786A       Linear Measurement Transit Case
      10787A       Straightness And Squareness Transit Case
    
      10790A/B/C   Receiver Cable
      10791A/B/C   Laser Head Cable  
      10793A/B/C   Laser Head Cable (5517A to 5507A and 5518A to 5508A)
    
      10880A/B/C   Receiver Cable
      10881A/B/C   Laser Head Cable (5517, Din Connector)
      10881D/E/F   Laser Head Cable (5517, Spade Lugs)
      10882A/B/C   Laser Head Cable (5519A/B To 10887P)
      10883A/B/C   Laser Head Cable
      10884A       Power Supply
      10884B       Power Supply
      10885A       PC Axis Board
      10886A       PC Compensation Board
      10887A/B     PC Calibrator Board (5518A)
      10887P       PC Programmable Calibrator Board (5519A/B)
      10887-60202  Laser Interferometer Cable
      10888A       Remote Control
      10889A       PC Servo Axis Board
    
      19895A       VME Laser Axis Board
      10897B       VME Laser Axis Board
      10898A       VME Dual Laser Axis Board
      55280A       Linear Measurement Kit
      55281A       Angular Optics Kit
      55282A       Flatness Accessory Kit
      55283A       Straight Measurement Kit
    
      C05-59995A   Reference Cable (5501A/B)
      C07-59995A   Power Cable (5501A/B)
      C08-59995A   Diagnostic Cable (5501A)
    
      E1203C       Precision Beam Translator
      E1204C       Precision Horizontal Beam Bender
      E1705A       Fiber Optic Cable
      E1706A       Remote Sensor
      E1207C       Precision Vertical Beam Bender
      E1208C       33% Bare Beam Splitter
      E1208D       40% Bare Beam Splitter
      E1208E       50% Bare Beam Splitter
      E1208F       66% Bare Beam Splitter
      E1208G       60% Bare Beam Splitter
    
      E1250A/B     High Performance Receiver Cable
      E1251A/B     High Performance Laser Head Cable
    
      E1708A       Remote Dynamic Receiver
      E1709A       Remote High Performance Optical Receiver
      E1713A       Scale Servo Axis Board for E1720A.
    
      E1826E/F/G   One-Axis Plane Mirror Interferometer
      E1827A       Two-Axis Vertical Beam Interferometer
    
      E1833C       15% Bare Beam Splitter
      E1833E       33% Bare Beam Splitter
      E1833G       50% Bare Beam Splitter
      E1833J       67% Bare Beam Splitter
      E1833M       100% Bare Beam Splitter (Beam Bender)
    
      E1837A       Two-Axis Vertical Beam Interferometer
    
      ET-319283    Interferometer Adapter Cable
    
      N1250A/B     High Performance Optical Receiver Cable
    
      Z4399A       Three-Axis Interferometer
      Z4420B       Five-Axis Interferometer
    

    Additional HP/Agilent Resources

    Complete operation and service manuals for HP products like the 5501A, 5501B, 5526A, and 5528A that are no longer supported - as well as other related information are available for download. Most of these have been provided by Jack Hudler scanned and enhanced from original manuals. Others are copies from the Agilent Web site. Please contact me via the Sci.Electronics.Repair FAQ Email Links Page for access information.

    Notes on the HP-5500A Two Frequency HeNe Laser

    The 5500A (or a substantially identical laser that had no model designation) is the original Hewlett Packard metrology laser and was unique compared to all that followed in having the interferometer optics built-in. Thus, the only required external optic required for a single-axis system was a retro-reflector (cube-corner). In fact, the 5505A might have been considered as an option to the 5505A Measurement Display as the only actual part number on it is: C01-5505A, though the decorative strips do say 5500A.

    The HeNe laser tube in the 5500A/B is generally similar to the one in the 5500C and 5501A, but isn't quite identical and thus is not interchangeable, at least not without some work. The original patent for the 5500A/B laser tube is: U.S. Patent #3,771,066: Gas Laser. The most notable obvious differences between the 5500A/B tube and the one in 5500C and 5501A are in the PZT connector at the rear which is a ring (rather than a center terminal) that allows the waste beam from the HR mirror to escape, and the optics assembly at the front of the tube assembly which only has the beam expander - the waveplates are mounted externally (though strictly speaking these aren't part of the tube itself). And the glass tube is simply clamped to the mounting feet, which are not part of the tube assembly.

    Rather than using a portion of the main beam for feedback, there's a shielded can with a photodiode behind a 1/4 waveplate and motor driven rotating polarizer that samples the waste beam from the back of the tube. The photodiode signal is used in a feedback loop to lock the laser so the modes are of equal amplitude. (See: U.S. Patent #3,701,042: D.C. Motor Circuit for Rotating a Polarizer and Providing a Detector Synchronizer Signal for a Laser Stabilizing System.) Ironically, this is actually closer in function to the LCD optical switch of the 5501B and later lasers, than the polarizing beam samplers of the 5500C and 5501A that followed the 5500A. Since the 5500C/5501A tube has no waste beam exiting the laser tube, duplicating this function would be a bit of a challenge.

    The 5500A is in the same size case as that of the 5500C. The main difference between the 5500A and 5500C is what's at the front of the laser. The 5500A has interferometer optics and detectors for both REF and MEAS within the case. The 5500C has two channels of optical receivers but no interferometer optics. However, it was possible to install linear interferometer optics inside the 5500C to give it 5500A functionality.

    The 5500A is also unique among HP lasers since it is the only one with a run-time (hour) meter!

    There are photos of a 5500A in the Laser Equipment Gallery (Version 2.49 or higher) under "Hewlett Packard HeNe Lasers".

    For several original articles introducing HP's interferometer-based measurement system using the 5500A, see the Hewlett Packard Journal, August 1970.

    Also see Dave Meier's HP Laser Interferometer Evolution Page which includes a links to the early HP catalog pages.

    I have a 5500A laser (see gallery pages, above) which appears to be from around 1970 based on the date code found on a 74H10 TTL IC in the optical receiver. Except for the shape of the beam expander mount and color of the ballast resistor cover, my 5500A appears identical to the laser shown on the last page of the August 1970 HP Journal. An external HeNe laser power supply was used to perform initial tested before being connected to a 5505A Measurement Display. The laser tube starts and runs flawlessly with a raw output (after the beam expander but before the waveplates) of at least 370 µW and possibly as high as 450 µW. (The power varies with temperature as the tube warms up if not feedback stabilized and I didn't run it long enough by itself to determine the actual maximum power.) Even the low end of 370 µW would be considered excellent power for a much newer 5501A tube. The output power of the laser is between 106 and 150 µW (again depending on the temperature as it's not locked). If the locked output is anywhere near the higher end of this range, then it's basically like it was when it was last serviced. There is a note inside the laser saying: "120 µW August 1978". Perhaps the tube was also replaced at that time. The reason for the large difference between tube output power and laser output power is that the waveplates cut the power by 15 to 20 percent, and the internal interferometer optics suck up approximately half of the remainder since most of the F1 frequency component doesn't exit the laser.

    When first attached to a 5505A, the laser powered up and locked instantly, and within a couple minutes, I was able to make sub-micron measurements! But, then at some point while my back was turned, the original HeNe laser power supply inside the laser head failed. Hard to believe! Not like the thing has probably been turned on for the first time in 20+ years! :) I don't know if the failure was in the two transistor driver, or inside the potted HV module, which is beautifully made in clear semi-flexible plastic with no obvious damage. But there could be a shorted turn in the inverter transformer or a capacitor breaking down. The driver transistors passed ohmmeter tests and were getting equally warm, but the output was only going to around 1 kV and then dropping to 0 V, never lighting the tube. So, I replaced it with a small brick power supply from a barcode scanner, installed inside the original aluminum can to preserve authenticity. Unless one knew exactly where to look, there would be no way to tell that it wasn't totally original.

    One thing that's probably only of curiosity value is that both the HeNe laser HV power supply and the PZT HV power supply are driven from a common oscillator which must be running for the PZT tuning to work. Without tuning, the 5505A readout may still function, but the RESET button will keep flashing. Newer versions of the 5500C, as well as the 5501A use independent self-oscillating inverters in ugly bricks made of hard tan potting compound for these two power supplies. The earliest 5500Cs are probably similar to the 5500A.

    It's extremely easy to align the interferometer with my home-built authentic replica of the retroreflector mount shown in the 1970 HP Journal article. As long as it adjusted so the return beam enters the optical receiver aperture or even the tiny alignment holes in the laser head turret, the system is happy.

    And here is the genuine imitation authentic setup hot off my time machine:

    More information and photos from early HP manuals and brochures, and elsewhere can be found at Dave Meier's HP Laser Interferometer Evolution Page.

    The cable wiring is given in the next section since it is the same for the 5500A and 5500C.

    More to come.

    Notes on the HP-5500C Two Frequency HeNe Laser

    The 5500C is one of the earliest of the HP metrology lasers (or "Laser Transducer" in HP-speak), only preceeded by the 5500A (described above). Thus, the 5500C is somewhat similar to a 5518A or 5519A/B in that it has an internal optical receiver for the return beam. This is actually comprised of 4 photodiodes wired as two pairs with lenses surrounding the laser output aperture. The aperture wheel may select either the horizontal or vertical pair. Thus, two independent optical receiver channels are actually available. The "A" channel is fed from the top and left photodiode, and the "B" channel fed from the left and bottom photodiode. Thus, with suitable interferometer optics, a pair of measurement axes may be handled by the 5500C. The 5518A and 5519A/B only have a single optical receiver, but the assumption by the time of their introduction was that the smaller 5501A/B and 5517 lasers with external optical receivers would be used for multiple axis machines.

    There are photos of a 5500C in the Laser Equipment Gallery (Version 2.48 or higher) under "Hewlett Packard/Agilent HeNe Lasers".

    Also see Dave Meier's HP Laser Interferometer Evolution Page which includes a link to the early HP catalog pages.

    The 5500C uses a HeNe laser tube with PZT tuning that appears identical to the one in the 5501A, though the part number differs. (The 5500A has a very similar, though not identical tube. See the description and patent reference in the previous section.) The beam sampler for the feedback stabilization is of the common modern polarizing beam-splitter variety with the control loop driving the PZT of the laser tube to adjust cavity length. But, unlike the 5501A which only requires DC power supplies, the 5500 requires the mating 5505A Measurement Display to even turn on and stabilize since its HeNe laser power supply and PZT power supply are controlled by the 5505A. Although the HeNe laser power supply could be run open loop with a variable DC voltage, this would not provide current regulation. However, the PZT power supply of later 5500Cs which appears to be a potted module inside more potting, can be used as a stand-alone PZT, PMT, or other variable HV low current power supply since its output is fairly linear with respect to input from 0 to 15 V, which is multiplied approximately by somewhere between 100 and 200 to produce the output voltage. (Although I have not seen it specifically stated, the PZT power supply appears to be capable of more than 2 kV based on the 5501A schematics.) Both HV Control and PZT Control are really just the power input to a self oscillating inverter. (Very early versions of the 5500C and the 5500A have the inverter transformers and other high voltage components potted inside metal cans with the driver circuitry on separate PCBs fed from a common oscillator.)

    The pinout for the self contained PZT power supply module is:

    5500A/C and 5505A connector pinout

       Pin      Function
     ------------------------
        A       Gnd
        B       DOPPLER (A)
        C       +5V
        D       LOCK (A)
        E       HV CON
        F       REF TRIP
        G       -15V
        H       BEAM AL
        J       PZT MON
        K       REF (A)
        L       GND
        M       REF (B)
        N       DOPPLER (B)
        P       NC
        R       LOCK (B)
        S       LASER I
        T       +15V
        U       PZT CON
    

    If constructing your own cable, the wires to pins B and N should be shielded twisted pair, shield to pin A, and the wires to pins K and M should be shielded twisted pair, shield to pin L. The shield probably isn't critial for relatively short cables, but use the twisted pair. Size the voltage (+5 and +15) and Gnd wires to handle a couple amps. HV Control may also need to supply some current.

    Notes on the HP-5501 Two Frequency HeNe Laser

    The 5501 is the successor to the 5500. However, the 5500C and 5500A had optical receivers for a return beam built-in. (Two in the case of the 5500C.) The 5501 requires an external optical receiver for each axis, but thus has no limitation on the number of axes other than there being adequate beam power.

    The 5501B is a functional replacement for the 5501A. Locking of the 5501B typically takes 5 to 9 minutes compared to 10 seconds or so for the 5501A, but this is of no consequence for machines that are run for hours or years. In terms of optical characteristics, and power requirements and reference signals (including connector pinouts), they are equivalent. However, the 5501B lacks the Diagnostic (J3) connector of the 5501A, so other system components may not be happy and some substitutes may need to be provided. Going the other way doesn't have this issue, but if a 5501A is installed in place of a 5501B, it may be necessary to press the Retune button from time-to-time whereas there is no such button or need on the 5501B! This may be anywhere from a few hours to never, but it would be a good idea to do this periodically at convenient times between measurement runs, at least until the system has reached thermal equilibrium. Performing a Retune cycle does not compromise the accuracy in any way. Once the Retune LED goes out, it's ready to go again. Even from a cold start, a laser may go 12 hours or more without requiring a Retune. After that, once a day may be more than sufficient.

    Operation and service manuals for the HP-5501A and HP-5501B may be found on the Hewlett Packard/Agilent Metrology Laser/Interferomter Page.

    HP-5501A

    The 5501A, 5500C, and 5500A appear to be the only models that use a special HeNe laser tube which has its cavity length fine tuned by a PieZo Transducer (PZT). (The 5501B and all successors still use a special HeNe laser tube but it has an internal heater for fine cavity length control rather than a PZT.) The OC (front) mirror and PZT with the HR (rear) mirror inside the laser tube are held in place by a spring against a spacer tube (the actual bore of the HeNe laser) made of Zerodur, a special very low thermal expansion coeffient glass/ceramic developed by the Schott Glass. In fact, longitudinal mode cycling is virtually non-existent with these lasers.

    Compared to the 5500C, the 5501A is in a much smaller lighter case with simplified optics and totally different electronics. See Interior of the HP-5501A Laser Head - Left Side and Interior of the HP-5501A Laser Head - Right Side. The HeNe laser tube dominates the interior space in both views. The high voltage piezo driver power supply brick is visible under the magnets at the center of the tube. The HeNe laser power supply brick is underneath the output end of the tube. The piezo driver electronics circuit board at the far right end of the right side view. The optical sensor circuit board is at the far left of the left side view.

    The naked tube is shown in HP-5501A Laser Tube Removed From Magnet and Output Optics Assembly. The normally enclosed part is really just a very think walled fine-ground bore inside an outer glass envelope. A spring (visible through the glass at the left) at the rear holds the PZT, HR mirror, bore, and OC mirror in place. No adjustment is possible. There are distinct multiple spots on the card because the output window is at a slight angle and not AR coated.

    Both the HeNe laser power supply and piezo power supply run off the -15 VDC power supply. An interlock switch (easily defeated) prevents operation with the cover removed. In the 5500A and 5500C, these power supplies are regulated by the 5505A Measurement Display. In the 5501A, the potted power supply bricks have no inputs other than power. Rather, current and voltage regulation are accomplished by controlling the input current. For the HeNe laser power supply in the 5501A, as well as later versions of the 5500C, while the passive HV components are buried in potting compound, the two 2N5192 driver transistors are mounted on the outside of the brick and are replaceable. However, from my experience, when the transistors blow, there is probably a fault in the potted section so replacing them doesn't help, I've successfully replaced the 5501A HeNe laser power supply with a common barcode scanner brick, the Laser Drive model 103-23. This has an input rance of 21 to 31 VDC at less than 0.5 A, and an output of 1.1 to 1.5 kV at 3.5 mA (fixed). The 3.5 mA is a bit higher than the labeled current on most 5501A tubes, but seems to be acceptable and actually beneficial for some high mileage tubes that like to run at a slightly higher current. But, adjustable versions of these supplies are readily available. I connected the supply between the HV Control (white/green wire) and -15 VDC (purple wire) with the pot set fully CCW (max current). This assures that the 5501A current regulator will not attempt to compete with the brick's internal regulator. However, with some HeNe laser power supplies, it may be possible to use the 5501A's regulator to *reduce* the current in a stable manner. This is left as an exercise for the student as it may not work in general.

    The output of the laser tube is passed through a quarter wave plate to convert the circular polarization to orthogonal linear polarization components, and then through a half wave plate to rotate the linear polarization by an arbitrary, but fixed angle to line the two linearly polarized components up with subsequent optics. These waveplates are adjustable with respect to orientation around the optical axis of the laser as expected. But the angle of each waveplate along one of its principle axes with respect to the optical axis of the laser is also adjustable - presumably to optimize the 1/4 or 1/2 wave performance. They are both very thin and may be zero order waveplates, possibly made of optical grade mica. The beam is then expanded and collimated and passed through an angled partially reflecting plate located just beyond the collimating lens on the laser tube assembly. This deflects about 20 percent of the beam to a polarizing beamsplitter which sends each component to its own photosensor to provide the frequency control feedback. A control loop uses these signals to adjust the PZT, and thus resonator length, so that the two signals are of equal amplitude. The difference of the two signals is the frequency/phase reference.

    The laser stabilization control algorithm is actually dirt simple: The voltages from the photodiodes corresponding to the two polarization components are compared in an integrator which maintains the PZT voltage at a level so they are equal. (There is an adjustment to compensate for slight differences in amplitude resulting from beamsplitter ratio and photodiode sensitivity.) While crude and simple to implement, this approach is adequate to achieve the needed stability. The electronic reference signal is derived from the slight residual difference frequency present in one of the polarization components.

    While the spacer rod has a very low coefficient of thermal expansion, it isn't exactly zero, so as the system heats up (over hours), the cavity length will still change slightly. Eventually, the PZT voltage may be unable to compensate. The PZT voltage is compared with fixed upper and lower limits which are well within the range over which locking is assured. When either limit is passed, the "Tune Fault" flag is set turning on the "Retune" LED and asserting the "Retune_Status" signal. The laser may be retuned via a pushbutton or external TTL signal). This clamps the PZT control voltage at its lowest value for a short time and then releases it to ramp up to the lock point. Requiring external intervention (whether manually or by computer) assures that a measurement will never be made when the laser isn't stable, nor will one in progress be interrupted due to the laser relocking unexpectedly.

    The 5501A laser head requires +15 VDC and -15 VDC for power. (There is also a +5 VDC pin but it is an output according to the manual.) The two voltages (and common) are all that is needed to operate the laser head but an interlock switch (on the right side at the rear of the case) must be depressed to turn on the laser tube. I haven't yet looked at the output with a photodiode or scanning Fabry-Perot interferometer but after a few seconds, the "Retune" LED goes off, similar to if the "Retune" button is pressed. And then there is a stable reference signal. I have since acquired an operation and service manual for the HP-5501A laser head which confirms the information above.

    HP-5501A reference connector J1

       Pin      Function                              Socket View
     ---------------------------------------------         A
        A       Accessory +15 VDC fused                    o
        B       +15 VDC return                         D o   o B
        C       Reference (difference) frequency           o
        D       Complement of J1-C                         C
    

    HP-5501A power connector J2

       Pin      Function                              Socket View
     ---------------------------------------------
        A       +15 VDC input                          D o  o A
        B       -15 VDC input
        C       +5 VDC output (test-point)             C o  o B
        D       Power ground
    

    HP-5501A diagnostic connector J3

       Pin  Function      I/O   Comments
     ------------------------------------------------------------------------------
        A   +15 VDC TEST   O    Sample for diagnostics
        B   -15 VDC TEST   O    Sample for diagnostics
        C   +5 VDC TEST    O    Sample for diagnostics
        D   SYS COM        -    Ground/return
        E   Retune_CMD-    I    Active low to initiate PZT tune/check cycle.
        F   Retune_Failure O    Active high output indicates failure of PZT
                                 tune/check cycle.
        J   Retune_Status  O    Active high when tune/check cycle is in progress.
        K   Laser_Cur_Err  O    Active high indicates laser tube current is
                                 outside acceptable limits.
        L   Error          O    Logical OR of J3-J, J3-K, and PZT voltage outside
                                 of specifications.
        M   L I Mon Test   O    Laser current sample for diagnostics.
        N   PZT Mon Test   O    PZT voltage sample for diagnostics.
        P   Ref OK Status  O    Active low diagnostic signal indicates laser
                                 is properly tuned.
    

    HP-5501B

    The 5501B uses a very different HeNe laser tube of more conventional contruction but which has an internal heating coil for cavity tuning. It is also physically similar to the tube in the 5517B/C/D/E lasers. The 5501A and 5501B are in the same size case and look very similar externally. See Interior of the HP-5501B Laser Head - Left Side and Interior of the HP-5501B Laser Head - Right Side. See Internal Structure of Hewlett Packard 5501B and 5517B/C/D Laser Tube Assemblies for the anatomy and Tube Assembly Used in HP-5501B and HP-5517B/C/D Two-Frequency HeNe Lasers for the a photo of the beast. The actual glass laser tube is shown in Hewlett Packard/Agilent 5501B and 5517B/C/D HeNe Laser Tube Removed from Magnet/Optics Assembly

    Like the 5501A, the 5501B also requires only +/-15 VDC to power up. There is no case interlock on the laser I have, though one is shown in the manual so I assume this is either an addition or deletion depending on version. When power is applied, at first, only the +/-15 VDC power LEDs come on. After 3 to 6 minutes, the "READY" LED begins to flash at about a 1 second rate. After another 1.5 minutes or so, the "Laser On" LED comes on and the beam appears. Finally, a minute or so after that, the READY LED comes on solid and remains that way. Note that this method of turning on the laser only after the temperature set-point has been reached is unique to the 5501B. All the other HP/Agilent lasers turn on the laser with power-on. Specific times for one test beginning from a cold start at an ambient temperature of about 65 °F were: (min:sec) 3:15, 1:35, and 0:48. The first of the times is called "preheat" and is determined by how long it takes for what HP calls the "laser rod" to reach operating temperature. The laser rod is the large glass bore of the laser tube to which the mirrors are clamped at either end. It thus controls cavity length. The temperature is sensed by disabling the heater drive and measuring the resistance of the heater coil every 25.6 seconds. The warmup is much shorter if the laser is restarted after having been running: 1:00, 1:20, and 0:50. Only after the READY LED is on solid, do the reference signals appear. The 5501B adjusts the cavity length so that the two polarized components of the beam (the Zeeman split longitudinal modes) have equal power. Interestingly, there is only one photodiode sensor which is alternatively switched between beams using a liquid crystal polarization rotator. A sample-and-hold then outputs to the error amplifier of the optical mode control feedback loop.

    There are two outputs of about 5 to 6 V p-p (centered about 0 V), 180 degrees out of phase. For this laser, the reference frequency is about 1.80 MHz. There is no need for a "Retune" button as with the PZT based system of the 5501A. Also unlike the 5501A, there are no other signals to or from the 5501B (no large connector), only the +5 VDC output on the power connector, and a fused +15 VDC output on the reference connector.

    Although the control board inside the 5501B looks similar to that of the "small" 5517 lasers, it is not interchangeable with them as some functions like the heater drive have are on the small "connector PCB" at the back-end of the case.

    There were two problems with the first 5501B I acquired that I had to deal with. The first was that the tube wouldn't stay on stably at the 3.5 mA setting (fixed) of the power supply but works fine at 4 mA. Such a condition is usually due to the tube having been run for a long time, which wouldn't be surprising with a surplus 5501B laser head. Since the existing power supply has no current adjustment, I needed to find a similar size HeNe laser power supply brick (1"x1.5"x4" or smaller) that will run on 15 VDC to replace it that can be set for 4 to 4.5 mA. The tube seemed healthy enough otherwise. I installed one that runs the tube at 4 mA but draws more DC input current than the original, and possibly for that reason, the controller aborts and resets after about 1 second when it turns the laser on. For now, to get around this, I have connected the HeNe laser power supply directly to the raw -15 VDC and added a transistor to drive its enable input when the original laser power turns on. That appeared to work fine. But after replacing the cover, the laser tube wouldn't come on. :( I discovered that it needed the room light to start! I had thought this to be a relatively rare malady for HeNe laser tubes, but more common for neon lamps and glow-tube fluorescent lamp starters. However, it turns out that a decent percentage of HP/Agilent HeNe lasers start more quickly when illuminated. So, there is now a decorative red LED shining on the back of the tube which is lit when the laser is powered. An HeNe laser power supply with a higher starting voltage would probably make this kludge, oops, feature, unnecessary. But no one will ever know about it. :) While many of these higher mileage HP/Agilent lasers can benefit from this addition, since the 5501B turns the laser on and expects it to come on quickly, it is more critical than with the other lasers like the 5517s that really don't care whether the laser is outputting a beam or not, until they actually try to lock. However, in either case, if the laser takes too long to lock, associated equipment like the 5508A Measurement Display may flag it as a failure.

    HP-5501B reference connector J1

       Pin      Function
     ---------------------------------------------
        A       Accessory +15 VDC fused
        B       +15 VDC return
        C       Reference (difference) frequency
        D       Complement of J1-C
    

    HP-5501B power connector J2

       Pin      Function
     ---------------------------------------------
        A       +15 VDC input
        B       -15 VDC input
        C       +5 VDC output (test-point)
        D       Power ground
    

    Notes on the Hewlett Packard 5505A Measurement Display

    The HP-5505A is a part of the HP-5525A/B and HP-5526A Laser Measurement Systems, which are designed for very precise single-axis measurements in a machine tool environment.

    The HP-5525A was used in the original HP interferometer introduced around 1970 and includes the HP-5505A Measurement Display and the HP-5500A two-frequency HeNe laser head. The 5500A laser has the interferometer optics built-in and thus only requires an external retroreflector (cube-corner) on the moving part to be mesaured. The HP-5525B upgraded to the 5500C laser head which requires external interformeter optics but allows for two axis measurements (with a pair of 5505As!). The 5526A seems to have added a variety of options and but it's not clear how it really differs from the 5525B.

    The 5525A/B and 5526A can be set up in the field with relative ease with a minimum number of individual components and no need for a control computer as its basic functions are built-in to the HP-5505A. It provides for the stand-alone precise measurement of position and velocity. But straightness and angle are not directly supported.

    The 5505A implementation of the display function is all done in MSI TTL logic with a pair of 36 bit counter/registers for REF and DOPPLER (same as MEAS for other HP lasers), with a decimal adder/subtractor to generate the result. This is all on multiple PCBs and while there is one labeled "Program", there is no actual microprocessor controlling the system.

    The 5525A, 5525B, and 5526 all require the 5505A display but differ in the laser and options. (There may be some minor changes required to convert an older 5505A to be used in a 5526A system.) The following is from the N4MW HP 5526A Documentation Page which also has links to the actual HP catalog pages for each system.

    5500Cs have also been showing up with internal linear interferometers like the 5500A. I haven't seen any reference to this as a standard product though. I wonder if they were retrofits for customers who found their original 5525A configuration adequate or whined when their 5500As went bad and wanted an exact replacement.

    The HP-5525A/B and HP-5526A are very obsolete, but many are still in use. 5505As show up on eBay, often for next to nothing. To non-interferometer geeks, the set of Nixie tubes is probably more valuable than a working unit! However, being so old, they often have problems, and at least some of the ICs like the Nixie tube drivers are proprietary parts and no longer available.

    For info (or lack thereof):

    The laser connector on the back of the 5505A is the same type and has the same pinout as that on the 5500A and 5500C heads. The 5508A supplies +/-15 VDC power for the laser head. It also controls both HeNe laser power supply current regulation and PZT laser tuning.

    To use the 5505A with a 5500A, all that's required is a 05500-60025 cable and a retroreflector (cube-corner) as shown in Original HP-5525A with HP-5500A, HP-5505A, and Retroreflector - View 1. (Additonal photos can be found in the section above on the 5500A laser.) It's straightforward to make a cable. The connectors are standard and everything is wired 1:1 at both ends. To use the 5505A with the HP-5500C laser also requires external interferometers optics. All of the standard configurations that have separate outgoing and return beams should work.

    To use the 5508A with other HP laser heads will require a custom cable and possibly a separate optical receiver which can be any version of the 10780 (A, B, C, F, U). However, some circuitry may need to be added to the 5505A to keep it happy by making it think it still has control of PZT tuning.

    FWIW in the "well that's interesting department", here is the board set from another 5505A. This is a rather vintage sample, S/N: 2016A01966, which puts its manufacturing date around 1970:

     Slot       Name           Part #        Additional Markings
    -----------------------------------------------------------------
      A1   Analog Board      05505-60001     Series 1920        03L
      A2   Clock Board       05505-60002  B3 Series 952-2       03F
      A3   Accumulator Board 05505-60034     Series 1920     01403F
      A4   Accumulator Board 05505-60034     Series 1920     01403F
      A5   Adder Board       05505-60005     Series 952         03F
      A6   Algorithm Board   05505-60006     Series 952      00203F
      A7   Program Board     05505-60007     Series 2240     23103F
      A8   Function Board    05505-60058     Series 1920     23903F
      A9   Multiplier Board  05505-60049     Series 1948     23103F
      A10  D-Register        05505-60010                        03F
      A11  Display Board     05505-60011     Series 1324        03F
      A12  Power Suppy Board 05505-60012     Series 1940     01503F
    

    Notes on the Hewlett Packard 5508A Measurement Display

    The HP-5508A is a part of the HP-5528A Laser Measurement System, which is designed for very precise single-axis measurements in a machine tool environment.

    The HP-5528A includes the HP-5508A Measurement Display, an HP-5518A two-frequency HeNe laser head on heavy duty tripod, a variety of interferometer optics, and optional environmental sensors, and other stuff. :) So, this system can be set up in the field with relative ease with a minimum number of individual components and no need for a control computer as its basic functions are built-in to the HP-5508A. It provides for the stand-alone precise measurement of position, velocity, angle, and straightness by using the appropriate interferometer assemblies.

    The 5508A implementation of the display function consists of X16 frequency multipliers for REF and MEAS, which are then applied to separate 16 bit up counters. These initiate a non-maskable interrupt to the microcontroller when either exceeds the half way point (the MSB gets set). They are then stopped while separate small "swallow counters" absorb pulses occurring while the interrupt is processed and the position is updated. The microcontroller is kept rather busy, but since it doesn't have all that much else do do, should be quite happy. :)

    Although the HP-5528A is considered obsolete by Agilent, it's still very useful and surplus systems or components are now much cheaper. The Agilent 5529A Dynamic Calibrator is the replacement for the 5528A. Rather than a dedicated display, it requires a PC (not included). But aside from the slightly higher REF frequency of the 5529A laser head (generally irrelevant in these types of typically slow speed applications), the precision is no better than that of the 5528A.

    The laser connector on the back of the 5508A is the same type and has the same pinout as that on the 5517 and 5518A laser heads. The one "No Connect" pin on the 5517 connector (pin A) is used to drive the MEAS beam level indicator on the front of the 5508A. The meter reading seems to be proportional to the current flowing out of this pin, from an internal +5 VDC source, with approximately -2 mA being full scale. Pins B and C that are also unused on the 5517 lasers now get ~MEAS and MEAS. (They are connected to line drivers on the 5517 lasers but only used for testing.) The 5508A supplies +/-15 VDC power for the 5518A laser head.

    There are several other connectors on the rear of the 5508A for various environmental sensors (temperature, pressure, etc.) and even a remote control. (I'd like to see that!) There is also a IEEE-488/GPIB/HP-IB interface for control and data acquisition.

    To use the 5508A with a 5518A, all that's required is a 10793A/B/C cable, which is wired 1:1 at both ends. To use the 5508A with other HP laser heads will require a custom cable, and possibly a separate optical receiver which can be any version of the 10780 (A, B, C, F, U). Except for ~MEAS and MEAS, all signals are wired 1:1 for 5517s. However, I'm not sure whether all versions of the 5508A will work over all velocities with any of these except the 5517A. The 5517B/C/D have REF frequencies, and result in possible MEAS frequencies, that may be too high, at least under some conditions. It is also possible to use the 5508A with the 5501A/B lasers, but the connectors are totally different. For the 5519A/B laser, these same issues apply. The connector of the 5519A/B provides only the REF and MEAS signals as this laser head has an internal power supply that plugs into the AC line. To use the 5508A with 5500A/C lasers would require additional circuitry to provide the HeNe laser current control and PZT tuning.

    One way to get around the REF frequency issue is to build a divide-by-two circuit for REF and MEAS that goes between the laser and 5508A. This is simply a dual differential line receiver, a pair of D flip-flops, and a dual differential line driver. Add a switch to select straight through or divide-by-two if desired. Of course, measurements values will now be halved unless a plane mirror interferometer is used, in which its doubling will be exactly offset by the halving in the divider!

    The only thing that won't work when using laser heads other than the 5518A without additional effort is the beam level meter on the 5508A front panel, fed from pin A of the laser connector. This seems to require a current of 0 to 2 mA to Ground from an internal +5 VDC source. The test-point on the outside of all 10780 receivers generates a voltage related to signal level, but simple voltage to current converter circuit (1 transistor and a few resistors) is then needed to interface to the meter input. If you're not a purist, this can just be ignored as it is not used anywhere and its only purpose is to aid in optical alignment and confirmation of adequate signal. But the 10780 test-point services the same function.

    In summary:

    I've attached my 5508A to my measurement test setup and initially have been using a 5517D laser head with it. I'm a bit surprised that this even works with the 5508A as it has a REF frequency almost double the maximum of even the later versions of the 5518A laser. I don't know if it will run at full velocity, but for modest speeds, the readings seem to be fine. But I intend to add the divide-by-two circuit eventually as insurance. I've installed a DB9 disconnect to make this easier. It has the following pinout (for my own reference!):

      Pin   Function
     ------------------
       1   +15 VDC
       2   Power GND
       3   ~MEAS
       4    MEAS Return
       5    MEAS
       6   NC
       7   ~REF
       8    REF Return
       9    REF
    

    The module would have an option for a gain of 1 or 5 and divide ratios of 1, 2, or 4. The higher gain is needed for some late model Agilent 5517s which have the Newest Digital Control PCB and appear to output lower amplitude REF and ~REF signals. So far, I've seen this on a 5517D and 5517E. With these, the 5508A would never acknowledge "LASr UP" even though the laser itself came READY, and would eventually time out with "LASr FAIL" even though my home-built measurement display was perfectly happy.

    Notes on the HP/Agilent 5517 Two Frequency HeNe Laser

    There are several versions of the HP (now Agilent) 5517 series. All are very similar to the 5501B in terms of the physical characteristics of the HeNe laser tube and the control electronics. Except for the 5517A, all versions are also in the same small rectangular case as the 5501B. However, the 5501B has a pair of connectors, one for power and the other for the reference signal; all 5517s have a single connector for power and signals. (Pinout below.) For the 5517s (and 5518A and 5519A/B), F1, the lower frequency output is horizontally polarized and F2, the higher frequency output is vertically polarized.

    When warming up, the difference frequency only appears for perhaps 10 percent of the time during mode sweep - only when the Zeeman modes are near equal amplitude on the split neon gain curves. The difference frequency is maximum and the output power is minimum at the center of this region, which is also where it will eventually lock. This is normal behavior for these lasers based on what is shown in Axial Zeeman Split HeNe Laser Mode Behavior. Note that while there may be some other longitudinal modes present, at least one of the other pairs is too weak to lase and thus there will be no beat except from the main pair, and then only when very close to being positioned symetrically on the Zeeman-split neon gain curves. In fact, only the main F1/F2 mode is present when locked. While other rogue mods would not produce any beat signal, they would result in problems in the interferometer and possible transient errors.

    HP/Agilent 5517A/B/C/D

    The 5517A is in the larger case with a trapezoidal shape but does not include an optical receiver with an external input like the physically similar 5518A and 5519A/B. Converting a 5517A into a 5518A is simply a matter of installing the internal optical receiver PCB and replacing or removing the shutter assembly on the front of the laser. Of all the 5517s, it is the only one to have a tube that might appear to be of lower manufacturing quality, and is larger than the the tube in the others, and the tube in the 5501B. But the real reason may be that it is cast with precise locating pegs so that a tube can be swapped without requiring even minimal alignment. And the 5517 tube is physically nearly interchangeable with the 5518A and 5519A/B. (For installation in a 5519A/B, there is a small piece of metal that needs to be removed to provide clearance for the 5519's internal DC power supply.) More below. Interior of the 5517A Laser Head shows the major laser/optics components of the Hewlett-Packard 5517A laser. The actual HeNe laser tube is inside the gray cylindrical housing which also has a cylindrical magnet for Zeeman splitting the HeNe laser lines to create the difference reference frequency in the interferometer application. See the previous sections for more information on these two-frequency lasers. It's very well made but not particularly unusual except for a heater coil wrapped around the bore inside the HeNe laser tube, used to control cavity length very precisely. Here are some photos and a detailed diagram of the internal construction:

    Unlike most other internal mirror HeNe laser tubes, there are no mirror adjustments. The mirrors are held in place against the thick glass bore (or mirror spacing rod as HP calls it) by spring pressure alone. So, the ends of the bore and mirrors must be ground to a precision sufficient for alignment to be near perfect. The distance between the mirrors is 127 mm in the 5517B corresponding to an FSR or longitudinal mode spacing of about 1.18 GHz. The purpose of having multiple glass backing disks behind the OC mirror is not really known. But along with the appropriate length spacing rod, these would provide a means of selecting cavity length by using a variable number of them without having to manufacture multiple (glass) tube designs. So, everything could be identical except for the spacing rod and number of backing disks installed behind the HR mirror and in front of the OC mirror, and perhaps the length of the associated spring(s). Having at least one backing disk may be desirable to reduce stress on the mirrors from the springs, but some tubes don't have any behind the HR mirror. So backing disks may simply be present to accomodate different spring sizes so the bore is properly positioned. Visual examination of 5501B, 5517A, 5517B, 5517D, and 5519B tube assemblies from the front after removing the beam expanders and waveplates has been inconclusive. And the glass tubes from 5517A, 5517B, and 5517D lasers appear physically identical in all dimensions that matter from the outside, which includes the cavity length. I was expecting the low REF frequency 5517A to have some obvious difference compared to the high REF frequency 5517D, but this does not appear to be the case. So, if they are all the same, what is the purpose of the extra space in front with 5 backing disks and the extra length spring? I have also checked a working 5517D, as well as a 5517E and 5517FL using a Scanning Fabry Perot Interferometer (SFPI) to measure the mode spacing during warmup (and thus cavity length based on c/2f) and confirmed what I already knew from physical examination - that the 5517D is the same as the 5517B (1.18 GHz, 127 mm). But the 5517E and 5517FL were found to be around 1.5 GHz implying a cavity length of around 100 mm. At this point, I assume the cavity length is also 127 mm in the 5517C, 5518A, and 5519A/B, but 100 mm in the 5517G. If anyone has taken more of these tubes to bits or X-rayed them, or has measured the longitudinal mode spacing either with an SFPI or a high speed photodiode and RF spectrum analyzer, and can thus shed some more coherent light on this topic :), please contact me via the Sci.Electronics.Repair FAQ Email Links Page. Also of particular interest would be the reflectivity of the OC mirror, which is also critical in determining the REF frequency. (See the section: Speculation on Axial Zeeman HeNe Laser Behavior.)

    In terms of cavity length, the 5517 tubes are very similar to typical 1 mW barcode scanner HeNe laser tubes. Their relatively low power - always under 1 mW and often much less - may be accounted for by the combination of the magnetic field, precisely specified OC mirror reflectance, and possibly isotopically pure helium and neon and fill pressure, to achieve the desired Zeeman beat frequency range for each model of the 5517. This is consistent with the generally lower output power for higher REF frequency lasers - 5517Ds almost always have much lower power than 5517As when new. The heater connections with red and purple wire stubs sticking out can be seen at the left of the tube. The purple one also attaches to the cathode via a piece of springy sheet metal, no welds. The anode connection goes through the outer glass envelope but there does not seem be a glass seal into the bore, but simply a hole drilled in it to coincide with the anode location. Even after totally disassembling a tube, it's not clear what prevents the discharge from bypassing the bore unless there was some other sealant present that got lost in the process. But there are no traces of any, nor anything visible in two other tubes. The fit between the bore and surrounding glass cylinder is quite close but even this wouldn't normally guarantee that the discharge goes through the bore, especially on hard-start tubes. The magnet is a single piece cylinder of what looks like AlNiCo with an inner diameter a few mm larger than the tube. It extends at the left and right ends of the tube to exactly where the discharge begins and ends.

    I have measured the tube+ballast voltage on one 5517B laser to be about 1,675 V at 3.5 mA. This may be a bit greater than typical since the tube was high mileage, though with normal start and run behavior. The ballast was 100K so the actual voltage across the tube was 1,325 V, This is still somewhat greater than expected for a laser with a power output of less than 1 mW. Part of the reason may be that the magnet probably slightly increases the tube voltage.

    Here are the 5517B HeNe laser tube parameters as best I have been able to determine them so far, compared with typical short barcode scanner HeNe laser tubes. Except for tube voltage (which I haven't measured and am only estimating) and tube current (which is always spec'd to be 3.5 mA), the parameters for the 5517B are for one very healthy sample. The total length, planar HR mirror, divergence (without beam expander), and beam diameter are similar or identical for other models. The parameters for the barcode scanner tube are for a range of typical models. The divergence for a particular model barcode scanner tube is usually achieved by either the specific curvature of the outer surface of the OC mirror glass or with an external lens glued to it, but the cavity design including the OC RoC (radius of curvature) is the same for all.

                                                            Barcode
       Parameter                         5517B Tube       Scanner Tube
     --------------------------------------------------------------------
       Output Power (Tube only)            0.8 mW        0.5 to 1.0 mW
       Total Length                        194 mm        125 to 155 mm
       Cavity Length                       127 mm        115 to 145 mm
       HR Mirror Radius of Curvature       Planar           Planar
       OC Mirror Radius of Curvature       140 mm        250 to 300 mm
       OC Mirror Reflectance               98.5 %        99.0 to 99.5 %
       Beam diameter at output surface      ~1 mm        0.4 to 0.6 mm
       Beam divergence                     ~10 mR        1.7, 2.7, 8 mR
       Discharge Length                    100 mm         60 to 75 mm
       Operating Current                   3.5 mA          3 to 4 mA
       Operating Voltage (tube only)       1.3 kV        0.7 to 1.1 kV
    

    Of note here is the relatively low OC reflectance of 98.5 percent. This is one of the primary parameters that determins the REF frequency. All other factors being equal, reducing the reflectance increases cavity loss which increases the REF frequency.

    Plot of Hewlett Packard Model 5517C Stabilized Laser During Warmup shows how a typical 5517 laser behaves. This is a 5517C rather than a 5517A but the locking algorithm is the same. Note that the entire warmup period from laser on to locked is only around 3.5 minutes because of the internal location of the heater for the active mode as noted above. A laser with the more common external heater could take 20 minutes or more to lock. The control algorithm is a bit more sophisticated than used on some other stabilized lasers, checking periodically for the status, and switching from "Warmup Mode" to "Optical Mode" about half way through the warmup period, at which point the READY LED starts flashing. A short while after it locks is when the READY LED comes on solid.

    Plot of Hewlett Packard Model 5517C Stabilized Laser Near End of Warmup shows the 5 mode cycles just before locking and the final transition to the locked state. The peculiar shape of these Zeeman-split modes is clearly evident in this expanded view.

    The beat frequency is shown for the last 5 cycles and after locking in both these plots. This is the actual measured frequency captured along with the vertically and horizontally polarized modes and total output power. (Showing the frequency plot earlier would be a mess.) The beat only appears for a small percentage of the mode cycles with some variation during the time it is present, peaking when the F1 and F2 amplitudes are equal, and only when F1 is falling with increasing temperature. There is no beam when F1 and F2 are equal but F2 is rising with increasing temperature. The reason for this becomes evident from the simplified diagram in Axial Zeeman Split HeNe Laser Mode Behavior, or more accurately in HP-5517 Zeeman Split HeNe Laser Mode Behavior. The second diagram has been specifically crafted based on the mode plots, above, and thus more accurately represents the actual mode behavior of the 5517. Both show 5 snapshots of most of a mode sweep cycle starting with the cavity being 1/4 wavelength too short at the top and ending with being 1/8 wavlength too long. (The case of 1/4 wavelength too long would be the same as the first, 1/4 wavelength too short). Only when the longitudinal mode is centered between the Zeeman-split neon gain curves will there be a beat. In addition, the mode amplitudes are changing rapidly as the cavity expands at those high slope locations on the gain curves. When the cavity length changes (longer or shorter) by 1/4 of the lasing wavelength of approximately 633 nm, the amplitudes are again equal, but the two separate longitudinal modes are oscillating far apart and there is no beat. Note that the red and blue plots include the F1 and F2 amplitudes, but also may have contributions from other longitudinal modes derived from the same split gain curve which thus have the same original circular polarization.

    Note that as the tube ages with use, the gain declines and the width of each gain curve that is above the lasing threshold decreases. Eventually, with a really high mileage tube, there may be no overlap at all and the beam will probably disappear for a part of the mode sweep cycle. But it is exactly at that point where the Zeeman beat would be generated, so it will also disappear entirely. Lasers are generally taken out of service long before this happens, but I recently found one whose output power was so low that this behavior was present - or absent depending on your point of view!

    Here is how the article "An Instant-On Laser for Length Measurement" by Glenn M .Burgwald and William P. Krugein describes the operation of the laser tube in the Hewlett-Parckard Journal, Aug., 1970.

    "If an axial magnetic field is applied to a laser which is free from polarization anisotropy in either the mirrors or the plasma tube, the output splits into two frequencies of left and right circular polarization. First-order theory predicts that the frequency splitting is proportional to magnetic field strength and to the ratio of line Q to cavity Q. In the new laser, magnetic field strength is adjusted for a difference frequency of about 2.0 MHz. Line center is virtually midway between the displaced lines, so proper cavity tuning can be assured by adjusting for equal intensities of the lines."

    And they show a gain curve diagram even simpler than the one above. See that article for more details.

    The reason for the peculiar shape of the mode plot is not clear. It is probably due to a combination of Zeeman modes and normal longitudinal modes. However, the gain profiles need to be asymmetric to account for it. The simpified explanation of Zeeman splitting rarely takes into account the actual profile of the magnetic field which no doubt stretches and distorted the gain profiles. So, drawing a pair of nice bell-shaped gain curves really isn't accurate. Near the ends of the tube, there will be little magnetic field and the gain curve there will not be shifted very much. In the center, the field will be maximum and the gain curves will be shifted by a large amount. Thus, the neon atoms will see a variation in magnetic field along the bore - a summation or integral of gain curves varying with frequency. This is depicted in HP-5517 Zeeman Split HeNe Laser Mode Behavior. Here, the gain curves have been modified so that the results would be roughly similar to what was in the plots, above. And HP-5517 Zeeman Split HeNe Laser Mode Behavior Versus Mode Position on Gain Curve shows one complete mode cycle along with little split gain curves.

    Note that the locked mode amplitudes are not equal and that the locked beat frequency varies a bit after locking and does not stay at its maximum value as would be expected if the stabilization was optimal. This is not a quirk of the particular laser I'm using for these experiments as I've tested dozens with similar behavior. So, they aren't as perfect as we might hope! :)

    Also note the two smaller modes on the tails of the gain curves in the diagrams. They aren't present in all of these lasers. But if they are and align with the X and Y axes, then the only effect will be to slightly decrease the MEAS or detected REF signal level with respect to laser output power. However, if the are not aligned with the X and Y axes (e.g., at 30 degrees), they will cause level changes in the envelope of the signal from the optical receiver's photodiode due to self-interference in the interferometer. This is similar to what would happen if the primary Zeeman modes were misaligned, or not pure. The consequences could be transient position errors but only during motion. The end-points would be accurate.

    Some later versions of the 5517 lasers have a totally redesigned electronics board with more digital circuitry. I don't know exactly when this changeover took place, or whether it's simply a special option, but several 5517Cs and 5517Ds from before 2000 had the old style and a 5517C from 2005 had the new style. So, perhaps Agilent insisted on reinventing the wheel. :)

    The glass 5517A HeNe laser tube itself is similar or identical to that of the 5501B. However, the tube's enclosure appears to have been cost-reduced: It is a base metal casting rather than being constructed from precision machined parts. (This is also true of the 5518A and 5519A/B.) All the other 5517s use tube enclosures that are physically similar to that of the 5501B. (Perhaps newer versions of the 5517A have the higher quality construction.)

    The output optics consists of a beam expander/collimator (the black and silver object just to the left of the power supply danger label) and an additional optical assembly to the left of this whose front and rear halves contain what appear to be AR coated optical quality mica pelicles oriented at slight, but different angles. The front and rear sections can be rotated independently and they were sealed with blue paint once the perfect orientations were found. The two mica (or whatever) pieces of the optics assembly (just after the beam expander) are adjustable waveplates. The first one is a 1/4 waveplate to convert the circular polarization of the Zeeman split output of the HeNe laser tube to linear polarization and the second one is a 1/2 waveplate to rotate the resulting linearly polarized components to be aligned along the horizontal and vertical axes. These can then be separated out with a polarizing beamsplitter at the detectors.

    The HP/Agilent lasers do not employ any sophisticated method of stabilization such as locking the Zeeman beat frequency (which chances slightly depending on where the modes are on the neon gain curve) to a crystal reference. They simply use the amplitudes of two orthogonally polarized signals in an analog feedback circuit as is common with most other stabilized HeNe lasers. However, here, the two polarizations are of the two Zeeman split components of the single oscillating mode rather than two separate longitudinal modes. The error signal is the difference between their amplitudes, which is forced to zero by temperature tuning of the cavity. And, in fact, there is no real need to have the frequency be precisely known or even constant over the long term, as long as it is stable over the short term. More below.

    The warmup/locking algorithm is straightforward, though just a bit different than used in many other stabilized lasers. When the laser is first turned on, it is in "Warmup Mode" and the heater, which is wrapped around the internal bore of the laser tube, is driven to reach a fixed temperature (set by the only pot on the electronics PCB). The temperature is sensed by periodically measuring the heater's resistance. This is done by disabling the heater driver, passing a small fixed current through heater wire (for 2.56 seconds out of each 25.6 second period), and storing the resulting voltage in a sample-and-hold. Since the heater wire changes resistance with temperature, this eliminates the need for a separate temperature sensor inside the tube. Once the temperature set-point is reached (the voltage from the pot approaches the voltage on the sample-and-hold), the feedback switches to Optimal Mode and alternately samples the two polarized Zeeman split sub-mode signals with their voltage difference being the error signal in the feedback loop, which is driven to zero by adjusting the temperature, and thus cavity length. In fact, from the relative shapes of the red and blue mode cycles, it can be seen that from about the last half dozen mode cycles till just before locking, the tube is actully steadily cooling rather than heating. With the heater located inside the laser tube, the time from power on to a locked condition is typically only about 4 minutes and should also be less susceptible to ambient conditions. In fact, from the relative shapes of the red and blue mode cycles, it can be seen that during most of the time from power on (a cold start) to lock, the laser tube is heating (about 75 cycles), but it switches to steady cooling (about 6 cycles) just before locking.

    Agilent 5517E/F/G

    The primary evidence of the existence of a 5517E is that I have one. :) Actually, I've also seen options lists for the 5517E and 5517FL, and have heard that there is also a 5517G! The 5517E is not documented on the Agilent Web site and all the major Web search engines including Google come up empty-handed. Interestingly, specifications (but not much else) for the 5517FL can be found in the Agilent Laser and Optics Manual.

    My 5517E is in the gold case, so perhaps it is only available as an OEM or "military calibrated" product? It runs at 6.3 MHz so the listed range above of 5.5 to 6.5 MHz is just a guess based on the only spec I have for the 5517E - 1.6 m/s maximum velocity. This is much higher than the 5517D's maximum of 4 MHz resulting in over a 50 percent greater velocity measurement capability. The textbook party line had been that axial Zeeman HeNe lasers above about 4 MHz were simply not viable. The requirement for a higher REF/split frequency likely means that the magnetic field is stronger and thus the total extent of the Zeeman-split neon gain curve is wider (necessitating a shorter cavity length with larger FSR to suppress rogue modes). An informal measurement of the magnetic field of a 5517E did show it to be 5 to 10 percent stronger than that of any other 5517 laser, though there was quite a bit of variability even for the same models (e.g., 5517B). The tube assembly looks almost identical to all the others except that it is about 0.75" shorter up front beyond the section with the magnet and the glass tube itself may be as much as 1.5" shorter - 6.5" compared to 8". In fact, in order to work at all with these high REF frequencies requires a careful balancing act among mirror reflectivity, magnetic field, cavity length, and no doubt many other parameters of the laser design.

    Although I haven't dissected a 5517E/F/G tube, the cavity length can be estimated by measuring the longitudinal mode spacing using a Scanning Fabry Perot Interferometer (SFPI). During warmup, two longtitudinal modes are present during a portion of the mode sweep cycle, so their spacing compared to the FSR of the SFPI provides a good estimate of the FSR of the laser, and thus its cavity length. For 5517E and 5517FL, the mode spacing/FSR is approximately 1.5 GHz implying a cavity length of 10 cm (~4 inches). This is about 20 percent shorter than the cavity length of the 5517A/B/C/D lasers. The HR-end of the tube has a metal cap on it instead of the glass with spring affair of all the other 5517 lasers. Presumably, this allows the mirror to be closer to the start of the discharge and increases the discharge length and thus available output power. At the OC-end of the tube, reducing the space between the mirror and discharge escape hole would further decrease the cavity length. A diagram of the construction (mostly guessed based on this limited information since I don't have X-ray vision!) can be found in Internal Structure of Agilent 5517E/F/G Laser Tube Assemblies. (Compare this with Internal Structure of Hewlett Packard 5501B and 5517B/C/D Laser Tube Assemblies.)

    Several views of a naked 5517E laser are shown in Agilent 5517E Laser Head With Cover Removed.

    This 5517E has the most incredibly complicated Control PCB of any HP/Agilent laser I had see before finding this laser, even compared to the newer digital Control PCB (see below). (I've since found a similar control PCB on a few other recent vintage 5517 lasers.) It includes a SHARC DSP, two Lattice FPLDs, and a lot of other digital circuitry, purpose unknown. They also seem to have gone back to PWM for the heater drive since there is no power transistor on a heatsink, as with the original analog and first digital Control PCBs. However, that collection of inductors visible in the lower left of the photo may be there to clean up the drive to the heater and remove the high frequency switching noise. Since the locking should be basically the same as for the other lasers, this level of complexity is perplexing unless this particular unit was designed to have much better stability - perhaps the "military calibrated" version. Unfortunately, the new Control PCB lacks all the familiar jumpers and the temperature set-point pot, and adds a couple of micro DIP-switches and connectors, purpose also unknown. Aside from the unknowns, everything else is obvious. :)

    Here are a pair of closeup photos of the overall laser:

    Functionally, the 5517E behaves more or less like the other 5517 lasers. The user LEDs are the same but there are 4 LEDs on the control board that I'm sure provide a wealth of information if one knows how to interpret them. My sample takes over 5 minutes for READY to start flashing. READY also stops flashing once or twice for a couple minutes, before it resumes flashing, and then locks after about 9 minutes. Whether these long times and peculiar flashing behavior are normal or indicate some problem, is also unknown. However, with a similar Control PCB and heslthy 5517B tube, the behavior is similarly strange. More on this in the sections: HP/Agilent 5517 Laser Control PCBs and Locking Sequence and Agilent 5517 Laser RS232 Communications.

    The tube is somewhat low power compared to what's normal for other 5517s - about 120 µW locked. But I have no specs on 5517E minimum output power, so with the shorter tube and likely stronger magnetic field, that might be acceptable. In fact, the minimum power spec for the 5517FL is only 65 µW, so the 5517E may be similar. Once locked, it's quite stable with minimal drift in REF frequency. Given the huge amount of computation power available, it may count mode sweep cycles instead of using a temperature set-point (or in addition), and might also adapt automagically to a replacement tube - or require a factory upload of tube parameters via one of those unlabeled connectors!

    The high REF frequency of 6.3 MHz works fine with my home-built SGMD1 measurement display, but comes up as "Laser Fail" on a 5508A. This isn't all that surprising since 6.3 MHz is almost twice the maximum REF frequency of the 5518A for which the 5508A was intended. However, the same type of control PCB with a 5517B tube locks fine but also fails keep the the 5508A happy, so it is more likely due to wimpy line drivers or something like that. :)

    This montage of Agilent 5517FL Laser and Components shows views of a 5517FL in various stages of disassembly. (Sorry about the photo quality - I do not have this laser.) It had a listed output power of 160 µW and REF/split frequency of 7.12 MHz. (The minimum specs are 65 µW and 7.0 MHz, respectively.) The overall construction is similar to that of the 5517E including the overhead-mounted ballast resistor, though the HeNe laser power supply brick is in a fully shielded metal box. The portion of the tube assembly housing the beam expander is longer than the one in the 5517E and the same as that of the 5517B/C/Ds, but of no significance since it doesn't affect anything beyond looks. The design and size of the tube is also similar except that it's a bit more polished with a real Agilent label! But, from the photos, it appears as though the heater resistance adapter found in my 5517E is not used, so the tube heater resistance must be higher. However, this unit had the normal (for recent Agilent lasers) Newer Digital Control PCB rather than the fancier one found in the 5517E. Another 5517FL had the fancy Newest Digital Control PCB, and indeed, no heater resistance adapter. Photos may be found in the Laser Equipment Gallery (Version 3.18 or higher) under "Hewlett Packard/Agilent HeNe Lasers".

    HP/Agilent 5517 Laser Construction

    The following also applies to the 5518A and the 5519A/B, though I rather doubt that they have equivalent digital Control PCBs like the 5517s.

    These lasers consist of the laser tube assembly, potted (brick) HeNe laser power supply, beam sampler, connector PCB, and Control PCB. mounted on a an metal chassis. Any of the parts can be replaced in under 5 minutes using common tools, with only minimal or no adjustment or alignment.

    Laser tube assembly:

    All of these consist of the actual glass HeNe laser tube potted with some sort of rubbery material inside its Zeeman magnet, beam expander, and adjustable waveplates. The heater/cathode is attached via a 2 pin plug while the anode has its own single pin high voltage connector. The HeNe laser tube ballast resistance of about 100K ohms is conformal molded into the silicone insulated HV cable. The bifilar-wound heater inside the laser tube has a typical resistance (cold) of 8 ohms on all tubes except the 5517E. When at operating temperature, the resistance is spec'd to be 1.285 times what it is when cold. For the 5517E (and possibly the F and G models as well), the resistance (cold) is around 4 ohms.

    Only 4 screws hold the tube assembly to the chassis for lasers in the small cases (all the 5517s except the 5517A, as well as the 5501B). One or two will be flat head screws which provide either a fixed axis for horizontal (pan) alignment, or self alignment (no adjustment permitted). All of these tube assemblies appear physically identical, except for the 5517E (and probably 5517F/G) which are slightly shorter. (They, of course, differ with respect to the REF/split frequency.) The larger tube assemblies found in the 5517A, 5518A, and 5519A/B mount with 3 bolts and have machined alignment pins so no adjustments are needed or possible. They, too, are physically identical except for one small area of the casting that needs to be cut away if installing a 5517A tube into a 5519A/B laser.

    HeNe laser power supply:

    Very old (perhaps roughly pre-1990) lasers used Laser Drive model 111-Adj-1 HeHe laser power supplies which had adjustable current via a pot on the laser connector PCB. At least most of them did. I did find one that had the same part number but no third wire. All later versions use VMI power supplies with a fixed current of 3.5 mA. However, the pot is still present on the connector PCB even on lasers made after 2006 (and probably to the present day), but does nothing.

    There are three versions of the VMI power supplies used in these lasers. The two older ones (VMI PS 148 and VMI PS 217) have the same HP part number of 0950-0470. The switchover came around the year 2000. The newest one (VMI PS 373) has an Agilent part number of 0950-4459 and is found on all recent lasers, at least since 2004. I know that going from the 148 to the 217 reduced the residual current ripple from over 3 percent to less than 1 percent because I measured it. I do not know what changes were made in the 373, nor what other differences there may be between these models. VMI claims it is proprietary information. Can you believe that? :) However, I have de-potted a dead model 373 and have reverse engineered its schematic. See the section: VMI 373 HeNe Laser Power Supply.

    With few exceptions, the only defective power supplies I've ever found in HP lasers were nearly all Laser Drive 111-Adj-1s. And one type of failure may result in the adjustment pot having no effect with the power supply pumping way excessive current (like 6 or 8 mA) through the tube. With luck, the ballast resistor catches fire and explodes before the tube is damaged. :( :) One exception was a VMI PS 148 that had excessive ripple, so something in its output filter had blown. But if I hadn't been checking ripple on a bunch of these power supplies, it probably would have gone unnoticed, since as long as the tube stays lit, performance of the laser probably wouldn't be affected in any significant way. However, the additional ripple would make the effective dropout current go up, so a reasonably healthy tube might start sputtering on that supply.

    Beam sampler:

    These consist of a first non-polarizing angled plate to sample a portion of the output beam and a second non-polarizing angled plate to take this and split it between the LCD switch, and reference photodiode. The LCD switch attaches to the Control PCB via a 4 pin connector - 2 pins for the LCD drive and 2 pincs for the photodiode behind it. The reference photodiode is actually mounted on the Control PCB and simply pokes its head into the beam sampler assembly. Beam samplers for all model lasers appear to be identical and interchangeable.

    Connector PCB:

    Aside from the Mil-style connector to the outside world and the 24 pin connector to the Control PCB, this has some filter capacitors; fuses for +15 VDC and -15 VDC; and the Power, Laser On (really same as Power), and READY LEDs. The one pot does nothing except for really old lasers with the Laser Drive 111-Adj-1 HeNe laser power supply brick.

    Very old lasers had a case interlock switch to disable the laser tube from being power if the covers were removed, and a service switch to override this. :) Both of these switches have been eliminated, though the PCB pads and wiring for them are still present, but bypassed.

    Control PCB:

    Since the introduction of the 5517A laser through the early 2000s, all the Control PCBs in these lasers were based on simple TTL logic for timing and an analog feedback loop. The one in the 5517A is physically larger and not interchangeable with those in the small lasers, but is nearly identical electrically. The main functional difference is the additiona of circuitry and a connector so it can also be used in 5518A and 5519A/B lasers. There had been virtually no change in the design over 15 years or more, except that a modification to the internal REF receiver makes newer lasers require somewhat higher optical power to lock than older ones.

    There are no useful indicators on the Analog Control PCB, only two wimpy LEDs that duplicate the functions of the LASER ON and READY LEDs on the backpanel.

    But since sometime after Agilent was created, at least two versions of a Control PCB with much more complex digital circuitry appeared. These still have a substantial number of analog parts like op-amps and comparators, but FPGAs/FPLDs and/or a reasonably high performance microprocessor or DSP have replaced discrete logic. If one wants to count transistors, I bet the digital Control PCBs have 1000 times the number of transistors as the Analog Control PCB! The PCBs are also almost entirely surface mount, with parts on both sides (at least for the newest version).

    The first of these digital Control PCBs (which I call the "newer digital control PCB) is electrically and physically interchangeable with the older Analog Control PCBs. It is based on a Xilinx FPGA and should be very reliable. But a failure for any reason other than an obvious problem like a blown fuse bad DC regulator with no underlying cause would likely render it non-repairable except by Agilent or an authorized service center since it's then just a black box with no real way to easily troubleshoot. A service manual may exist but I've never seen one. And even if it did, sophisticated test equipment and a surface mount rework station would be required to have any chance at repair. However, this version has all the same jumpers and temperature set pot, so normal testing and adjustment is similar to that of the Analog Control PCB. The solution would then be to simply swap in a known good board (either version).

    It's not clear what, if anything, the digital controller adds to the laser, other than to make it more proprietary and difficult to service. After all, features are not being constantly changed or added, nor will there be security issues due to computer viruses - it doesn't run Windows! :) So, periodic firmware upgrades and bug fixes really aren't required. Whether this digital Control PCB is now used for all Agilent 5517B/C/D lasers or only for those with "military calibration" or OEM or special requests, I do not know. Swapping in an analog Control PCB resulted in no obvious differences in performance, which would be my suggested method of repair unless there were special requirements. And now (2009) 5517B lasers with these digital Control PCBs have been turning up from 2004, 2005, and 2006, probably removed from service in wafer fabs after degrading to the edge of Agilent specs for REF frequency, or a specific number or hours of service.

    All the jumpers and their approximate locations are identical and the time spent in the major states and time to lock (READY on solid) are about the same as with the Analog Control PCB. The behavior while warming up and after locking is indistinguishable from that of the Analog Control PCB so it really isn't even possible to determine which one is inside the laser without removing the cover. The only obviously similar electronic component common to the two is the large white film capacitor for the feedback integrator, and perhaps the heater driver transistor. It's possible that the objective of this redesign was simply to eiliminate all the older SSI/MSI TTL logic and other obsolete through-hole parts, but that it is functionally identical to the Analog Control PCB with essentially the same logic inside the Xilinx FPGA and linear circuitry in SMT ICs.

    More recently, another type has appeared which I call the Newest Digital Control PCB. I first found one in a 5517E (this model laser isn't documented anywhere except here!) and thought it might be unique to the -E version. But I have since also seen one in a 5517D-C29, manufactured in 2004. (And the Newer Digital Control PCBs seem to have become mainstream around 2004 also, at least in OEM 5517Bs, 5517Cs, and 5517Ds as well.) But the Newest Digital Control PCB seems to be a total redesign, with no effort made to be at all similar to the Analog Control PCB with its jumpers and test points.

    Here are the three types of Control PCBs used in the 5517B/C/D/E lasers.

    (As noted, the one in the 5517A differs very slightly in form factor and has a small amount of extra circuitry for use in the 5518A and 5519A/B. As far as I know, it only comes in the original analog version.)

    The Analog Control PCB and the Newer Digital Control PCB function in a virtually identical manner, requiring about 2 minutes for the READY LED to start flashing, and another 2 minutes to come on solid. And as noted, they also have more or less the same jumpers and testpoints, as well as the temperature set-point pot. The Newer Digital Control PCB may in fact simply be essentially an emulation of the Analog Control PCB using an FPGA and more modern surface mount parts. That same large integrator capacitor is present, though the smaller sample-and-hold caps are not, or at least they aren't the same type.

    The third type of control PCB is not at all similar to the others. It has none of the same jumpers and several different testpoints, no pots, an RS232 port no doubt for setup and testing and almost certainly a digitally-maintained run-time meter, another unused connector, and a large unpopulated header. There are also a pair of micro-DIP switches - and a pushbutton, which I fianally dared to push, and as expected, seems to be master reset. :) There are many SMT parts on the back side of the PCB, including another Lattice FPLD.

    For more, on the Control PCBs and their operation, see the sections: HP/Agilent 5517 Laser Control PCBs and Locking Sequence, HP-5517E/F, and Agilent 5517 Laser RS232 Communications.

    The entire purpose of redesigning the controller more than once is somewhat perplexing. Doing it the second time with such complex digital hardware seems totally nuts, unless it was actually done for some custom high performance application, and it was simply convenient to use in late model 5517 lasers. (Given the likely relatively high cost of components including the SHARC processor and Lattice FPLDs, I find this rather hard to believe though. And it's a multilayer PCB with components on both sides, additional connectors, and other stuff.) The same very limited inputs (a pair of photodiodes sensing the modes through the relatively slow speed LCD switch and another photodiode behind a polarizer generating the REF signal) and outputs (tube heater current) are used in all 5517 lasers so it wouldn't seem to be possible to implement a significantly higher level of frequency accuracy or stabilization no matter what sort of control scheme is used. About the only thing that might be done is to actually compute the REF frequency from the REF photodiode signal and fine tune the lock position to maximize it once the basic stabilization using mode balance has been achieved. The peak of the REF frequency function may be a more accurate means of locating the Zeeman-split gain curve center. But except for NIST-level precision, the analog method is really just fine, so even if this scheme is implement, it's not clear what customers would require it. And from my observations of the REF frequency while locking, it doesn't seem to make any effort to maximize it, but stabilizes at a point much lower than the peak, with the same sort of slow variation once locked as the analog Control PCB.

    So, a combination of several explanations make the most sense:

    1. Components used on the Analog Control PCB were likely becoming obsolete so that some (though likely trivial) redesign would have been needed in any event.

    2. Basing the controller on a digital platform makes it easy to set up and test in the factory. Those extra connectors must have a purpose! (And we now know that one is for the RS232 port.)

    3. Basing the controller on a digital platform makes it virtually impossible for anyone other than Agilent to repair or adjust (such as when a laser tube is replaced). The RS232 port is no doubt part of this, though the unidentified connector next to it probably provides additional status information. Of course, with the amount of computing power on this thing, it should be able to determine the relevant parameters of any laser tube without any need for calibration (and predict the weather at the same time!).

    4. Digital anything is "in" and the complex appearance helps to justify the five figure price tag. :-)

    If anyone has more information on these digital control PCBs, please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    Also see the section: Common Problems with HP/Agilent 5517 Lasers.

    HP/Agilent 5517 Laser Control PCBs and Locking Sequence

    The stabilization technique used with 5517 lasers is really dirt simple: Start by heating the "mirror spacing rod" to a set-point temperature, fine tune the temperature, then switch to optical feedback which forces the two orthogonal Zeeman-split polarized modes to be equal. However, the details vary slightly depending on the specific Control PCB used in the laser.

    Locking sequence with Analog Control PCB:

    This also applies to the 5518A and 5519A/B since they all use the 5517A Analog Control PCB. It is also generally applicable to the 5501B though some details differ slightly.

    From power-on to READY takes around 4 minutes for most 5517 lasers - all those NOT using the Newest Digital Control PCB. Even on the original Analog Control PCB, a state machine based on counters, flip-flops, and gates determines the timing. This may be true of the Newer Digital Control PCB as well, except that the state machine would be inside a Xilinx FPGA. Who knows how the Newest Digital Control PCB with its SHARC CPU implements this algorithm (which tends to take much longer than 4 minutes, reason unknown)! The following is paraphrased from the 5517A manual, which assumes the Analog Control PCB implementation. (All timing is approximate as the main clock is a 555 timer on the Analog Control PCB!):

    Thus, under normal conditions, the laser will be locked and ready to make a measurement (approximately) 150 seconds after the READY LED starts flashing. Note that the only check to make sure the laser is locked is that the REF signal is present. Since this only occurs for a small percentage of the entire longitudinal mode sweep cycle, REF will not remain on for long without active feedback, so this is a reliable test. The laser will in fact continue to repeat the above sequence forever if REFON is not detected. Typically, this will occur when the output power from the laser tube has declined to below the REF detection threshold of the internal optical receiver after years of hard work. However, some marginal lasers will go through the sequence several times when powered up as the output power from the laser tube gradually increases with warmup until the amplitude of the difference frequency signal exceeds the REF detection threshold.

    Locking sequence with Newer Digital Control PCB:

    The Newer Digital Control PCB has 3 yellow state LEDs near the top right corner of the large Xilinx chip. These provide some information about where the controller is in the warmup process and they have a 1:1 correspondance with the major modes of the Analog Control PCB. I'm not sure the times in each state are identical for the two but they are close. Here is a rough chart of their behavior for a normal 5517C laser:

       Time    READY   State    Comments
     --------------------------------------------------------------------
       0:00             000     Power on (WARMUP Mode)
       0:01             001
       0:02             000
    
       0:07             00X     001-000-001-000 in three seconds.
       0:10             000     Remain here 3 seconds.
    
                The previous two entries repeat approximately 14 times,
                 dependant on time to reach set-point temperature.
    
       1:26   Blinking  010     (HEATER QUALIFIED Mode)
    
       1:32   Blinking  01X     011-010-011-010 in three seconds.
       1:35   Blinking  010     Remain here 3 seconds,
    
                The previous two entries repeat approximately 16 times.
    
       3:03   Blinking  110     (OPTICAL Mode)
    
       3:09   Blinking  11X     111-110-111-110 in three seconds.
       3:12   Blinking  110     Remain here 3 seconds.
    
                The previous two entries repeat approximately 9 times.
    
       3:58      ON     000     (LASER READY)
    

    The blink rate for READY is about 1.5 Hz.

    Locking sequence with Newest Digital Control PCB:

    The Newest Digital Control PCB seems to go through many more gyrations during warmup than either of the others, including several times where READY flashes multiple times separated by a period of inactivity, and then finally flashing READY continuously for two minutes until it locks - the latter being similar to what the Analog Control PCB does. The entire process consistently takes much longer than the 4 minutes typical of a laser with the Analog Control PCB, up to 10 minutes or more. The behavior is not obviously different whether a weak (but functional) laser tube, or one that greatly exceeds minimun output power specs is used, though it may take slightly longer with a below-spec tube. After all this, the end result seems to be exactly the same.

    Here is a chart of the typical startup behavior for a very healthy 5517B tube installed in a (previously) 5517D laser with this newest digital Control PCB:

       Time    READY   State    Comments
     --------------------------------------------------------------------------
       0:00             1111    Power on
       0:01             0000
       0:02.0           0001    0001-0010-0100 sequence in less than 0.5 sec.
       0:02.1           0010
       0:02.2           0100
       0:03    Flash    1100    MSB LED and READY LED flash briefly.
       0:04.0           0001    0001-0010-0100 sequence in less than 0.5 s.
       0:04.1           0010
       0:04.2           0100
       0:05    Flash    1100    MSB LED and READY LED flash briefly.
       0:06.0           0001    0001-0010-0100 sequence in less than 0.5 s.
       0:06.1           0010
       0:06.2           0100
       0:07    Flash    1100    MSB LED and READY LED flash briefly.
       0:08             0100    3x(Step 0, Step 3, Step 5) sent on RS323 port.
       0:15             0101    "LASER" sent out RS232 port.
       1:00             0110
       1:35   Flash 5   X110    1110,0000,5x(1110,0110).
       1:40             0110
       2:14             0000
       2:15   Flash 5   X110    1110,0000,5x(1110,0110).
       2:20             0110
       2:55   Flash 6   X110    1110,0000,6x(1110,0110).
       3:01             0110
       3:35   Flash 7   X110    1110,0000,7x(1110,0110).
       3:42             0110
       4:12   Flash 8   X110    1110,0000,8x(1110,0110).
       4:20             0110
       5:02   Flash 8   X110    1110,0000,8x(1110,0110).
       5:12             0110
       5:40   Flash 10  X110    1110,0000,10x(1110,0110).
       5:50             0110
       6:45   Flash 32  X110    1110,0000,32x(1110,0110).
       7:17   Flash 96  X111    96x(1111,0111).
       8:53      ON     X000    1000,0000,1000,0000,...
    

    The State refers to the 4 SMT LEDs above the upper left corner of the Lattice chip near the center of the PCB. The MSB is green while the three LSBs are red. All times are approximate. "Flash" is just the briefest pulse of light. "Flash n" denotes "n" flashes at a 1 Hz rate with a 50 percent duty cycle The duration for the 1110,0000 state changes in each "Flash n" sequence is relatively short (perhaps 100 ms for each of the two states). The minimum value for "n" seems to be 5, but it tends to increase as the laser warms up. (I'm not positive it's monotonically increasing though.) Once the REF frequency can be sustained by the feedback loop, it continues flashing for 32 seconds, and then switches to state 0111 for 96 seconds prior to becoming READY. Until that time, the READY LED and the MSB state bit track each other almost perfectly. But then, the MSB state bit (1000) continues to flash (but now at about a 1.2 Hz rate) while the READY LED remains on solid, And there is just a hint of the 0100 state bit flickering dimly, possibly the actual feedback loop in operation. :)

    Multiple runs from a cold start may differ slightly in the number of "Flash n" sequences and the values of n, as well as other details, but always take much more than 4 minutes (typical of the analog Control PCB). The very healthy tube will lock in 7 to 9 minutes while a weak but usable one might take 11 minutes or more. In all cases where the laser successfully locks, the last two minutes will be identical in behavior to that of the other two Control PCBs with READY flashing continuously until it stays on solid. A tube that is very weak or with no detectable beat (REF) frequency will result in only occasional very short abortive flashes, and no conclusion (at least not in 15 or 20 minutes).

    One annoying difference between this Control PCB and the others is that the signal level for REF and ~REF seems to be much lower - about 2 V p-p open circuit and less than 1 V (maybe as low as 0.5 V p-p) terminated, instead of more than 5 V p-p, and the 5508A display apparently doesn't accept this as a valid signal. So even if the laser comes READY within 10 minutes (the maximum allowed by the 5508A), it still comes up as "Laser Fail", which isn't recoverable without power cycling the 5508A (which means the laser as well if it receives DC power from the 5508A). However, my home-built SGMD1 display has no problemss. :) I wouldn't be at all surprised to learn that the signal levels are programmable - somehow.

    Here is a chart of the typical startup behavior for the 5517E with its similar Newest Digital Control PCB. The tube is probably below the Agilent spec for minimum power, but locks without problems so the sequence of event is probably not affected singificantly:

       Time    READY   State    Comments
     -----------------------------------------------------------------------
       0:00             1111    Power on.
       0:01             0000
       0:02.0           0001    0001-0010-0100 sequence in less than 0.5 s.
       0:02.1           0010
       0:02.2           0100
       0:03    Flash    1100    MSB LED and READY LED flash briefly.
    
       (Repeat the previous 4 events 27 times.)
    
       1:15.0           0001    0001-0010-0100 sequence in less than 0.5 s.
       1:15.1           0010
       1:15.2           0100
       1:16    Flash    1100    MSB LED and READY LED flash briefly.
       1:17             0100    3x(Step 0, Step 3, Step 5) sent on RS323 port.   
       1:27             0101    LASER sent on RS232 port.
       1:52             0110
       2:25             0000
       2:26   Mode 16   XXX0    16x(1110,0000,....,0000).
       3:35   Mode 12   XXX0    12x(1110,0000,1110,0000,....,0000).
       5:10   Flash 15  X110    1110,0000,15x(1110,0110).
       5:40             0110
       5:48   Flash 32  X110    1110,0000,32x(1110,0110).
       6:10   Flash 96  X111    96x(1111,0111).
       7:46      ON     X000    1000,0000,1000,0000,...
    

    The last part of the sequence is essentially identical to that of the other laser, but the initial behavior differs significantly. This one appears to keep track of the mode cycles, or at least flash the State LEDs in response to them! "Mode n" denotes "n" times where the Zeeman beat is on, at least momentarily. Also note that the READY LED only tracks the MSB State bit near the end. I assume that the AM29F040B (4 Mbit flash memory) is the firmware NVRAM but there is no version number so I don't know that they differ, but they must as everything else on the two digital Control PCBs appears identical including the DIP-switch settings.

    For more on the 5517 laser Control PCBs, see the section: HP/Agilent 5517 Laser Construction.

    Improving the Frequency Stability of HP/Agilent 5517 Lasers

    The following also applies directly to the 5518A and 5519A/B, which are virtually identical to the 5517s. It could also be used with the 5501B, but additional modifications would be needed since these have additional issues.

    While HP/Agilent lasers are very good for their intended metrology applications, they can't compare to the best stabilized HeNe lasers like those from Laboratory For Science, Spectra-Physics, and others. There are issues with both short term variation in optical frequency as well as long term frequency drift. The 3 most significant are probably:

    1. HeNe laser power supply ripple: This shows up as a frequency modulation of the optical frequency at 30 to 50 kHz (and harmonics) with a deviation of up to 1 MHz or more.

      Replace the HeNe laser power supply with a low noise/low ripple type or add an external ripple reducing circuit to its output. The older VMI 148 had particularly high ripple, but even the VMI 217 can stand improvement. I have not tested the older Laser Drive 111-Adj-1 or the newest VMI 373 for ripple. But the VMI-373 already has a ripple reducer built in. See the section: Reducing Ripple and Noise in a HeNe Laser Power Supply with a Switchmode Regulator.

    2. LCD beam sampler and sample-and-hold voltage decay: This shows up as a sort of "hunting" behavior with a period of around 2.56 seconds, resulting in a variation in optical frequency of 50 kHz to 1 MHz or more.

      (I'm assuming modifications to the common Analog Control PCB. I do not know if it's possible to do this easily to either of the digital control PCBs. At the very least, cuts and jumpers would be much more difficult on the dense surface mount PCB. And, since there are so many of the older Analog Control PCBs available, why would you want to!)

      Remove the LCD panel and its photodiode. Drill a hole in the beam sampler PCB and mount a polarizing beam sampler (e.g., polarizing beam splitter cube and a pair of silicon photodiodes) on top of the PCB. (It might even be possible to build this into the plastic housing instead.)

      The schematic for one possible modification is shown in Upgraded Electronics for HP-5517 Lasers 1. This references the part numbers found on the 5517A/B/C/D Analog Control PCB, and probably the 5518A and 5519A/B as well.

      The dual trans-impedance preamp for the photodiodes generates separate S and P mode signals. These feed the "Subtracting-Sample-and-Hold" circuit modified so that when in "Optical Mode" under feedback control, it passes both straight through - no holds allowed! During "Warmup Mode", it must pass the normal heater drive signal. The added preamp can be made from any stable dual op-amp mounted on a little circuit board perhaps stuck on top of U12, the LF13331D quad JFET analog switch, and attached to the photodiodes via twisted pairs. A 1M ohm pot in parallel with a 1 nF capacitor should suffice for the op-amp feedback, providing enough gain for all but the weakest laser tubes. The op-amp, U101, isn't critical - something like an LT082 would suffice. A few cuts and jumpers will be required, but on the wide open through-hole layout of the Analog Control PCBs, that shouldn't be too difficult. An alternative would be to remove the LF13331D and install an IC socket in its place. Then, build a little PCB that plugs into that with the LF13331D and preamp circuitry on it. Add an offset pot and it will then be possible to fine tune the optical frequency. It may not end up pretty, but should work great! It may be easiest to do the modifications in two stages: First replace the LCD and its PD with the polarizing beam sampler and preamp, and confirm that the correct polarizations are selected - the system should lock normally. Then disable the LCD selection logic so that both signals are passed at all times.

      I later implemented a simpler set of modifications as shown in Upgraded Electronics for HP-5517 Lasers 2. This should produce similar results but with a wee bit less flexibility:

      • The LCD and the original photodiode PCB were removed and replaced with a polarizing beam-splitter cube and a pair of photodiodes. However, rather than adding separate pre-amps for each PD, they were wired in parallel with opposite polarity and connected to the same pins as the original PD. The resulting signal is then proportional to the difference in photocurrent. A two pin connector was installed so that the polarity of the PD signal could be determined by experiment rather than trying to figure it out based on theory. :) This would also enable a 5501B tube assembly to be easily tested, as it would require opposite polarity due to the swap of the F1/F2 axes.

      • Three cuts and jumpers reconfigured the subtracting sample-and-hold. All could be done without removing any ICs:

        1. The upper PD input switch (U12D) was disabled (pin 16 tied to ground).
        2. The upper grounding switch (U12C) was permanently enabled (pin 9 tied high through a 10K ohm resistor).
        3. The lower PD input switch (U12A) was enabled by having pin 1 tied to "-WARM UP" (U3-5).

        Wiring of the lower "POWER AMP DRIVE" switch (U12B) was unchanged (enabled by "DISABLE").

      • A circuit for offset was added so the mode balance could be fine tuned. This can vary the net photocurrent by approximately +/-7.5 uA. If I'm ambitious, I'll even drill a hole in the cover so the pot can be twiddled without affecting the thermal environment. As it turned out, the response of the two photodiodes was close enough that the laser locked fine and was very close to optimal with zero offset. But the control would enable the optical frequency to be moved a few 10s of MHz one way or the other.

      • Pin 4, the RESET input of U10, (LM555, main clock) was isolated from its direct connection to +5 VDC and jumpered to "-REF ON" (U1-1). This turns off all the digital circuitry once the laser is locked (READY on solid) to eliminate control PCB-originated digital noise.

      See Modified Beam Sampler and Offset Adjust Circuit for HP/Agilent 5517 Laser for a photo of these modifications.

      The first two sets of changes were implemented first. These worked fine with the locking characteristic after warmup, total time-to-lock, uncertainty in REF frequency, and slow oscillation in REF frequency appearing very similar to the behavior of an unmodified laser. This is actually a rather surprising and unexpected result, so more study will be required. :) A discrete time system has been converted into a continuous time system without doing anything to the loop parameters and there were no dramatic changes the system response. Interesting.... However, later I did confirm that actual locking to the modes occurred almost immediately after the laser entered "Optical" mode (about 100 seconds after READY starts flashing). I also confirmed that if the photodiode polarity was incorrect, it would repetitively pass through the lock point at a rapid rate but never stabilize there. I had expected it to lock to the opposite crossover point of the two modes, but apparently the slope there is so much different that it never latched on, so to speak. Or, possibly it would have locked there eventually but I did not wait long enough.

      For the record, the laser first tested with these modifications was a somewhat high mileage 5517C with a power output of around 240 µW and a REF frequency of around 3.3 MHz, the latter being outside the spec'd range for the 5517C (2.4 to 3.0 MHz). The uncertainty in REF frequency may be 200 Hz or more. The variation starts out with a period of around 16 seconds and deviation of around 0.003 MHz. Over a few hours, it slows down and finally stops (or becomes so long as to not be obvious).

      Some tests:

      • The "Power Amp" jumper was pulled off which prevents the optical signal from affecting the short term stability. The integrator is good enough that the lock-point will remain nearly constant for many seconds. Both the randomness and slow frequency variation in REF frequency were essentially unchanged. The heater driver (Power Amp) could still be acting in a peculiar way though this doesn't seem all that likely.

      • The main clock was disabled after the laser locked by grounding pin 2 of U10, the 555. This made no detectable difference in anything, indicating that these anomalies aren't an issue of either some digital signal still affecting the feedback directly or digital noise being picked up by the analog electronics. EVen though there was no detectable in behavior, modification (4) was installed just for insurance as it was easy enough.

      • A 2M ohm resistor was added in series with the Power Amp input. (The normal input is approximately 2.15M in parallel with 0.22 uF). This made no difference (not surprising given the previous result).

        Then a 220K ohm resistor placed in parallel with the 2.15M ohm resistor. This also had no detectable effect once locked. But, while the time-to-lock didn't change that much, the locking behavior was more rapid after the initial warmup.

      • A low ripple low noise linear HeNe laser power supply (modified SP-248) was substituted for the original VMI-148. This also made no significant difference.

      • An adjustable HeNe laser power supply was also installed to see if running at slightly higher current would have any effect. There may have been a very slight decrease in the randomness but nothing conclusive. At the normal current, this tube does have some amplitude ripple that disappears at higher current, so that may be related.

      • The PBS cube was mounted at a slight angle to reduce the possibility of back-reflections but this also had no obvious effect. The remaining (original) optics of the beam sampler still have perpendicular surfaces that could produce back-reflections, but these are still present when used in the normal way with the LCD and may be worse, since the LCD has several additional perpendicular reflective surfaces.

      Later, I installed the modified control PCB and beam sampler in a certifiably healthy low mileage 5517B (510 µW, 2.32 MHz). Locking was fine and both the randomness and periodic variation in REF were still present, though subtly different. The amplitude of the randomness was slightly lower - perhaps averaging 50 Hz compared to 200 Hz. The period started at about 10 seconds and the deviation was about 0.0045 MHz. However, running this laser with its original Newer Digital Control PCB resulted in essentially identical behavior. However, the deviation as well as the amplitude of the randomness also appear to be affected by exactly which longitudinal mode pair (i.e., exact temperature) at which the laser locks. A later power-on cycle resulted in a deviation of almost 0.01 MHz. Turning my offset control too far (accidentally!) resulted in the laser losing lock and then reaquiring it after the offset was turned back toward center. But the behavior had changed! The deviation in particular had dropped from 0.01 MHz to 0.004 MHz or less. Nothing else was different other than (presumably) where it locked!

      The detailed character of these artifacts remains a mystery. The randomness may in fact be a faster but lower amplitude oscillation in REF frequency superimposed on the larger slower one but it's hard to tell without actually recording it, which I'm not sure I am eager to do. :) Since other evidence suggests that there isn't a corresponding variation in optical frequency to go along with the variation in REF frequency, this peculiarity may be a fundamental characteristic of the Zeeman laser and have nothing to do with the stabilization at all. Or, they may be the result of some sort of etalon effect. The time constant of the slow down in the periodic variation in REF frequency is too long to be anything but thermal in origin. HP/Agilent laser tubes have at least 4 planar uncoated glass surfaces outside the laser cavity and these are not wedged or set at an obvious angle to minimize back-reflections. In addition, the optics of the beam expander telescope and beam sampler have several optical surfaces. Since the structures these are mounted on are all mostly temperature independent of the controlled thermal environment of the mirror spacing rod, it's possible that one or more are forming some sort of external resonator with its longitudinal modes interfering with the normal lasing process very slightly. Maybe.

      And eventually, I will have to set up the dual laser setup to check the optical frequency stability.

    3. Thermal effects due to ambient conditions: While the laser cavity itself is well thermally isolated from the environment, the outside of the tube and electronics are not. Variations in the temperature of the overall tube assembly can affect gas pressure inside the tube, and thus the optical frequency. Changes in the temperature of the electronic components can also affect the location where the optical frequency is locked. One simple way to remedy this might be to add a small fan forcing air through the case, with a sensor to control its speed based on interior temperature. Then, put the entire laser in a box to minimize air currents. A more sophisticated scheme could use a heater or TEC(s) to actively control the case temperature.

    A second order effect is external magnetic fields, but this really shouldn't be significant unless other Zeeman lasers are living nearby, or you want to run this inside an MRI machine. :) And for the purist, air pressure and seismic activity also affect optical frequency, but the three modifications described above should reduce the short and medium term (up to days, probably not years) variation by more than an order of magnitude. Long term drift of optical frequency will be dominated by changes in the laser tube gas pressure and fill ratio from use, and this can't be easily controlled. But periodic diddling with the offset pot can compensate for those. :)

    Agilent 5517 Laser RS232 Communications

    The following likely only applies to the Newest Digital Control PCB. I haven't poked around on the Newer Digital Control PCBs yet but there doesn't seem to be a connector that's obviously suitable for RS232 on that one.

    And, no, there is nothing labeled "RS232 Port", even on the newest digital Control PCB. But, there was a header a with a suspiciously appropriate number of pins (10) near the SHARC chip, so I started looking at voltages, and sure enough, pin 3 on the header had -9 VDC on it, and was occasionally pulsing to 0 V. So, I made up a cable to my ancient Kiwi laptop, and sure enough, there was ASCII being spit out at 9600 baud! :)

       Header Pin   DB9 Pin   Signal
     ----------------------------------------------------
           3           2      Data from 5517 (transmit)
           5           3      Data to 5517 (receive)
           9           5      Ground
    

    The "DB9 Pin" is the result of using an IDC cable wired directly to a DB9 connector. These may be salvaged from old PCs as they are often used to attach the mainboard to the rear panel. The pin numbers will be the same on the PC (not swapped). It's 9600 baud and full duplex (the laser echos characters typed). I have no idea about start and stop bits and parity, but suspect they don't much matter.

    The few interesting things I've discovered so far are:

    Perhaps flipping one of those DIP-switches will put it into Verbose mode, but picking the wrong one might erase the Universe, so I'm not willing to risk that - just yet. :)

    At the very least, the runing time is probably maintained in NVRAM and it would be nice to know how to access that!

    If anyone has more information on these digital control PCBs and their RS232 or other diagnostic port, please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    More to follow, maybe, but these "First Contacts" are encouraging. :)

    HP/Agilent 5517 Connectors

    All versions of the 5517 lasers have the same rear panel connector for power and REF. (The connector on the 5518A adds MEAS and MEAS signal level but is otherwise compatible.)

    HP-5517 power/reference connector J2

       Pin      Function
     ------------------------------------------------------------------
        A       No Connection on 5517 (MEAS signal level on 5508A) (1)
        B       ~MEAS (Not used on 5517) (2)
        C        MEAS    "          "
        D       Signal Return (MEAS)
        E       ~REF (Zeeman beat signal from internal optical
        F        REF  receiver's differential line driver.)
       G,H      Ground
        J       +15 VDC Sense
        K       +15 VDC
        L       -15 VDC
        M       +15 VDC
       N,P      Cable Shield
        R       Signal Return (REF)
        S       Ground
        T       +15 VDC
        U       Cable Shield
    

    Notes:

    1. Pin A goes to the signal level meter on the 5508A display which provides a reading proportional to the current flowing from an internal +5 VDC source to Ground, approximately -2 mA full scale. This is normally provided by the 5518A laser head.

    2. Pins B and C are the ~MEAS and MEAS differential signals, along with the associated return, D. These go from the built-in optical receiver of the 5518A to the 5508A display, but normally only used for test purposes on 5517s. However, they are driven to the de-asserte state by the line driver.

    The 5517 laser head connector looks like a standard MIL-style bayonet lock type but apparently is special built for HP by Amphenol. The part number may be PT06A-14-18PZ but this has not been confirmed. It's possible the mate to these is available direct from Amphenol or a distributor, but probably not a standard item at an electronics distributor. It's actually the same connector as used on the 5500C laser head (PT06A-14-18P), but the keying is rotated 270 degrees, not that that helps much. I did check a Mouser catalog and indeed, the standard connector has the keying rotated by 270 degrees like the 5500C. One mating connector from an original HP cable is labeled: 97 USA/CTI 26SOU 851-06P14-18PX50-44. Searching for any of these part numbers only seems to result in non-stock items with no listed prices - "Ask for Quote". You know you're in trouble when this is involved! :) Used cables for the 5517 are available for around $100 from various surplus dealers and often on eBay. But the standard cables may be 10 or 20 feet long and much more than is needed if the other parts of the interferometry system aren't being used. For power and the reference signal, the mating connector and a few wires should suffice. It would be a pity to chop up an expensive high quality cable simply for the connector. However, it is possible to modify a standard PT06A-14-18P See the section: Making HP Interferometer Cables.

    The pinout for HP cables with a 5 pin power connector (looks like an old PC AT keyboard connector) is as follows:

       Pin      Function                 5 Pin DIN Female
     ------------------------                  2 o
        1       Ground                     4 o     5 o
        2       Ground        
        3       No Connect?               1 o       3 o      
        4       -15 VDC                         ___
        5       +15 VDC                        [KEY]
    

    Common Problems with HP/Agilent 5517 Lasers

    Here are some of the faults that typically occur with the 5517A/B/C/D lasers. Those found in the 5501B, and the 5518A/B and 5519A/B should be similar since the tube design and electronics are virtually identical. The comments with respect to tube issues also apply to the 5501A and 5500C lasers, though the details of the remedies will be slightly different.

    The following are causes that may produce a variety of symptoms:

    For detailed service information, see the section: Additional HP/Agilent Resources. While there is nothing on the 5517 laser specifically, the electronics of the 5518A (part of the 5528A Measurement System) and 5517A is identical except that the 5518A has an additional PCB (the internal optical receiver). And the electronics of 5517B/C/D lasers using the Analog Control PCB is close enough to that of the 5518A to be useful for troubleshooting and repair.

    Notes on the HP-5518A Two Frequency HeNe Laser

    The 5518A is basically a 5517A with an optical receiver added for the return beam. It was the successor to the 5500C intended for use in the 5528A Laser Measurement System along with the 5508A Measurement Display. (The 5518A has the same REF frequency specifications as the 5517A and can be used exactly like a 5517A laser by ignoring the optical receiver.) The chassis, laser tube, and Connector and Control PCBs are identical to those of the 5517A. An additional optical receiver PCB which plugs into the control PCB is added inside the front of the laser, and the front bezel and shutter assembly differ for the 5518A.

    There are two apertures at the output-end of the laser. The top one is the normal laser output, with the usual control wheel for a large opening (normal), small opening (alignment), and closed. It is also the return port for straightness measurements only. A second aperture below it is for the optical receiver. This aperture is used for all measurements except straightness. It has a control wheel for large (normal) and closed (which then has an alignment target printed on the exposed surface). A large Turret Ring behind the apertures has two positions: Straight and Other. For straightness measurements, it inserts optics in the normal laser output aperture to direct a return beam there to the optical receiver, and a microswitch is activated to change the gain of the optical receiver. (The laser output power is also reduced somewhat in this position, so the optical receiver needs to be more sensitive.) There are also "Laser ON" and "Signal" LEDs on the front bezel. Laser On is the same as the LED on the back panel. Signal is lit when there is enough of a return beam to the optical receiver to be useful.

    Several photos of a 5518A laser head can be found in the Laser Equipment Gallery (Version 2.42 or higher) under "Hewlett Packard/Agilent HeNe Lasers".

    The 5518A I have includes one additional component, not present in any other HP/Agilent laser I've seen, and that is a shield or cover surrounding the area of the beam expander, purpose unknown.

    I have reverse engineered the schematic for the Optical Receiver PCB shown in Photos of HP-5518A Optical Receiver PCB. See HP-5518A Optical Receiver Schematic. Most of the component designations are arbitrary since very few had anything on the artwork. Although it performs a function similar to that of external optical receivers like the 10780C, the circuit is considerably simpler and nearly identical to that of the reference receiver on the Control PCB. The built-in photodiode can be seen below the hole through which the output beam passes. The two pin header attaches to the microswitch in the current assembly that selects gain based on whether it is set for "Straight(ness)" or "Other". The gain is increased in Straightness mode since the outgoing beam passes through a non-polarizing beam-splitter and the return beam reflects off of it

    I would also like to find the non-HP equivalent of the receiver IC U1, HP part number 1826-0775, listed as 1DA7Q on the HP schematic of the 5517B laser, which (among others) uses the same IC. If anyone has a standard part number and/or datasheet, please contact me via the Sci.Electronics.Repair FAQ Email Links Page. Of course, maybe 1DA7Q was just a random text string intended to be replaced by the actual part number and that never happened! :) A different revision of the schematic shows the manufacturer part number as 1826-0075 which could be another typo.

    Notes on the HP-5519 Two Frequency HeNe Laser

    The 5519A and 5519B are unique among all the (non-obsolete) HP lasers in that they include an optical receiver for the return beam. (The original 5500C also did but was considered too bulky for many applications. The 5518A also had an internal receiver but it has been superseded by the 5519A/B. All of the lasers include another optical receiver to monitor the output of the laser but it does not have an external optical input.) Like the 5500C and the 5518A, the optical receiver enables them to be easily set up in the field with only interferometer optics and a remote mirror or cube-corner retro-reflector on a tool or other device whose motion needs to be measured. In this regard, the 5519A/B also run on AC power, rather than the +/-15 VDC required by the other HP lasers, so the cable to the electronics only needs the REF and MEAS signals. The 5519A and 5519B differ in the REF (beat) frequency with the 5519A being similar to the 5517C and the 5519B being similar to the 5517D. (The 5518A had a REF frequency corresponding to either a 5517A or 5517B depending on serial number.)

    The case style of the 5519A/B (and 5518A) is similar to that of the 5517A and the tube assembly is very nearly physically interchangeable. The "very nearly" means that a small piece of the tube assembly casting needs to be cut away to provide clearance for the internal DC power supply, not present on the 5517A or 5518A. So, where the higher REF is not needed, a 5517A tube assembly can be installed relatively easily. Presumably, the modification wouldn't be needed for the 5518A since it uses external DC power like the 5517A.)

    The control electronics of the 5519A/B laser heads is functionally the same as all the other thermally tuned HP lasers. But the main PCB is probably not interchangeable among the 5519A/B and the 5517A and 5518A. The 5519A/B seems to have done away with any need for -15 VDC unless it's generated on the main or connector PCBs as the built-in DC power supply only provides +15 VDC. And the 5517A may not have the needed connections for the optical receiver of the 5519A/B.

    Many photos of a 5519A laser head can be found in the Laser Equipment Gallery (Version 2.31 or higher) under "Hewlett Packard/Agilent HeNe Lasers".

    HP-10780 Optical Receivers

    There are several HP/Agilent optical receivers that are used as the detectors for the return (reflected) beam in interferometry measurement systems. The main differences are in their bandwidth and whether they have free-space or fiber-coupled inputs.

    The 10780A is used in interferometry systems using the 5501A, 5501B, 5517A, or 5517B laser heads. It contains a silicon photodiode behind a focusing lens and polarizing filter oriented at 45 degrees, a preamp, a comparator to generate a digital signal from the heterodyne beat of the two polarized modes of the Zeeman-split lasers, and a differential line driver. The primary output is called called "MEAS" and its complement "~MEAS". There is also a Beam Indicator LED which will be lit when there is enough power to produce a reliable beat frequency signal. (This threshold is adjustable.) The 10780B appears substantially similar to the 10780A except that the threshold pot is accessible without removing the receiver cover.

    The pinout of the main connector (J1) is:

       BNC Pin   PCB Pin   Function
     -----------------------------------------------------------------
      1 (LL,F)      1      ~MEAS (Zeeman beat signal pair from
      2 (UL,F)      2       MEAS  differential line driver.)
      3 (LR,M)      3      Return (also BNC shell and receiver case).
      4 (UR,M)      4      +15 VDC
    

    The PCB pins are counted from the edge of the board. I don't know the official designations of the pins on the funny bi-sex 4 pin BNC connector. LL (Lower Left, etc., F for female and M for male) reference the connector with the receiver oriented vertially - with the optical input and Beam Indicator LED at the top. (Rather than buying the way overpriced mating cable, I fashioned a 2 pin female header for power that fits only one way into the male pins, and a separate 2 pin male header for the MEAS signal. These were then glued into a BNC shell. It's not as pretty as the original but it works. I have not found any supplier for the mate that sounded like the cost would be less than an arm and two legs for each one.

    (Note that the 10791, one of the types of cables that is used to connect the 5517 laser heads to DC power and the measurement electronics, has a 4 pin BNC plug like the one that mates with the optical receivers. The REF outputs of the laser are on the male pins with +15 VDC and GND on the female pins. This connector should normally NOT be attached to the optical receiver!)

    There is also an external test-point called "Beam Monitor" on a feed-through pin sticking out above J1. This is the intermediate rectified and filtered signal used for the threshold detection.

    The case should not be connected to the optical metal chassis or Earth ground (I assume for single point grounding noise considerations). Use Nylon screws through the plastic insulated mounting holes at each end.

    The 10780A and 10780B are now considered obsolete as they are not guaranteed to work with 5517C/D and later interferometer lasers over the full specified velocity range since the spec'd upper cutoff frequency is too low (5 MHz). However, HP/Agilent specs are often very conservative. A 10780A I tested using a function generator and LED operated from below 40 kHz to over 8 MHz. It actually would probably be usable down to around 10 kHz but the waveform was somewhat distorted below 40 kHz. The sensitivity as determined by the voltage on the Beam Monitor test-point was down to about 50 percent of what it was at 5 MHz, but some of that fall-off might have been due to my LED/driver. The replacements are the 10780C (free space optical input) and 10780F (fiber optic input, though some of these may actually have the 10780C model number and/or be designated 10780U). The 10780C and 10780F have a guaranteed frequency range from 100 kHz to 7.2 MHz. But for experimental use, when using a single interferometer, or when not requiring high velocity in one direction, the 10780A or 10780B should be fine and typically much less expensive on eBay. :-)

    Any of these HP receivers make good general detectors for optical heterodyne beat signals within their frequency bandwidth since they will operate over a wide range of input optical power from a few µW to 1 mW or more without adjustment. They will also operate with similar optical pulsed signals and work fine to detect the chopped drive of some of my LED flashlights! :) However, note that although the 10780F/U can be used with free-space input, to do so will require a polarizer at 45 degrees to be added, and since there is no lens to focus the light onto the small area photodiode, the maximum sensitivity is much lower than for the other optical receivers.

    HP-10887A PC Calibrator PCB

    This is a full length PC ISA card that is part of the 5529A Dynamic Calibrator. The 10887A has a "Laser Head" input (REF and MEAS) from a 5519A/B laser head, the normal two-frequency heterodyne interferometer signals. But it could also be used with a 5518A laser (and interferometer optics) with an adapter cable, or any other HP/Agilent laser, interferometer optics, and a separate optical receiver, also with suitable adapter cables. The 10887A also has an "A quad B" which I assume to be a quadrature baseband input for use with a homodyne interferometer system.

    Here is a high resolution scan of a 10887A card:

    I'm looking for PC software to make this card useful. :) If anyone has more information on the 10887A, or software they would be willing to share, please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    Making HP Interferometer Cables

    Rather than dedicating special and expensive genuine HP/Agilent cables for each interferometer configuration, I am constructing adapter cables to allow any 5517 (and 5518A) laser or 5508A display to be easily attached to other parts of a system. Each short adapter cable will have the special HP circular male connector at one end and a DB25 female connector at the other. The DB25 will also have additional -15 VDC pins reserved so that it could be used with similar adapters for the 5501A or 5501B lasers as well.

    5517/5508A adapter pin-out

       Mil  DB25
       Pin  Pin    Function
     -----------------------------------------------------------------------------
        A    1     MTR (MEAS signal level to meter on 5508A)
        B    2     ~MEAS
        C    3      MEAS
        D   15     Signal Return (MEAS)
        E    5     ~REF
        F    6      REF
       G,H  7,10   Ground
        J   11     +15 VDC Sense
        K   12     +15 VDC
        L    8     -15 VDC
             9     -15 VDC Sense
            20     -15 VDC
            21     -15 VDC
        M   23     +15 VDC
       N,P 13,16   Cable Shield
        R   18     Signal Return (REF)
        S   19     Ground
            17     Ground
            22     Ground
        T   24     +15 VDC
        U   25     Cable Shield
    
             4     NC
            14     NC
    

    DB25 male:

    MTR  ~MEAS  MEAS  NC   ~REF   REF   GND   -15  -15S    GND   +15S   +15   CSHLD
     1     2     3     4     5     6     7     8     9     10     11     12     13
    
        14    15    16    17    18    19    20    21     22    23     24     25
        NC   MGND  CSHLD  GND  RGND   GND   -15   -15    GND   +15    +15   CSHLD
    

    The Power, REF, and MEAS signals are also brought out to terminal blocks so they can be monitored or easily attached to test equipment like a frequency counter or oscilloscope, as well as the 5501A reference connector.

    For testing 5517s (all versions) and 5518As, the 5508A is used directly with the DB25 adapter. For testing 5501As, a separate DC power supply is used but with the 5508A powered and fed with REF and MEAS via the terminal blocks.

    5517/5508A connectors

    Since the circular MIL-Spec connector for the 5517/5508A is non-standard, the connectors available from electronics distributors need to be modified. The part number is PT06A-14-18P-SR with strain relief. Mouser has them for around $20 in single quanity, which is actually quite good as these things go. The modification turned out to be easier than I had anticipated. The pin block is made of rubber and can be pushed out with a piece of 1/2" copper pipe in a drill press. (Though a very slightly larger cylinder would be a bit better.) First, go around the periphery from both ends with a thin blade which will free most of the rubber from the adhesive used to secure it in place. The pipe fits around the pins without mashing them and only contacts the rubber. Push in increments, making sure the rubber doesn't get too misshapened or skewed in the process. The screw-on strain relief (if present) or some other suitable spacer with a hole in it will be needed under the connector to allow the rubber block to be pressed clear of the shell. Then reinstall in a similar way after aligning with respect to the 5517 or 5508A connector. There will be some damage to the rubber, but it should not affect anything unless you're a purist. Even without any adhesive, it's really snug enough but won't be a Mil-Spec connector that's waterproof. :) It would also be straightforward to fabricate a "punch" that matches the pin pattern. That may reduce collateral damage, but doesn't seem to be worth the effort unless 1,000 of these connectors need to be modified.

    And a note about trying to salvage HP cables if all the required connections aren't already present: Forget it. The cover on the laser-end connector consists of a thick rubber boot on top of a hard plastic conformal molded inner core. While the boot can be slit from end-to-end and peeled off, I doubt it is realistic to remove the core without damage to the connector and pins. I gave up after seeing what would be involved since I didn't have any TNT handy. :) So, for example, an ET-319283 adapter cable which has the 5517 connector at one end and a 7 pin LEMO at the other, possibly intended to connect a 5519A/B to a 5508A Measurement Display isn't useful to power a laser since the DC power connections are not present. (The 5519A/B has a built-in switchmode power supply that runs off the AC line.)

    5501A/B connectors

    The 5501A and 5501B use a pair of 4 pin circular connectors. The power connector is standard with a suitable mate being Amphenol PT06-8-4P-SR. (This is also available from Mouser, but is more expensive than the 18 pin connector!) The reference connector has the keying rotated 45 degrees but a similar push out and reinsert approach works, though more care is needed to assure that the rubber doesn't get destroyed. The diagnostic connector (present only on the 5501A) mates with the standard PT06-14-18P-SR. Unless you're into automated monitoring, building a cable for that is probably not worth it. See the sections on the 5501A/B, above, for pinouts.



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Laboratory for Science Stabilized HeNe Lasers

    Laboratory for Science (LFS, now defunct) may have produced the best commercial stabilized HeNe lasers based only on self (internal) reference techniques - those that use feedback based on some property of the laser itself like dual mode polarization and/or Zeeman splitting. While higher precision lasers are available, they depend on external devices like gas absorption cells or temperature controlled Fabry-Perot resonators for the frequency reference. As would be expected, such equipment is much much more complex and expensive and may require an engineer to be shipped along with the laser to adjust and maintain it. :)

    Scans of original product brochures for the Model 200, 220, and 260 lasers, and html versions, as well as general desciptions and a price list can be found at Vintage Lasers and Accessories Brochures under "Laboratory for Science". The brochures include a nice description of the principles of operation and applications considerations in addition to the specifications.

    The following brief descriptions include extensive contributions from David Woolsey (http://www.davidwoolsey.com/).)

    There were three Laboratory for Science stabilized HeNe lasers known to have been produced and sold:

    All three models had the same size power supply/control box but the laser head for the Model 260 was longer than those for the models 200 and 220. The user controls and general operating procedures are also basically the same for all models.

    A number of features and attention to detail set these lasers apart from most other commercial stabilized HeNe lasers that are or have been available. These are described with respect to each model in the following sections. Unfortunately, clever ideas and implementation are often not the most important factors in determining the success of a product or business.

    Even with the superb technology, not many of any of these lasers were ever sold. The total production run for all the years of the product line from the early 1980s to sometime in 1995 was soemthing like: 300 for the Model 200, 60 for the Model 220, and only 10 for the Model 260. There are references to other models ranging up to 280 in the product literature, but someone who actually worked at Laser for Science throughout the years of ultra stable laser production never heard of them going beyond the discussion stage.

    Ironically, the extensive discussion of retro-reflections in the product brochures may have scared off potential buyers. Nearly half the text in the brochures for the LFS-200, LFS-220, and LFS-260 is related to the effects and mitigation of retro-reflections which some people might interpret as a deficiency with these lasers. Retro-reflections are a problem with all lasers, but especially with lasers designed to have the best stability performance. Other manufacturers tend to simply mention retro-reflections in the operation manual - not the product brochures! - as something to be avoided, but even there, they don't dwell on it.

    Even experienced laser jocks find it hard to understand how reflected light with a power level 1/100,000,000 or less compared to the intra-cavity power can have an effect on the behavior but it definitely can with these type of lasers.

    If anyone has schematics, a service manual, or other detailed documentation for any of the Laboratory for Science lasers (or an actual Laboratory for Science laser!) stached away they no longer need, please contact me via the Sci.Electronics.Repair FAQ Email Links Page.

    The LFS-200 and LFS-220 are described in the order in which I acquired samples.

    Laboratory for Science Model 220 Ultra Stable HeNe Laser

    The Model 220 uses a set of transverse (side-mounted) magnets to produce Zeeman splitting of a single lasing line (at least over a range of positions relative to the center of the gain curve). This results in a beat frequency in the 100 to 500 kHz range. However, since the exact beat frequency depends on the position relative to the center of the gain curve of the single Zeeman-split lasing mode, a Phase Locked Loop (PLL) can be used to park the lasing line very precisely, much more accurately than with the common one or two mode stabilization techniques. Even the very expensive HP/Agilent 5501A/B (and other) metrology lasers used the basic dual mode polarization stabilization technique, despite being Zeeman lasers! The approach of the Model 220 is probably about as precise as possible short of the much more complex and expensive ones using an iodine absorption cell or external Fabry-Perot resonator for the reference. None of the common stabilized HeNe lasers available today like the SP-117A even come close. And while that performance did come at a price compared to a common HeNe laser, it was probably simlar to that of other vanilla-flavored stabilized HeNe lasers. In the early 1980s when the Model 220 first came out, the cost was under $5,000, though a brochure from 1992 shows that the price had increased to $6,750. But, given the level of performance, they could probably have charged 2 or 3 times that price - and sold more of them! :)

    Although LFS is now out of business, other companies do offer transverse Zeeman stabilized HeNe lasers. One example is NEOARK (Japan).

    Among the features and attention to detail that sets the Model 220 laser apart are:

    The well known commercially available HeNe lasers I'm aware of implement very few, if any of these. And note that many duel frequency Zeeman like the 5501A/B, 5517, and others, use simple dual polarization mode stabilization techniques despite their being Zeeman lasers.

    Scans of an original product brochure for the LFS-220 can be found at Vintage Lasers and Accessories Brochures under "Laboratory for Science". A much more compact html version is at Model 220 Ultra Stable Laser Brochure. The brochure includes a nice description of the principles of operation and of course, the specifications.

    The first Model 220 I (Sam) acquired on eBay - S/N 51 - has IC date codes and PCB fab dates between 1981 and 1986. But if the serial numbers started at 1 (or even 10 as has been suggested) rather than 50 and only 60 lasers were ever built, it may be much newer than 1986, possibly between 1988 and 1992. So what if the chips are a bit moldy, they haven't changed in any way other than dropping in price by 1 or 2 orders of magnitude since 1981. :) Maybe LFS bought their chips from PolyPacks (a popular surplus outfit for cheap chips that also no longer exists). ;-)

    While the tube in this laser is weak - around 0.8 mW on a good day which is about half the minimum power spec - this is more than adequate to provide a stable beat frequency signal. Originally, the laser was going through what appeared to be normal warmup, but would not lock after the warmup period and the Lock Level indicator came on. The Model 220 has a headphone jack to permit listening to the PLL error signal (as do the other models as well) and a knob to adjust the PLL gain. And while the knob affected the sound in the headphones, there was little correlation with anything else. It was like a bad SciFi movie sound track! I was thinking there must be electronic problems preventing a stable lock from being achieved. Fortunately, all ICs are standard 4000-series CMOS and common analog parts. Unfortunately, it's not likely that a schematic will be available given how few of these were probably built. Google gas been totally incapable of finding much of any useful information beyond the brochure for the Model 200 on David Woolsey's Web site and a few journel references citing the use of Laboratory for Science lasers for such-an-such research.

    However, a miracle happened. Someone sent me the user manual for the Model 220 and lo and behold, that empty socket under the controller I had been pondering since acquiring the laser needed a jumper plug to complete the internal signal paths! It provides access to all the critical input and outputs of the internal architecture of the Model 220 controller with the intent to permit the use of an external frequency reference, remote control and monitoring, and other advanced functions. The jumper block must have either fallen out in shipping, or the previous owner had been using the remote hookup and kept the cable. No wonder it didn't work. Nothing was connected together! So there were electronic problems of sorts. :)

    With the default jumper plug constructed and installed, *everything* started working in a manner that actually made sense. The Mode LED went on and off as the modes cycled in the HeNe laser tube during warmup and the headphones produced a satisfying chirp a couple times during each mode cycle. When the 30 minutes or so warmup time was completed, the laser locked instantly!

    The sound from the headphones is nearly pure white noise and the beat frequency appears rock stable on an oscilloscope and around the 425.8317 kHz it should be based on the PLL synthesizer BCD switch settings of 511. (The frequency is: 3.4 MHz*m/M where M is the switch setting and m is 32, 64, or 128, preselected based on the laser tube to provide the maximum number of possible discrete Zeeman frequencies.) I intend to check it on a frequency counter but have little doubt that will also show the correct frequency with crystal accuracy. Unfortunately, I don't have a spectrum analyzer or an iodine stabilized laser to check it more precisely. The stability should increase is allowed to warm up for longer - 90 minutes is the time to reach spec'd performance. Originally, I thought it might not be working quite correctly due to the sound from the headphone jack having rumbling and other non-white noise components, but I now believe that may have been due to acoustic feedback since I was actually listening using a stereo amp.

    Here are some photos:

    The HeNe laser tube construction is nothing special, at least on the outside. Like the two mode stabilized HeNe lasers, a Spectra-Physics 088-2 or similar tube would work. But the actual tube used by LFS was apparently custom built though. Some, if not all, were filled with isotopically pure Ne20 or Ne22 to provide the narrowest linewidth and/or to select the precise line center, and possibly He3 as well. Later ones were made with a special bore support spider that eliminated the "slip-stick" behavior during warmup of some other designs.

    The waste beam from the HR-end of the tube is used for the reference beat tone. It has a polarizing filter between the tube and the photodiode and a glued-on wedge to make sure the waste beam can't reflect back into the bore. There is an AGC circuit of sorts for the photodiode so that a usable signal can be obtained as the tube ages regardless of a (reasonable) decline in tube power.

    The tube is rather elaborately suspended as can be seen in the photos. The suspension provides some degree of vibration isolation and there is even a fine thread screw (visible on the top of the laser head) to rotate the tube by a few degrees. The complex suspension was designed to minimize stress in the glass envelope and eliminate stick-slip noise due to length changes of the overall tube. It also allowed the tube to be rotated by a 100 pitch screw adjustment without twisting the tube at all. This was desirable to align the tube's birefringence axis (mode orientation) precisely with the magnetic field.

    The entire laser head is thermally regulated by a temperature controller which is the circuitry on the lone PCB inside the head. The temperature set-point can be adjusted via a pot accessible from underneath the laser head. Power resistors attached to the baseplate on which the tube and magnet assembly is mounted provide the heating and an LED on the rear panel of the laser head shows the amount of power to the heater by its intensity. The baseplate bolts to the outer aluminum case with close-fitting end-plates. Although perhaps not obvious from the photos, the wall thickness is much greater than that of most other HeNe lasers.

    There is also a rather elaborate transducer attached to the tube. While serving a similar function to the heaters on many mode stabilized lasers, the design was optimized for fast response. Power to this heater is what is controlled by the PLL responsible for locking the laser.

    The transducer consists of a dense "zig-zag" run of copper wire about 3.5 inches long Epoxied directly to the outside of the glass tube envelope. The wire is oriented (back and forth) along the long axis of the tube, *not* as a helix or coil (it is not an inductor). When a current is run through the winding the wire heats up and immediately pulls (stretches) the glass with it. The response bandwidth is something like 10 kHz since the length change between the mirrors did not have to wait for the glass to heat up. With the wire arranged along the tube axis all of its change in length was in the intended direction - unlike with a the the more common coil arrangement.

    With a simple coil, the initial change in dimension when current is applied is an increase in winding *diameter* which pulls the glass with it (expands the tube diameter) and causes an initial *shortening* of the tube. The shortening is followed by a lengthening as the heat from the transducer diffuses into the glass. This is not a good way to make a fast feedback loop. Also unlike other heater schemes generally used, with the wire directly attached to the tube glass, there is nothing in between to limit the response as with taped on thin-film heaters.

    On the anode-end of the HeNe laser tube (the front of the laser head and output) is a collar with two LEDs on it and a trim pot. Only the anode wire connects to this collar. One LED is lit when the tube is first turned on. Inside the collar is a temperature regulator for the output mirror. There is a small amount of internal reflection in the mirror that gets back into the laser cavity and this is the way it was tamed. There is a thermistor regulated heater in there that uses the laser discharge current for power. The voltage drop across the heater box will vary, but the current through it is held constant. So, the mirror temperature is regulated so that the etalon formed by its front and rear surfaces has a peak covering the neon gain curve resulting in a constant transmission without retro-reflections. For the approiximately 5 mm thick mirror - 7.5 mm optical length - the FSR is 40 GHz, compared to 1.5 GHz for the Doppler-broadened neon gain curve. So, the peak is rather broad in comparison, but keeping it centered helps long term stability.

    The rear mirror had a simple prism made of cover glass that was Epoxied onto it so that the internal reflection was removed by putting it off axis. The Epoxy was made to be thicker at one side than on the other by supporting one side of the cover glass with little tabs of tape. This method couldn't be used on the output mirror.

    When the output window is under proper thermal regulation both of the LEDs on the thermal regulator enclosure should be half lit. The upper one lit means heating and lower one lit means cooling. The pot adjusts the temperature set-point.

    And note that neither anode or cathode is at ground potential! Don't ask how I (Sam) found this out. :( :) This was apparently for noise suppression. Grounding one end of the tube will risk inserting some 60 Hz hum onto the tube current through ground loops and such. Talk about paying attention to every last detail!

    The HeNe laser tube is driven by a linear power supply with totally exposed components once the controller cover is removed. Not even a plastic shield! It is the typical voltage doubler with parasitic voltage multipler for starting. Four power transistors provide current regulation in the cathode return. While at first glance it looks similar to many other linear power supplies of the early 1980s, it was designed to put out 5 mA at 1,200 Volts with a supply ripple of about 1 mV! That gives it a SNR of around 127 dB. This was necessary in order to reduce the very small fluctuations in laser power output due to supply ripple, and their corresponding phase noise, to a minimum. This was somewhat tricky to do back then. Specifically, the current regulation control circuit has better components and additional filtering compared to common commercial HeNe laser power supplies. The PCB traces were also apparently arranged to minimize pickup of hum and noise from the nearby power transformers. A partial schematic I traced of the Model 200 HeNe laser power supply can be found in the section: Laboratory for Science Model 220 Laser Power Supply (LS-220). I still need to determine the details of the current regulation circuit (lower right in the schematic) but it's diffiult to make out because the PCB can't easily be removed from the controller case.

    And speaking of details. There are some zener diodes in the power supply. If they are clear glass, room light getting in via the ventilation slots will end up modulating the power supply current, so they should be painted or replaced! Mine has the silver painted variety so I guess it's OK.

    The controller has two PLLs. One is used as a frequency synthesizer to produce a highly stable reference derived from a 3.4 MHz crystal. The reference frequency may be set via 3 rotary BCD switches accessible through holes in the case. The other PLL then locks the Zeeman beat to the reference once the laser has reached operating temperature (about 1/2 hour). Thus, the reference determines the exact place on the neon gain curve where the laser will operate. (A little typewritten note on the unit I have states that the center of the Ne20 lasing line corresponds to a setting of 511.) So, maybe my laser tube is filled with isotopically pure gases.

    There are 3 indicators on the front panel. The "Lock Signal" lamp on the right shows by its intensity, the approximate power to the heater transducer attached to the tube. The indicator on the left is called "Reference" and is on all the time at relatively low intensity. It is a power indicator but at a reference brightness that should be similar to the "Lock Signal" indicator when the laser is optimally stabilized. The LED at the top is called "Mode" and goes on and off during warmup as the modes cycle. When locked, it will be on at partial brightness.

    A switch on the rear panel can be used to override the PLL output and select heater at max or off, to adjust the lock temperature, either because the tube is at too high or too low a temperature for stable locking, or should it lock onto a "bad" point of the Zeeman frequency response function.

    The headphone jack is used not only to check on the laser during warmup and to confirm that stabilization has occurred, but also is a sensitive detector of back-reflections, which may be a destabilizing influence. Effects of optics resulting in back-reflections will be heard as transient tones in the headphones. (The headphone output may also be connected to the "Line", "CD", or "Tape" input of an audio amplifier.) Waving anything in front of the laser is audibly detectable, as are any sort of vibrations including gently touching the laser or even the table it's on, or walking across the floor. If the output is piped through loud speakers, having the volume above a very low level will result in acoustic coupling into the laser tube and a very noticeable increase in audio level as well as a change more toward non-white noise.

    There is also a calibration jack which provides a beat frequency signal and DC power source for the Model 225 Zeeman Beat Frequency Range Register, whatever that is. :)

    For an overview of the operating principles, which seem to track the actual implementation quite closely, see the following patents. (For the model 220, the main patent of interest will be #4,468,773.)

    And a non-LFS patent for a green (543.5 nm) transverse Zeeman laser (though I don't know if it actually uses that term):

    The patents also include a number of relevant references.

    About two months after snagging the first LFS-220, I obtained another one, also on eBay - S/N 36. Its tube is a bit hard starting but has slightly higher power than the first - about 1.1 mW. After replacing 2 transistors and a diode which may have been bad or may have been killed when I accidentally shorted the high voltage to the Mode light bulb socket (don't ask!), it also works quite well. Internal construction appears virtually identical to S/N 51.

    At some point in the future, I plan to combine the beams of the two LFS-220s and record and plot the frequency of the beat signal to determine the actual stability. I'll have to complain to the LFS QC department if they don't meet published specifications!

    I have also built an experimental setup using a normal barcode scanner tube in a transverse magnetic field. While turning this into a stabilized transverse Zeeman laser is unlikely to occur, I have captured some plots of it's behavior. See the section: Two Frequency HeNe Lasers Based on Zeeman Splitting.

    I have acquired a scan of the operation manual for the Model 220 laser but have not gotten permission to make it public as yet. However, much of the same technical information with respect to theory of operation can be found in the brochures at Vintage Lasers and Accessories Brochures and in the patents. In fact, the block diagram in the operation manual is taken directly from Fig. 1 of Patent #4,468,773.

    Laboratory for Science Model 200 Ultra Stable HeNe Laser

    The Model 200 uses the simpler and more common dual polarization mode control technique for stabilization. However, through innovative design and careful attention to detail, it routinely achieves a frequency stability 5 to 10 times better than other commercially available stabilized HeNe lasers. At least if one is to believe their specifications.

    Among the features and attention to detail that sets the Model 200 laser apart are:

    The well known commercially available HeNe lasers I'm aware of implement very few, if any of these except for tube testing, which would be essential.

    At the same time, the electronic implementation (see the schematics) is a bit too simple and could benefit from a few things like an integrator in the feedback loop and bypass capacitors!

    There were also a few LFS-210s which added a 10 turn pot on the back that provided adjustment of the mode position on the neon gain curve and thus the optical frequency.

    Scans of an original product brochure for the LFS-200 can be found at Vintage Lasers and Accessories Brochures under "Laboratory for Science". A much more compact html version is at LFS Model 200 Ultra Stable Laser Brochure. The brochure includes a nice description of the principles of operation and of course, the specifications.

    Both the laser head and controller for the LFS-200 are superficially identical to those of the LFS-220 except for the lack of a tube rotation knob on the laser head. Operation is generally similar as well, including the use of the audio headphones for locating back-reflections. However, the tube lacks the heated OC mirror and of course, the additional rotation hardware. The shutter lever on the laser head selects among NP (Non Polarized), off, and LP (Linear Polarized). (This contrary to the manual which says the latter is CP (Circular Polarized).

    The interior of the laser head also differs in a number of ways. The HeNe laser tube appears to be a bit shorter than the one in the LFS-220 and the anode is at the HR-end. The mode pickoff optics and photodetectors are in a little box behind the HR mirror with their premap mounted on the side. There is an offset trimpot for the mode position accessible from under the laser head. The heaters and temperature controller are mounted on the baseplate as with the LFS-220.

    The controller box is arranged roughly the same way as for the LFS-220 but the locking circuitry is substantially simpler having a total of three 8 pin DIPs: LF412 and LM358 op-amps, and an LM2905 timer, presumably for the warmup delay. But there are 6 pots for adjustment (in addition to the user accessible "volume control" servo gain knob). The HeNe laser power supply is similar to the one in the LFS-220 but several additional high voltage filter capacitors have been added on the Control PCB to zap the unsuspecting. There is also an additional pot, as well as an unidentified object in the vicinity of its control circuit, purpose unknown.

    Here are some photos:

    The designers at Laboratory for Science appear to be more obsessed with retro-reflection or back-reflection (same thing) than at any other stabilized laser company. This is understandable considering the higher level of performance that is being achieved with the higher bandwidth servo system more sensitive to cavity perturbation. For example, while other stabilized HeNe lasers will simply use a polarizing beam splitter or two to separate the modes making sure to angle all reflective surfaces to prevent back-reflection, the LFS-200 has added the QWPs after the polarizers. The optics stack sandwich for each mode visible in the photo of the HR-end of the LFS-200, above, is something like:

      Plexiglas back-plate | Amber filter | Polarizer | QWP | Plexiglas front-plate -> PD
    

    Two passes through the QWP (out and back) result in a 90 degree rotation of the polarization axis so any reflected light is blocked by the polarizer.

    Schematics of Laboratory for Science Model 200 Controller

    Here are schematics for most of what is in the control box including the HeNe laser power supply and stabilization circuitry. There are two sheets:

    There is also a significant amount of electronics in the laser head including the laser head temperature controller and photodiode amplifiers. Reverse engineering those would require ripping apart a laser head - something I'm not planning on doing any time soon.

    Note that there were many engineering changes over the course of manufacturing relatively few lasers, with little if any documentation or revision numbering on the PCBs. So, don't be alarmed if there are discrepancies between the schematics and the PCBs in your laser!

    Some Comments From David Woolsey

    The manufacture date is probably closer to the mid or late 1980s for your LFS-220 S/N 51.

    The tube in the 260 was 15 inches long. It lased on three modes, giving it a more complex inter-combinational beat frequency pattern. About 50% of the power was in the central mode and a polarizer could be used to discard the other two modes since they were polarized orthogonally to it. This would get rid of the beats.

    Some of the tubes were filled with single isotopic neon. Most were not. The isotopic mix did not depend on the model type though.

    The tubes used in some of the later lasers were custom made by Shasta Glass (R.I.P.). These tubes had a specially designed capillary support "spider" that produced no "stick-slip" noise as the tube changed length under regulation. Other than that, there was nothing any different between the tubes used in the Laboratory for Science lasers and the tubes used in supermarket barcode scanners. We did exploit mirror defects that were typical of the type of laser tube though. Some types of sputtering artifacts can make a laser less prone to mode hopping. Also, since the mirrors were imperfect, there was a small amount of birefringence in them that we exploited as well. They were cheap tubes, but with lots of sorting and characterizing. We used about 2/3 of the tubes we bought.

    The transducer was one of the fundamental, and patented, ideas that made the Laboratory for Science lasers better than any others. All the lasers used the same transducer system. One of the other patents was related to the phase locked loop electronics on the Model 220. (See the patent list above.)

    A Model 220 was used by IBM in the first Atomic Force Microscope (AFM). The Model 220 could be used to measure distance changes on the order of 1/20 of an Angstrom right out of the box. Compare that to what the "competing" HP laser could do ("Position/distance resolution down to better than 10 nm") and then compare the price tags.

    NASA bought a 260 for the robot that they made to test the tiles on the Space Shuttle. The robot had a YAG laser to hit the tile with a high power pulse that, due to the resulting thermal shock, would make the tile ring. The 260 was used to detect the ring modes. All this was done without contact or close proximity to the surface.

    If you need a tube replacement, the right thing to do is contact Dr. Seaton. He may be able to supply you with one (even though the Lab is nominally out of business). It'll cost a bit over $1,000 installed, I would guess. There are quite a number of subtle things about tube replacement and it is best left up to someone who has done it before (unless you consider your time to be of very little value).

    Why aren't there other lasers like these available today?

    There are much simpler solutions available now for lasers with a coherence length of a few hundred meters. Distributed FeedBack (DFB) diode Lasers can have coherence lengths of a couple hundred meters, power outputs of many times what the Model 200 put out, cost much less than the Model 200, turn on and stabilize quicker, and don't die as easily when abused. (However, DFB lasers do not provide a self-referenced absolute frequency, as do stabilized HeNe lasers. --- Sam.)

    As for the Model 220, I am not quite sure why nobody is making an equivalent system now. I suppose that there is just no significant demand for 1 mW of optical power with 20 km of coherence length. Also, there is only so much that modern manufacturing will get you in this case because there is just too much "hand tweaking" that went into these lasers.

    LFS could have charged 2 or 3 times as much as they did and not lost sales. There was no place else to turn, short of much more complex and expensive iodine stabilized lasers and such, for the 220 and 260 levels of performance. The Lab almost got involved in making an iodine stabilized system. I think I recall Dr. Seaton claiming that it would have something like 0.01 Hz stability.

    Successor to LFS Lasers?

    I'm trying to gather information to help decide whether to the design of a replacement laser for the LFS-200 or LFS-220.



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Laseangle Stabilized HeNe lasers

    Laseangle RB-1 Stabilized HeNe Laser

    Laseangle was a company that built some stabilized HeNe lasers in the early 1980s. I don't know if they ever actually sold anything commercially but they are immortalized by a reference in a paper to the use of 6 of their systems. Perhaps, the researchers were related to the company founder. :)

    The RB-1 consisted of two pieces. The first RB-1 I saw had laser head S/N 1 and controller S/N 2, so at least two of these systems were built and I had mismatched pieces. However, I have photos (below) of RB-1 S/N 8 with very similar construction, which still looks like someone's science fair project. :) The thing clearly wouldn't be caught dead going out to a paying customer, though it's almost certain that the RB-1 or its successor eventually morphed into the Newport NL-1 (maybe "Newport Laseangle 1"?) as a result of a merger or buy-out. However, I've yet to see an NL-1 (or production RB-1 if there ever was such a thing) in person.

    The RB-1 laser head contains the HeNe laser tube, with wrap-around heater, a beam sampler assembly that diverted all of one polarization to a photodiode and part of the orthogonal polarization to another photodiode, and preamps for the photodiodes. The base is a 3/4 inch thick aluminum slab with a 1/8 inch aluminum cover sealed with foam rubber.

    The HeNe laser tube was from Uniphase, a garden variety model with a length of about 8 inches, which is somewhat unusual, probably rated around 2 mW. A tube length of 6 or 9.5 inches being more common, at least today.

    The beam sampler includes a polarizing beamsplitter cube to extract one of the mode signals and prevent it from reaching the output at all, and a separate angled plate to extract a portion of the orthogonal mode. A pair of EG&G SGD-100A photodiodes (may be similar to the Perkin Elmer FFD-100) fed LF356 op-amps.

    The controller houses a linear DC power supply, standard Laser Drive HeNe laser power supply brick, feedback circuitry, and heater driver. There were controls on the front clearly not for an end-user, like 8 or 10 gain settings and a fine gain control for one of the op-amps, selection of which mode signal to pass to an output, a current meter for the heater, and so forth. People who typically use these things would have no clue of what to do with the knobs and switches. I've yet to see a user manual for the RB-1.

    While the mounting of the HeNe laser tube is somewhat overkill and the beam sampler is a nice solid unit with an adequate number of adjustments, the electronic construction of both the laser head and controller are, to put it politely, a disaster. Everything is on those copper strip prototyping boards, with capacitor upon capacitor added in various places no doubt to tame noise pickup or instability. (Someone must have had stock in a capacitor company!) The designers must have had a goal of using strange and hard to find connectors wherever possible which they did for the separate cables of the photodiode signals (blue multipin) and heater drive (microphone two pin). Power for the HeNe laser tube in S/Ns 1 and 2 came from a standard Alden on the controller but at the laser head had both the medium voltage BNC on top for the positive and the normal BNC on the bottom for the negative. In S/N 8, the high voltage cable is hard-wired into the laser head. Maybe the engineers were getting zapped too often. :)

    Here is a composite photo of S/Ns 1 and 2:

    Here are some photos of Laseangle RB-1 S/N 8 courtesy of eBay seller rdr-electronics:



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Mark-Tech Stabilized HeNe Lasers

    Mark-Tech 7900 Frequency Stabilized HeNe Laser

    Mark-Tech Laser, Inc. is a maker of stabilized HeNe lasers and other laser related equipment for measurement, tooling, and alignment. It's not clear how active they are. The Mark-Tech Web site is somewhat in a state of disrepair and the contact email address is through AOL. :)

    The 7900 is a dual mode polarization stabilized laser essentially similar to the Coherent 200, Spectra-Physics 117/A, and others. It consists of a rectangular laser head which contains the controller and HeNe laser power supply, and a separate box with DC power supplies and possibly a status indicator.

    The specifications and a photo for the 7900 can be found at Mark-Tech Model 7900 Frequency Stabilized Laser. And the 7910 at Mark-Tech Model 7910 Frequency Stabilized Laser

    The HeNe laser tube appears to be a Uniphase 098-2 or similar, 2 to 3 mW. It uses a Laser Drive power supply.

    The one interesting difference between the 7900 and most other similar lasers is that the heater to control the length of the HeNe laser tube is painted or coated on the outside of the tube, rather than being a thin film heater or wound with wire. This should potentially have a more predictable response and thus lower frequency/phase noise once locked.

    During initial warmup, the controller runs the heater at rather high power until a reference temperature is reached, and then closes the feedback loop. Since it doesn't need to wait for the temperature to reach equilibrium, this greatly reduces the lock time to under 5 minutes. This is similar to that of the HP/Agilent lasers which use custom and expensive HeNe laser tubes which have an internal heater wrapped around the bore. Most other stabilized HeNe lasers using off-the-shelf tubes take 10 to 20 minutes to lock. The tube does run rather hot though, but this is probably normal.

    The stabilization feedback is implemented in 2 op-amps with some other stuff to monitor the heater temperature, do the switchover from preheat to feedback mode, and generate status signals.

    The output power when locked on the sample I have is about 0.9 mW. (The spec'd minimum locked power is 0.5 mW.) So, this one appears to be basically in like-new condition even though it has a manufacturing date of 1984 making it 24 years old, with a serial number of 112. And I bet they started at S/N 100! :)

    In addition to connections for +/-15 VDC and ground, there are a pair of status signals from the laser head. One goes high (around +12 VDC) a few seconds after the laser locks. The other is open collector, and turns on at the same time. However, this signal will start pulsing if the lock is interrupted - for example, if the beam to the photodetectors is momentarily blocked. The pulsing continues even after lock is re-established. There is a third signal, also open collector, that is always on. I have no idea what that does.

    Here is the pinout for the circular connector (J5, mating connector is AMP/Tyco part number 206434-1 with possible pin part number 66507-9). The same pin numbering is also used on the internal PCB header:

      Pin    Function               Commecnts
     ----------------------------------------------------------=-------------------
       1     Ground
       2     Ground
       3     +15 VDC                Direct to HeNe laser PSU, +12 V reg elsewhere.
       4     -15 VDC
       5     Ground
       6     Lock/Error (Blink)     OC, on when locked, 2 Hz for loss of lock.
       7     Lock                   Low initially, +12 VDC when locked.
       8     Unknown                OC, on all the time.
    

    (There are two similar circular connectors on this laser but only one of them is wired to anything internally.)

    The same case seems to be used for a fancier Mark-Tech laser as there are obvious tapped hole locations for mounting additional optics and other stuff. This may be for the model 7910 which appears to be an interferometer laser used in measurement/calibration systems. However, unlike those for similar applications from HP/Agilent and Zygo, it is probably NOT a two-frequency Zeeman laser but simply a model 7900 with an internal optical receiver using simple quadrature A/B fringe counting in the interferometer. See Model 7910 Interferometer System.

    Here are two photos of the interior:



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Melles Griot Stabilized HeNe Lasers

    Melles Griot 05-STP-901 Frequency and Intensity Stabilized HeNe Laser

    Melles Griot used to offer several frequency and/or intensity stabilized HeNe lasers. Some had the stabilization circuitry built in to the laser head, only requiring a HeNe laser power supply and source of low voltage DC to run - these were all derived from the Aerotech Syncrolase design. (See the section: Aerotech Syncrolase 100. However, the stabilized laser with the highest performance (and no doubt highest price) was the 05-STP-901 which consists of a controller box and separate HeNe laser head. This is actually a Spectra-Physics 117A system in every respect except for the front panel layout and the color of the paint job. In fact, the circuit boards inside have Spectra-Physics 117 "Fab" and "Assy" numbers! The circuit diagrams may be found in the section: Schematics of Spectra-Physics Model 117A Controller. Everything here is past tense because this laser is no longer offered by either Melles Griot or Spectra-Physics, since taken over by Newport and they now sells a stabilized HeNe laser from REO!

    Here are the optical and stabilization specifications for the 05-STP-901 (from Melles Griot):

     Optical Specifications
    ----------------------------------------------
     Output Wavelength:               633 nm
     Output Power:                      1 mW
     M2:                               <1.1
     Beam Diameter (1/e2):            0.5 mm
     Far-Field Divergence (1/e2):    1.60 mrad
     Polarization:                 Linear, >1000
     Mode:                            TEM00
    
     Stabilization Characteristics - Frequency Stabilized Mode
    -------------------------------------------------------------
     Frequency Stability (1 min/1 hr/8 hr):   +/-0.5/2.0/3.0 MHz
     Power Stability (1 min/1 hr/8 hr):       1.0% rms
     Frequency Offset:                        +/-150 MHz
     Temperature Dependence:                  0.5 MHz/°C
    
     Stabilization Characteristics - Intensity Stabilized Mode
    --------------------------------------------------------------
     Frequency Stability (1 min/1 hr/8 hr):   +/-3.0/5.0/5.0 MHz
     Power Stability (1 min/1 hr/8 hr):       +/-0.1/0.2/0.2% rms
     Frequency Offset:                        +/-50 MHz
    
     Stabilization Characteristics - General
    ---------------------------------------------------------------
     Noise:                   0.05% rms
     Lock Temperature Range:  10 °C to 30 °C
     Time to Lock:            <30 minutes
    

    The specifications for the SP-117A should be similar.

    The unit I acquired is of relatively recent manufacture (as these things go) - 1996. The only major problem I found with it was a dead HeNe laser power supply brick - a Laser Drive unit rated 4.5 mA at 1,600 V, similar to the one in the SP-117. It appears to be a standard model except for a hand-printed label with "0.03 percent noise". So, it's either built with better filtering or is specially selected for this application from standard units. Using an external HeNe laser power supply temporarily allowed the controller to be tested. However, it appears to be much more finicky than the original SP-117 in Frequency Mode and would only stabilize with one of my SP-117-compatible laser heads. It basically ignored one that had a slightly leaky photodiode and my home-built clone, simply turning on the "Locked" LED but not actually doing anything. All three of these laser heads stabilize reliably on the much older SP-117 controller. I suspect that an adjustment of the gain of the photodiode preamps would take care of this - probably just turning it up all the way. So, perhaps I shouldn't be so hard on it. :)

    Switching to Intensity Mode at first resulted in the heater simply turning on. The offset pot had to be adjusted to get the mode signals to be in the required range for the locking circuitry to operate, but then it would lock with perhaps 30 seconds required to settle down.

    Switching from Frequency to Intensity Mode or back caused the Locked LED to flash and the Locked relay to chatter for a few seconds. This seems to be normal behavior, as it happens on every one of these lasers I've seen. (The Locked relay provides a set of SPDT contacts that can be used to control auxiliary equipment, though there is no external connector for it. But a cable could be wired to the PCB pads and snaked out through the ventilation slots on the bottom of the case.)

    Monitoring the heater drive signal on an oscilloscope shows how sensitive this feedback scheme really is. Even playing music at a moderate level evoked a detectable response. Tapping on the concrete floor resulted in an oscillation that took a few seconds to die out.

    The electronic design of the SP-117A and 05-STP-901 clearly has it roots in that of the SP-117 (no A) but in addition to the added circuitry for the intensity mode, the frequency control loop has been upgraded to a pure integrator (with a few additional op-amp stages). The PCB layout is completely new and TL084s have replaced LF347s. But much of it looks like it is unchanged from the SP-117 design. The heater drive is a crude pulse width modulator rather than linear pass transistor. Considering the care with which the PCB is laid out with separate analog and digital grounds and linear everything else, it seems strange to have this source of high level digital noise. And, in fact, the varying heater current results in either thermal or magneticly induced vibration of the tube. This can be detected in the spectrum of the laser output if one looks hard enough. For example, if it is heterodyned with another clean laser. (An upgrade is described later to eliminate this.)The input is +12 VDC from a linear power supply. A source of +9 VDC is provided by a LM317 linear regulator. A 555 timer generates the PWM clock and also the -9 VDC power via a charge pump. These power most of the analog circuitry. Timing delays are implemented using several CMOS monostables. Grrrr. :)

    Also see the sections starting with: SP-117 and SP-117A Stabilized Single Frequency HeNe Lasers for more details of the electronics including complete schematics for both the SP-117 and SP-117A/05-STP-901.

    Adjustment of 05-STP-901/SP-117A and SP-117

    The following also applies without changes to the Spectra-Physics 117A, and includes info for the SP-117 as well.

    For best performance, the controller should be adjusted to match a the specific laser head. (Intensity mode may not work at all if swapping heads without readjustment.) There are only three pots inside, so this isn't that complex a procedure! Remove the cover by taking out the 4 Philips-head screws on the bottom near the feet at the edge of the case.

    First, power up the laser and check the 12 VDC power supply at the mainboard PCB connector. There are two pots on the power supply PCB. The one closer to the HeNe laser power supply-side is the voltage adjust. (The other one is unlikely to need adjustment.) You'll need a tiny right angle flat blade screwdriver or bent flat hairpin turn the pot. Set the voltage for 12 V +0.25/-0.0 V.

    Adjustment procedure for MG-05-STP-901 and SP-117A:

    Pots R9 and R10 (500K ohms) set photodiode preamp gain while R13 (50K ohms) sets balance in INTENSITY mode. All measurements should be made with respect to AGnd (TP7). An oscilloscope is desirable for the INTENSITY mode adjustments but not essential.

    It's possible that an earlier PCB revision of the MG-05-STP-901 or SP-117A may have different parts designations. In particular, the SP-117 - no "A" - has other part numbers but it should be obvious which pots and test points to use.

    FREQUENCY mode:

    1. Power up the laser in FREQUENCY mode and monitor TP3 with respect to AGnd (TP7) using a voltmeter or oscilloscope. The signal on TP3 is the output of the preamp for one of the photodiode channels and should vary from near 0 V to as much as 9 V in a cycle that takes a few seconds (increasing in length as the tube heats up and expansion slows). Adjust R9 so the peak is about 6 V or as much as possible if it won't go that high. Where the laser head is weak and it won't reach 6 V, just set the pot fully CW. The laser seems to lock even with only 2 or 3 V max for the photodiode signals.

    2. Monitor TP6 and do the same for R10, the other photodiode channel. It's more important that the signal amplitudes be equal than maximum. So, for a weak tube where the signal amplitudes can't both be set for 6 V, the pot for the smaller one should be fully CW and the other pot should be set to make them equal.

      The LOCKED LED should not be on at this point (even if it was before the adjustments were made).

    3. Give the system time to stabilize (about 15 to 20 minutes from a cold start). The LOCKED LED should come on and then within a minute or so, the output should be stabilized with the voltages on TP3 and TP6 approximately equal and quite constant. Note that the exact values will change somewhat as the tube warms up and its output power changes, especially if the photodiode signal amplitudes are not equal.

      The system is basically working at this point but for the final adjustments, let it remain this way for at least another hour to give everything time to warmup fully.

    4. Very *Carefully* install a jumper between the base of Q1 (LM395T, TO220 package) and DGnd (TP11). This disables the heater and allows the tube to cool causing the modes to cycle again. (The LOCKED LED may go off or start flashing.)

    5. Check that the voltages on TP3 and TP6 don't go above about 8 V (since the HeNe laser tube may produce slightly more power when it is at operating temperature) - readjust R9 and R10 if required to prevent clipping.

    6. Remove the jumper. The system should reacquire lock.

    7. Fine tune R9 or R10 so that the voltages on TP3 and TP6 are equal, giving the system a few seconds to reach the stable position between adjustments.

    8. If it is desired to use both the S and P polarized modes from the output, the Polarizing Beam-Splitter (PBS) rhombus (a squashed cube!) must be removed from the end of the laser head. Loosen the setscrews at the very end of the laser headn and pull off the end-cap. Remove the two screws holding the PBS assembly in place and pull it out. Replace the end-cap, and tighten the setscrews.

      Use the PBS rhombus or any polarizer to compare the power in the two modes. (There should be a mark on the laser head case to indicate the plane of polarization.) If it isn't adequately equal, adjust R9 or R10 to make it so. Only a series of small adjustments should be needed, giving the system a few seconds to reach the stable position.

    INTENSITY mode:

    This adjustment can be done without electronic test equipment but it's much quicker with at least a DMM and easier with an oscilloscope:

    1. Allow the laser to warm up and lock with the stabilization keylock switch in the FREQUENCY position.

    2. Measure the voltage on TP10 (PWM Ramp) and note its value (minimum and maximum if using a scope).

    3. Monitor the voltage on TP9 (HDrv).

    4. Switch the stabilization keylock switch to the INTENSITY position.

    5. Give the system a few seconds to settle down. If the voltage on TP9 is below the minimum of the ramp or above the maximum of the ramp, slowly adjust R13 to bring it within that range. (If without a scope, set it close to the measured value on TP10.) If the voltage on TP9 remains between the ramp limits, the system has already locked and adjustment may not be needed.

    6. Monitor the voltage on TP16 (Htr On). It should be an approximately 500 Hz PWM signal (not a constant value) whose duty cycle is varying to maintain the laser stabilization. If measuring with a DMM, the voltage should be somewhere between 0 V and 9 V (but not at either extreme) and also varying. It may take a minute or more for it to settle down but if nothing happens, try readjusting R13 slightly higher or lower.

    7. It should be possible to do fine adjustment of R13 to place the INTENSITY mode operating point either equal to that of the other mode, or fairly close to the peak of the gain curve. However, going too close will result in loss of lock and the inability to lock when switching modes. And, when switching on from a cold start where the output power of the laser tube is lower, it may not lock at all in INTENSITY mode until it warms up.

    8. Confirm that the system reliably reacquires lock when switching from FREQUENCY to INTENSITY mode and back again.

    Here is the procedure without electronic test equipment, but a laser power meter (or calibrated eyeball) and polarizer will be required:

    1. Allow the laser to warm up and lock with the stabilization keylock switch in the FREQUENCY position.

    2. Rotate R13 to its counterclockwise until it is at the end (will probably start clicking).

    3. Move the Stabilization keylock switch to the INTENSITY position. The polarized mode will start changing as the controller hunts for a stable position.

    4. Rotate R13 clockwise in 1/2 turn increments, waiting a few seconds after each while monitoring the laser output power in one polarized mode. At some point, the polarized mode will change its behavior and the laser should then lock.

    5. It should be possible to do fine adjustment of R13 to place the INTENSITY mode operating point either equal to that of the other mode, or fairly close to the peak of the gain curve. However, going too close will result in loss of lock and the inability to lock when switching modes. And, when switching on from a cold start where the output power of the laser tube is lower, it may not lock at all in INTENSITY mode until it warms up.

    6. Confirm that the system reliably reacquires lock when switching from FREQUENCY to INTENSITY mode and back again.

    Further adjustment is left as an exercise for the student and your mileage may vary. :)

    Adjustments procedure for SP-117:

    The SP-117 only has FREQUENCY mode, which is functionally identical to FREQUENCY mode of the SP-117A. There are also 3 pots. R11 and R13 (the two 500K pots) are equivalent to R9 and R10 of the SP-117A with R12 (the 50K pot) adjusting the position on the gain curve. I do not know why a similar function to this last pot isn't included in the SP-117A. Or, perhaps it is and my schematics have errors. Errors? Nah. :)

    1. Set R12 to approximately mid-range.

    2. Power up the SP-117 and monitor TP8 with respect to AGnd (TP6) using a voltmeter or oscilloscope. The signal on TP8 is the output of the preamp for one of the photodiode channels and should vary from near 0 V to as much as 9 V in a cycle that takes a few seconds (increasing in length as the tube heats up and expansion slows). Adjust R11 so the peak is about 6 V or as much as possible if it won't go that high. Where the laser head is weak and it won't reach 6 V, just set it fully CW. The laser seems to lock even with only 2 or 3 V max for the photodiode signals.

    3. Monitor TP7 and do the same for R13, the other photodiode channel. It's more important that the signal amplitudes be equal than maximum. So, for a weak tube where the signal amplitudes can't both be set for 6 V, the pot for the smaller one should be fully CW and the other pot should be set to make them equal.

      The LOCKED LED should not be on at this point (even if it was before the adjustments were made).

    4. Give the system time to stabilize (about 15 to 20 minutes from a cold start). The LOCKED LED should come on and then within a minute or so, the output should be stabilized with the voltages on TP5 and TP6 approximately equal and quite constant.

      The system is basically working at this point but for the final adjustments, let it remain this way for at least another hour or so to give everything time to warmup fully.

    5. While alternately monitoring each of the photodiode channels, or better yet, the actual laser output for each polarization, very slowly adjust R12 to make them equal. This places the stabilization setpoint at the optimal location on the gain curve. R12 can also be used to move it elsewhere if desired. Note that this point will change somewhat as the tube warms up and its output power changes, especially if the photodiode signal amplitudes are not equal.

    6. Very *Carefully* install a jumper between the base of Q1 (LM395T, TO220 package) and DGnd (TP17). This disables the heater and allows the tube to cool causing the modes to cycle again. (The LOCKED LED may go off or start flashing.)

    7. Check that the voltages on TP8 and TP7 don't go above about 8 V (since the HeNe laser tube may produce slightly more power when it is at operating temperature) - readjust R11 and R13 if required to prevent clipping.

    8. Remove the jumper. The system should reacquire lock.

    Reducing Electronics-Induced Frequency Ripple in SP-117/A Lasers

    For some applications that require the ultimate in a nice stabilized laser, minimizing optical frequency variation is of fundamental importance. Gross variations in optical frequency are due environment conditions like temperature and pressure, mechanical effects like vibration, and drift of electronic components. But there may be small periodic variations in optical frequency due to the electronic design. The only practical ways to detect this would be either by heterodyning the laser's output with a known clean laser or by testing with a super high finesse Fabry Perot interferometer. The beat frequency would not be a pure tone and would have typical FM sidebands on an RF spectrum analyzer. And the sidebands would also show up on the interferometer.

    There are two sources of periodic modulation of the optical frequency in the SP-117 and SP-117A:



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Newport Stabilized HeNe Lasers

    Newport NL-1 Stabilized HeNe Laser

    Before Newport merged with Spectra-Physics, they were usually thought of as providing components like optics, mounts, and maybe measurement instruments, but not lasers. However, it turns out that they did offer a stabilized HeNe laser in the early 1980s when they were still called Newport Research Corporation or NRC. The NL-1 was probably the only such laser manufactured and sold by Newport. Now (2010), they offer the REO SHL but at most, may simply put their logo on it, if that.

    The Newport NL-1 is the "polished" version of the Laseangle RB-1 described above, so Newport didn't actually design it either. The primary changes compared to the RB-1 are cosmetic with much nicer construction of both the laser head and controller. However, the underlying design appears to be virtually identical. It is a dual polarization mode thermally stabilized HeNe laser with very rudimentary electronics (as if the circuitry can be more basic than those in most other similar lasers!). However, while it doesn't even have automatic switch-over from preheat to stabilized operation, there are controls to set "Servo Gain", which could be used to optimize the locking behavior. It also may be locked to either the low ("red") or high ("blue") side of the neon gain curve, along with a trim-pot accessible from the front panel to set the locking point on the neon gain curve. There are front panel outputs to monitor the two polarized mode signals from the photodiodes and something called "Null Test" which seems to be the amplified difference signal from the photodiodes, as well as a tuning input. And there is a meter to monitor the heater current so it's easy to tell if it is within the stable operating region as thermal equilibrium is reached or the ambient temperature changes.

    The laser head uses the same 8 inch long Uniphase HeNe laser tube as the RB-1, along with the same rather overly elaborate beam sampler assembly. The photodiode preamp with two ICs is now on a real PCB as opposed to a prototyping board, and the power supply filters caps are also on a PCB attached to the rear panel. There are no controls or indicators on the laser head (the interlock plug has been moved to the controller), only the cables (still 3 of them - HeNe laser tube power, heater, and feedback) permanently attached via strain reliefs. The main difference compared to the RB-1 is that the laser head cover is now in three pieces - two thin (but nicely made) side panels and a thick milled top plate! But now 12 to 20 screws need to be removed to get the entire cover off, although only 8 screws (woopie!) for a side panel. :) Of course, there's generally no need to go inside the laser head other than to get zapped on the laser tube anode voltage. :) The one adjustment in there (mode balance) shouldn't really change.

    The controller is in a 3.5 inch high 10 inch deep cabinet that could probably be rack-mounted with a suitable adapter kit. It has to win the award for the largest controller for a commercial dual polarized mode stabilized HeNe laser ever made and is about 90 percent empty space! All of the control electronics are on a PCB mounted behind the front panel, with an attached heat-sink for the heater driver transistors. A simple DC power supply is mounted on one side, the Laser Drive HeNe laser power supply brick is mounted on the other, and the power entrance assembly and 2 pin Jones socket interlock connector are on the back. The entire central area is totally empty! :) The laser head cables attach to the front panel probably only because that's where they were on the RB-1. (Although this would make sense if it were rack-mounted.) As with the RB-1, there is both an AC power switch AND keylock switch. Geez, it's not like this is a 10 kW cutting laser!

    Operation is very straightforward based on empirical evidence, as I have no user manual: Power up with the Heater Current set at an intermediate value like 0.3 A and Servo Gain controls (Coarse and Fine) set to minimum. Allow it to warmup for 10 or 15 minutes or until a full mode sweep cycle takes more than 20 or 30 seconds. (It would have been nice if they had provided mode indicators - a pair of LEDs monitoring the Passed and Blocked outputs. But a laser power would need to be used for this.) Then, turn up the Servo Gain controls to achieve a stable lock. The Fine control appears to be most important. It seems that almost any setting that isn't minimum will result in a stable lock, but if set very low, there will be an offset. So, the feedback loop probably has no "I" term. Coarse (which is a multiposition switch) only affects loop stability in some obscure way - the meter needle tends to oscillate if set too low. Or something. So, either the labeling isn't quite accurate or Coarse is broken. :) But Fine definitely does have a much more dramatic effect than its name implies! No doubt, the Servo controls could be set more optimally by monitoring the short and long term variation in the output power or optical frequency. But locking is very rapid at almost any settings, and the output then becomes rock stable, with only a very slight drift as the entire system reaches thermal equilibrium. Flipping the Red/Blue Lock switch results in a shift in lock position over a few seconds to the opposite side of the gain curve with nearly identical power. If the tube temperature is too low or too high as evidenced by the heater current being near or at zero or maximum (about 0.6 A), it's a simple matter to move it to a different lock point by turning the Servo Fine control to minimum and allowing the tube to cool or heat by a few mode sweep cycles. So, unlike most boring stabilized HeNe lasers, this one does have a few fun controls to fiddle with. :)

    The locked output power on my sample can be set between about 0.6 mW and 0.725 mW using the front panel Tuning trim-pot, though it hits 1 mW during mode sweep. The output power is almost the same for the "Red" and "Blue" lock points. (These probably only differ due to non-ideal components.) There is also a trim-pot in the laser head to adjust the mode balance between the "Passed" and "Blocked" photodiode signals. I do not know what the specs are, but the performance of this laser could be close to the new values. An 8 inch laser tube may not have a rated output power much higher than 1 mW, and the beam sampler inside the laser head diverts 100 percent of the horizontally polarized (blocked) mode and about 10 percent of the vertically polarized (passed) mode to the feedback photodiodes.

    Here are photos of the laser head and controller:

    Many more photos of an NL-1 laser head and controller can be found in the Laser Equipment Gallery, (Version 3.13 or higher) under "Newport HeNe Lasers".



  • Back to Commercial Stabilized HeNe Lasers Sub-Table of Contents.

    Nikon Stabilized Lasers

    And you always thought Nikon made cameras. :)

    Nikon apparently developed their own custom HeNe lasers at some point in the past, though it's not known what the intended applications were. More recently, Nikon has manufactured equipment like wafer steppers but they typically use Hewlett Packard/Agilent lasers for positioning. And other Nikon products such as confocal microscopes and other optical instruments that contain HeNe lasers have used standard models from other suppliers, possibly re-badged Nikon but not made by them or significantly modified for the specific application.

    Nikon NKL-85 Stabilized HeNe Laser

    This is a single frequency stabilized HeNe laser with a rated output power of around 0.5 mW. The rather massive laser head (over 15 pounds) includes the HeNe laser tube, HeNe laser and PZT power supplies (both low voltage DC input inverters), a photodiode preamp using an LM308 op-amp on a circular PCB mounted behind the HR of the laser tube, and what is probably another similar amplifier of unknown purpose in a little metal box near the main connector. The rest of the electronics are in a separate box, including control of PZT power supply output voltage, and logic and analog circuitry for stabilization. Regulation of the HeNe laser power supply current may be there as well. All the DC input power and signals are on a 27 pin circular military-style connector except for a single input via a mini-coax, which looks like it might be amplified and then turned right around and send back down the main cable back to the electronics box without being used inside the laser head! More below.

    The tube appears to be of generally similar design to the one in the SP-119. However, it is not physically identical, so swapping a Nikon tube into an SP-119 head or vice versa, would not be possible. But it's probably functionally equivalent, though probably with a longer active region and definitely longer cavity. Why else make such a fancy piece of glasswork that's mostly hidden with the cover in place? In fact, it's not easy to even see if it has Brewster windows or where exactly the mirrors are located without extensive disassembly, so little about the details are certain!

    Based on the physical characteristics of the NKL-85 tube resembling the SP-119 tube, the presence of a PZT power supply, and the behavior of the output power versus cavity tuning that indicates that there is a Lamb Dip (more below), it's almost certain that the NKL-85 uses Lamb Dip stabilization like the SP-119, the only other commercial stabilized HeNe laser known to do so. What's not likely would be for it to use the equally uncommon gain peak technique, since two peaks are present with a Lamb Dip. But it could use an overly elaborate implementation of conventional really boring single mode stabilization. I only have a laser head at the present time so much remains unknown. Or, it may simply have manual tuning via adjustment of the PZT voltage with no automatic frequency control of any kind! Given what's contained in the laser head, any of these techniques would be possible, but it would be silly to go to the effort and expense to manufacture a custom laser tube with a Lamb Dip and allow it to go to waste.

    The HeNe laser power supply and PZT power supply are both low voltage DC-input high frequency inverters using ICs and transistors - no vacuum tubes, can you believe it?! :) But they are still about 10 times the size of modern equivalents.

    Photos of the NKL-85 laser head can be found in the Laser Equipment Gallery, (Version 3.15 or higher) under "Nikon HeNe Lasers".

    The circular military-style connector is about 1.5 inches in diameter with 1-1/2 coarse threads (not fine threaded and not bayonet). Here is its pinout determined visually with the aid of an ohmmeter and some labels on the HeNe laser and PZT power supply PCBs. The signal names are mostly mine:

      Pin  Color   Function           Description/Comments
     ------------------------------------------------------------------------------
       1   Black   Signal Shield      Shield of gray coax from Aux Box.
       2           NC
       3   Red     +30 VDC            HeNe laser PS +DC power.
           Orange  HeNe Interlock     COM terminal of inner microswitch.
       4   Yellow  HeNe Interlock X   NO terminal of inner microswitch.
       5   White   Signal             Center of gray coax from Aux Box.
       6           NC
       7   Red     PZT INT            PZT power supply INT input.
           Brown   PZT INT            NO terminal of outer microswitch.
       8   White   0 V/+30 VDC RET    Twisted with pin 3/red to HeNe laser PS.
       9           NC
      10           NC
      11   White   PD Out             Center of green coax from PD Preamp PCB.
      12   Black   PD Out Shield      Shield of green coax from PD Preamp PCB.
      13   Black   PZT- Shield        Shield of black coax from PZT-.
      14   Clear   PZT-               Center of b